Depth reading

Hi everyone,

Here is my probleme,

I have a plate of 1.5 in. At 0,750in I have a hole of 0,062in.
With my USN58L I can know the distance from my transducer, the sound path and the depth of this 0.062in hole.

But, when I move my transducer to Leg 2 with this same hole I still get the good distance from my transducer and also the good sound path, but the depth reading is 3.18in instead of 0,750in. I use a 45 deg. angle transducer and the thickness of the plate on my machine is set at 1.750in.

Is it possible that my machine can't give me the depth after Leg 1 and that I always have to use formulas.

Thanks a lot,

LouisB



 
 Reply 
 
Ed
Ed
06:25 Apr-09-2006
Re: Depth reading
----------- Start Original Message -----------
: Hi everyone,
: Here is my probleme,
: I have a plate of 1.5 in. At 0,750in I have a hole of 0,062in.
: With my USN58L I can know the distance from my transducer, the sound path and the depth of this 0.062in hole.
: But, when I move my transducer to Leg 2 with this same hole I still get the good distance from my transducer and also the good sound path, but the depth reading is 3.18in instead of 0,750in. I use a 45 deg. angle transducer and the thickness of the plate on my machine is set at 1.750in.
: Is it possible that my machine can't give me the depth after Leg 1 and that I always have to use formulas.
: Thanks a lot,
: LouisB
------------ End Original Message ------------

I don't believe the machine is incapable of calculating it. Try setting the thickness of the block at 1.5" as you stated it was and not 1.75". You may also want to check the actual angle in the material. Last but not least make sure you have the correct velocity entered in the machine. Then lastest but not leastest, make sure you are looking at the correct reflector and not an inclusion in the block. Did you try to plot the hole on the second leg?



 
 Reply 
 
Nigel
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
08:20 Apr-09-2006
Re: Depth reading
Louis

3.18" is the beam path skip distance to the hole (1.5 + 0.75)/cos 45 and you say that you get this distance as the "good sound path" which should be just that - 3.18". The fact that your machine is giving this figure as the depth of the hole cannot be mere coincidence, thus somehow the machine must have been set to give the beam path distance as the depth (maybe through a gate).

For a new and expensive piece of kit the sales rep or tech services of the supplier should assist in untangling your problem.

Good luck

Nigel

----------- Start Original Message -----------
: Hi everyone,
: Here is my probleme,
: I have a plate of 1.5 in. At 0,750in I have a hole of 0,062in.
: With my USN58L I can know the distance from my transducer, the sound path and the depth of this 0.062in hole.
: But, when I move my transducer to Leg 2 with this same hole I still get the good distance from my transducer and also the good sound path, but the depth reading is 3.18in instead of 0,750in. I use a 45 deg. angle transducer and the thickness of the plate on my machine is set at 1.750in.
: Is it possible that my machine can't give me the depth after Leg 1 and that I always have to use formulas.
: Thanks a lot,
: LouisB
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

Varex Imaging Large Field of View (FOV) Digital Detector Arrays (DDAs)

A larger FOV DDA can reduce the space and volume of the X-ray inspection system on the factory floor
...
, enable faster scanning times, better throughput and better resolution images at a lower dose. Customers can also save time and money. With these benefits in mind, Varex Imaging has designed a family of large FOV detectors (4343HE, XRD 1611, 4343DX-I, 4343CT) for our industrial imaging customers.
>

IRIS 9000Plus - Introducing the next generation of heat exchanger inspection.

Representing the seventh generation of the IRIS system, the IRIS 9000 Plus has nearly 200 years of c
...
ombined field inspection experience incorporated in its design. This experience combined with a strong commitment to quality and a history of innovation has made Iris Inspection Services® the undisputed leader in IRIS technology.
>

TraiNDE UT

TraiNDE UT is a virtual tool associated with a signal database which simulates real inspection con
...
ditions for numerous applications (Type A/V1 block, DAC block, welds and plates).
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window