steve hornby
steve hornby
01:41 Mar-03-2004
equivalent wire IQI sensitivity

If an ASME plague type penetrameter of a thickness of 2% of the specimen thickness under test is used and the 4T hole is the only hole visible is it right to claim 2% IQI sensitivity. Article 2 of ASME 5 in the appendix section appears to bear this fact up in the wire equivalence table however having talked to various people in the UK they seem to indicate that the 2T hole must be seen even if a plague thickness representing 2% is used. The 4T hole would only indicate a sensitivity of around 2.6%. Can anybody give me further guideance or is this a long standing difference in interpretation between US & European thinking


 
 Reply 
 
Sasaki
Sasaki
01:27 Mar-04-2004
Re: equivalent wire IQI sensitivity
US standards define the acceptance IQI sensitivity in term of e.g. "2-2T" in which "2" stands for the thickness of IQI with regard to the "material" thickness of interest and "2T" denotes "the hole size of the IQI" is what I have been understanding until now.

Regards,


 
 Reply 
 
John O'Brien
Consultant, -
Chevron ETC , USA, Joined Jan 2000, 280

John O'Brien

Consultant, -
Chevron ETC ,
USA,
Joined Jan 2000
280
09:45 Mar-11-2004
Re: equivalent wire IQI sensitivity
You are correct and most Europeans do not read the code properly plus they are hung up on the figure 2% sensitivity which does not appear anywhere in ASME.

ASME is quite clear it tells you you must see a specified hole on a specified thickness. The equivelent penetrameter sensitivity table goes on to say that when you do not have the correct hole IQI you can use the one listed in the EPS Table. Simple no calculations involved.


 
 Reply 
 
ilyes
ilyes
03:27 Oct-27-2004
Re: equivalent wire IQI sensitivity
----------- Start Original Message -----------
:
: Two percent sensitivity is the actual IQI, 2% of the tickness, remember you need to see three sides of the IQI, the 2T hole then is the quality level.
: Exerpt from ASME Section V, Article 22
:
: The quality level usually required for radiography
: is 2% (2-2T when using hole type IQI) unless a higher
: or lower quality is agreed upon between the purchaser
: and the supplier. At the 2% subject contrast level, three quality levels of inspection, 2-1T, 2-2T, and 2-4T, are available through the design and application of the IQI (Practice E 1025, Table 1). Other levels of inspection are available in Practice E 1025, Table 1. The level of inspection specified should be based on the service requirements of the product. Great care should be taken in specifying quality levels 2-1T, 1-1T, and 1-2T byഊSE-94 ARTICLE 22 — RADIOGRAPHIC STANDARDS first determining that these quality levels can be main-tained in production radiography.
:: If an ASME plague type penetrameter of a thickness of 2% of the specimen thickness under test is used and the 4T hole is the only hole visible is it right to claim 2% IQI sensitivity. Article 2 of ASME 5 in the appendix section appears to bear this fact up in the wire equivalence table however having talked to various people in the UK they seem to indicate that the 2T hole must be seen even if a plague thickness representing 2% is used. The 4T hole would only indicate a sensitivity of around 2.6%. Can anybody give me further guideance or is this a long standing difference in interpretation between US & European thinking
------------ End Original Message ------------




 
 Reply 
 
william brennan
william brennan
04:10 Dec-13-2004
IQI Wires and vynal plasic
----------- Start Original Message -----------
: If an ASME plague type penetrameter of a thickness of 2% of the specimen thickness under test is used and the 4T hole is the only hole visible is it right to claim 2% IQI sensitivity. Article 2 of ASME 5 in the appendix section appears to bear this fact up in the wire equivalence table however having talked to various people in the UK they seem to indicate that the 2T hole must be seen even if a plague thickness representing 2% is used. The 4T hole would only indicate a sensitivity of around 2.6%. Can anybody give me further guideance or is this a long standing difference in interpretation between US & European thinking
------------ End Original Message ------------

Can anybody point me in the right direction for finding a supplier of wires,plastic and laminators for making IQI's?



 
 Reply 
 

Product Spotlight

Immersion systems

ScanMaster ultrasonic immersion systems are designed for high throughput, multi shift operation in a
...
n industrial or lab environment. These fully integrated systems provide various scanning configurations and incorporate conventional and phased arrays technologies to support diverse applications, such as inspection of disks, bars, shafts, billets and plates. All of ScanMaster immersion systems are built from high accuracy scanning frames allowing for scanning of complex parts and include a multi-channel ultrasonic instrument with exceptional performance. The systems are approved by all major manufacturers for C-scan inspection of jet engine forged discs. Together with a comprehensive set of software modules these flexible series of systems provide the customer with the best price performance solutions.
>

CIVA 2020 UT Module

CIVA NDE Simulation Software is the world leader of NDT Simulation. The UT simulation Module incl
...
udes: - "Beam computation": Beam propagation simulation - "Inspection Simulation": Beam interaction with flaws or specimens The user can simulate a whole inspection process (pulse echo, tandem or TOFD) with a wide range of probes (conventional, Phased- arrays or EMAT), components, and flaws.
>

Silverwing RMS PA - Phased-array corrosion mapping

Eddyfi Technologies integrated two market leading solutions, Silverwing’s RMS and M2M’s instru
...
ments to provide a high-speed, remote access ultrasonic phased-array inspection system with live total focusing method (TFM).
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window