00:49 May-10-1999
Magnetic permeabilities

Hello therr

Does anyone out there have any idea where I can find values for the magnetic permeabilities of steels such as duplex and super duplex? I am looking into the possibility of carrying out MPI, and so would like to know how "magnetic" these materials are. I know that there are methods to carry out such evaluations, but unfortunately, I do not have access to the equipment.

Regards

Ishbel




    
 
 
Kåre Johansson
Kåre Johansson
04:23 May-10-1999
Re: Magnetic permeabilities
: Hello therr

: Does anyone out there have any idea where I can find values for the magnetic permeabilities of steels such as duplex and super duplex? I am looking into the possibility of carrying out MPI, and so would like to know how "magnetic" these materials are. I know that there are methods to carry out such evaluations, but unfortunately, I do not have access to the equipment.

: Regards

: Ishbel
We did some MPI of 2205 some years back with good result. We assumed a permeability of 60% of carbon steel and it worked reasonably well. This is not theoretically correct as
the permeability will depend on both the amount of ferrite and the chemical composition of the ferrite. Watch for false indications in the HAZ due to changes in ferrite content. If you want more exact values yu could look into the proceedings of some of the duplex conferences which have been held. Maastricht (1997), Glasgow (1994), Baune (1991).Steel producers like Sandvik and Creusot Loire should also be able to help.
Kåre


    
 
 
LK
NDT Inspector,
Norway, Joined May 2008, 104

LK

NDT Inspector,
Norway,
Joined May 2008
104
19:24 Jan-11-2012
Re: Magnetic permeabilities
In Reply to Kåre Johansson at 04:23 May-10-1999 .

Good afternoon.

Any changes within last 12 years ? :)
Anybody have more information about MPI of duplex bars?
One customer requested both PT and MT on them.
I have some doubts about efficiency of mag particle testing.

Looking forward for your comments.
LK

    
 
 
John H
NDT Inspector,
Norway, Joined Nov 2012, 1

John H

NDT Inspector,
Norway,
Joined Nov 2012
1
20:24 Nov-14-2012
Re: Magnetic permeabilities
In Reply to LK at 19:24 Jan-11-2012 .

I look forward to more information on this subject.

    
 
 
John
John
02:14 Nov-15-2012
Re: Magnetic permeabilities
In Reply to John H at 20:24 Nov-14-2012 .

All u can think of is a lift test of the material

    
 
 
Michel Couture
NDT Inspector,
consultant, Canada, Joined Sep 2006, 829

Michel Couture

NDT Inspector,
consultant,
Canada,
Joined Sep 2006
829
01:30 Nov-22-2012
Re: Magnetic permeabilities
In Reply to Ishbel Macdonald at 00:49 May-10-1999 (Opening).

Ishbel,

You're not saying wht type of equipment you have or which MPI method you will be using (yoke or mag bench).

Regardless, what you have to remember is that your magnetic field should be strong enough to magnetise your part and show the indication you are seeking. ASTM E709 talks about two way to ensure your magnetic inspection will be succesful is to use a pie gauge and see the artificial defects in it or with a Gauss meter tha will indicate a field of 30 - 60 gauss in the part.

    
 
 
Jianjun Wu
Engineering, NDE
China, Joined Nov 2012, 20

Jianjun Wu

Engineering, NDE
China,
Joined Nov 2012
20
02:22 Nov-22-2012
Re: Magnetic permeabilities
In Reply to Michel Couture at 01:30 Nov-22-2012 .

Recently, I read a book regarding the permeaility of steel materials.

This book specifies the Magnetic characteristic curve of the material which will help us to determine the proper parameters in MT techinque selection. But this book is written in Chinese. However, the symbols and curves could be easily understood. If any one want this book, please leave your mail. Br, Hc are indicated in the curve.

    
 
 
Jianjun Wu
Engineering, NDE
China, Joined Nov 2012, 20

Jianjun Wu

Engineering, NDE
China,
Joined Nov 2012
20
02:29 Nov-22-2012
Re: Magnetic permeabilities
In Reply to Ishbel Macdonald at 00:49 May-10-1999 (Opening).

zoom image
According to the Curve, if Br>0.8T, Hc>1000A/m or (H*B)max>0.4kJ/m3, residual method would be better for this material; otherwise, continuous method will be applied.

Hope this will do some help to you.
    
 
 

Product Spotlight

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

NDT Master Lecturer

In the program both university professors and practitioners will give lectures, which guarantees the
...
oretical depth and practical inside. Academic Director: Prof. Dr. Christian Boller
The following lecturers are not complete: Prof. Tadeusz Stepinski, Prof. Wieslaw Staszewski, Prof. Frank Walther, Prof. Giovanni Bruno, Prof. Gerd Dobmann, Prof. Philippe Guy
>

Exertus Dual 120

The Exertus Dual 120 Projector has the ability to accept Iridium 192 sources or Selenium 75 source
...
s. This projector incorporates design and safety features that make it flexible, compact and lightweight. The Projector is lighter than most of its competitors. It incorporates an improved source channel, based on a new helicoidal design, which makes maintenance easier. The helicoidal design also allows smoother movement of the source assembly inside the device, making it easier for the operator and improving safety. The Projector also has a unique safety feature not found in competitive products. The source assembly locking mechanism is triggered by the source holder capsule at the front of the source assembly, thereby always assuring the operator that the source has returned to the safe position. The Exertus Dual 120 is ISO3999:2004 compliant.
>

PROlineTOP Plug & Play Ultrasonic inspection device

As Plug & Play solution it units all control and operation elements in a small housing and therefo
...
re replaces the typical control cabinet...
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window