Louis
Louis
03:58 Apr-08-2006
Depth reading

Hi everyone,

Here is my probleme,

I have a plate of 1.5 in. At 0,750in I have a hole of 0,062in.
With my USN58L I can know the distance from my transducer, the sound path and the depth of this 0.062in hole.

But, when I move my transducer to Leg 2 with this same hole I still get the good distance from my transducer and also the good sound path, but the depth reading is 3.18in instead of 0,750in. I use a 45 deg. angle transducer and the thickness of the plate on my machine is set at 1.750in.

Is it possible that my machine can't give me the depth after Leg 1 and that I always have to use formulas.

Thanks a lot,

LouisB



 
 Reply 
 
Ed
Ed
06:25 Apr-09-2006
Re: Depth reading
----------- Start Original Message -----------
: Hi everyone,
: Here is my probleme,
: I have a plate of 1.5 in. At 0,750in I have a hole of 0,062in.
: With my USN58L I can know the distance from my transducer, the sound path and the depth of this 0.062in hole.
: But, when I move my transducer to Leg 2 with this same hole I still get the good distance from my transducer and also the good sound path, but the depth reading is 3.18in instead of 0,750in. I use a 45 deg. angle transducer and the thickness of the plate on my machine is set at 1.750in.
: Is it possible that my machine can't give me the depth after Leg 1 and that I always have to use formulas.
: Thanks a lot,
: LouisB
------------ End Original Message ------------

I don't believe the machine is incapable of calculating it. Try setting the thickness of the block at 1.5" as you stated it was and not 1.75". You may also want to check the actual angle in the material. Last but not least make sure you have the correct velocity entered in the machine. Then lastest but not leastest, make sure you are looking at the correct reflector and not an inclusion in the block. Did you try to plot the hole on the second leg?



 
 Reply 
 
Nigel
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
08:20 Apr-09-2006
Re: Depth reading
Louis

3.18" is the beam path skip distance to the hole (1.5 + 0.75)/cos 45 and you say that you get this distance as the "good sound path" which should be just that - 3.18". The fact that your machine is giving this figure as the depth of the hole cannot be mere coincidence, thus somehow the machine must have been set to give the beam path distance as the depth (maybe through a gate).

For a new and expensive piece of kit the sales rep or tech services of the supplier should assist in untangling your problem.

Good luck

Nigel

----------- Start Original Message -----------
: Hi everyone,
: Here is my probleme,
: I have a plate of 1.5 in. At 0,750in I have a hole of 0,062in.
: With my USN58L I can know the distance from my transducer, the sound path and the depth of this 0.062in hole.
: But, when I move my transducer to Leg 2 with this same hole I still get the good distance from my transducer and also the good sound path, but the depth reading is 3.18in instead of 0,750in. I use a 45 deg. angle transducer and the thickness of the plate on my machine is set at 1.750in.
: Is it possible that my machine can't give me the depth after Leg 1 and that I always have to use formulas.
: Thanks a lot,
: LouisB
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

Robotic laser shearography enables 100% inspection of complex, flight-critical composite structures

An article in “Composites World Magazine” showcases Non Destructive Testing of aero-structures
...
with Laser Shearography. Over the years Dantec Dynamics has supplied many solutions for the aerospace industry. Referring to specific customer projects several of these cases are examined to outline the advantages of using Laser Shearography for automated defect detection.
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

Wireless TOFD scanner

Quick, accurate and highly reproducible welds testing. The System operates wirelessly and is compat
...
ible with any type of Windows based Laptop, Desktop or Tablet.
>

Combination of Digital Image Correlation and Thermographic Measurements

The combination of measuring results from the digital image correlation (ARAMIS, DIC) and temperat
...
ure measuring data from infrared cameras permits the simultaneous analysis of the thermal and mechanical behavior of test specimens in the materials and components testing field.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window