09:41 Dec-07-2004
Flatness of ut probe face

Hi ,


In EN12668-2, it is mentioned that to measure flatness of contact probes use a ruler or feeler gauges. The gap over the whole probe face shall not be larger than 0.05mm. Can anyone please advice on the method to be used. Thanks.


 
 Reply 
 
Phil Herman Jr.
Sales, - Manufacture of NDT Reference Standards/Test Blocks
PH Tool Reference Standards, USA, Joined Oct 1999, 79

Phil Herman Jr.

Sales, - Manufacture of NDT Reference Standards/Test Blocks
PH Tool Reference Standards,
USA,
Joined Oct 1999
79
09:35 Dec-07-2004
Re: Flatness of ut probe face
Amry,

It will be difficult using a ruler or feeler gage to determine if the flatness over the probe face is less than 0.05mm (0.002"). To perform this check accurately, you need a dial indicator (aka test indicator). This is a common inspection tool used in the toolmaking trades, that measures flatness and other dimensional features by the amount of movement of the indicating arm. The movement, or variation is displayed on a clock-like dial which reads in MMs or thousandths of an inch. Set up the probe in a small vise or vee-block with the face up, and place the indicator in a surface gage to hold it. Move the surface gage base over a granite surface plate or other flat surface so that the indicator arm travels over the probe face. Set a 'zero' on the indicator by rotating the ring to 0. Observe the minumum and maximum readings and ensure the total indicator reading (TIR) does not exceed the allowable flatness.


----------- Start Original Message -----------
: Hi ,
:
: In EN12668-2, it is mentioned that to measure flatness of contact probes use a ruler or feeler gauges. The gap over the whole probe face shall not be larger than 0.05mm. Can anyone please advice on the method to be used. Thanks.
------------ End Original Message ------------




 
 Reply 
 
DREW
DREW
01:08 Sep-02-2005
Re: Flatness of ut probe face
----------- Start Original Message -----------
: Amry,
: It will be difficult using a ruler or feeler gage to determine if the flatness over the probe face is less than 0.05mm (0.002"). To perform this check accurately, you need a dial indicator (aka test indicator). This is a common inspection tool used in the toolmaking trades, that measures flatness and other dimensional features by the amount of movement of the indicating arm. The movement, or variation is displayed on a clock-like dial which reads in MMs or thousandths of an inch. Set up the probe in a small vise or vee-block with the face up, and place the indicator in a surface gage to hold it. Move the surface gage base over a granite surface plate or other flat surface so that the indicator arm travels over the probe face. Set a 'zero' on the indicator by rotating the ring to 0. Observe the minumum and maximum readings and ensure the total indicator reading (TIR) does not exceed the allowable flatness.
:
: : Hi ,
: :
: : In EN12668-2, it is mentioned that to measure flatness of contact probes use a ruler or feeler gauges. The gap over the whole probe face shall not be larger than 0.05mm. Can anyone please advice on the method to be used. Thanks.
------------ End Original Message ------------




 
 Reply 
 
DREW
DREW
01:08 Sep-02-2005
Re: Flatness of ut probe face
----------- Start Original Message -----------
: Amry,
: It will be difficult using a ruler or feeler gage to determine if the flatness over the probe face is less than 0.05mm (0.002"). To perform this check accurately, you need a dial indicator (aka test indicator). This is a common inspection tool used in the toolmaking trades, that measures flatness and other dimensional features by the amount of movement of the indicating arm. The movement, or variation is displayed on a clock-like dial which reads in MMs or thousandths of an inch. Set up the probe in a small vise or vee-block with the face up, and place the indicator in a surface gage to hold it. Move the surface gage base over a granite surface plate or other flat surface so that the indicator arm travels over the probe face. Set a 'zero' on the indicator by rotating the ring to 0. Observe the minumum and maximum readings and ensure the total indicator reading (TIR) does not exceed the allowable flatness.
:
: : Hi ,
: :
: : In EN12668-2, it is mentioned that to measure flatness of contact probes use a ruler or feeler gauges. The gap over the whole probe face shall not be larger than 0.05mm. Can anyone please advice on the method to be used. Thanks.
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

Robotic laser shearography enables 100% inspection of complex, flight-critical composite structures

An article in “Composites World Magazine” showcases Non Destructive Testing of aero-structures
...
with Laser Shearography. Over the years Dantec Dynamics has supplied many solutions for the aerospace industry. Referring to specific customer projects several of these cases are examined to outline the advantages of using Laser Shearography for automated defect detection.
>

NDTkit RT

NDTkit RT, TESTIA's Digital Radiography software The NDTkit product line software for X-ray analysi
...
s. NDTkit RT is a software benefiting from the Ultis kernel which is dedicated to radiographic image analysis. It offers a set of tools and filtering processes to assist RT operators in finding relevant flaws.
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window