Technical Discussions
CHARLIE TOH
CHARLIE TOH
02:07 Jun-10-2005
ASTM E1742-05

With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?


 
 Reply 
 
Mathias
Mathias
08:43 Jun-13-2005
Re: ASTM E1742-05
I believe you have to purchase a new lightmeter. The new requirement applies to the excisting viewers whenever the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used. This is my interpretation.
----------- Start Original Message -----------
: With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
: Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?
------------ End Original Message ------------




 
 Reply 
 
DJ Kallhof
DJ Kallhof
07:40 Jun-22-2005
Re: ASTM E1742-05
Light meters currently available register up to 19,990 lux. That level will only qualify the film viewer to interprate 3.30 density film, according to ASTM E 1742-05 Fig. 2. If I am incorrect thinking this please respond.

----------- Start Original Message -----------
: I believe you have to purchase a new lightmeter. The new requirement applies to the excisting viewers whenever the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used. This is my interpretation.
: : With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
: : Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?
------------ End Original Message ------------




 
 Reply 
 
Mathias
Mathias
00:55 Jun-27-2005
Re: ASTM E1742-05
Yes, seems correct. But I hope you don´t mean currently availabe on the market?!

----------- Start Original Message -----------
: Light meters currently available register up to 19,990 lux. That level will only qualify the film viewer to interprate 3.30 density film, according to ASTM E 1742-05 Fig. 2. If I am incorrect thinking this please respond.
: : I believe you have to purchase a new lightmeter. The new requirement applies to the excisting viewers whenever the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used. This is my interpretation.
: : : With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
: : : Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

NEW! The PragmaPro Instrument Platform

The PragmaPro is based on a modular cartridge technology and supports various NDT instrument modal
...
ities such as UT, PAUT, ECT and many more. This new platform is based on a machined, powder-coated aluminum frame for shock-proofness, best sealing qualities and maximum heat dissipation. This is practical to extend the outdoor temperature range and/or to extend the power injected in the transducers. The PragmaPro is aiming at a very wide range of applications, such as weld scanning, corrosion mapping and composite testing.
>

CIVA 2020 UT Module

CIVA NDE Simulation Software is the world leader of NDT Simulation. The UT simulation Module incl
...
udes: - "Beam computation": Beam propagation simulation - "Inspection Simulation": Beam interaction with flaws or specimens The user can simulate a whole inspection process (pulse echo, tandem or TOFD) with a wide range of probes (conventional, Phased- arrays or EMAT), components, and flaws.
>

Wireless TOFD scanner

Quick, accurate and highly reproducible welds testing. The System operates wirelessly and is compat
...
ible with any type of Windows based Laptop, Desktop or Tablet.
>

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window