07:58 May-26-2006
Re-calibration of EMCOL 6K Portable Current Flow Magnetic Particle Inspection Unit


We are using the Current Flow Magnetic Particle Inspection Unit. Its re-calibration is overdue and we want to re-calibrate it. Can anyone guide me how to re-calibrate (in-house) or help me in developing the re-calibration procedure.
We have some other equipment as digital fiield strength indicator which is calibrated, may be it will help us.

Thanks & Regards


 
 Reply 
 
Nigel
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
06:59 May-27-2006
Re: Re-calibration of EMCOL 6K Portable Current Flow Magnetic Particle Inspection Unit
Jamil

Calibration of the Emcol unit will primarily be to verify the accuacy of the mounted ammeters, so you will need someone who can check their accuracy back to National Standards. For the magnetic flux function which ia a %age, no calibration is feasible, just a function test. Best bet is to find an approved calibration authority in your area.

Hope this helps. Nigel

----------- Start Original Message -----------
:
: We are using the Current Flow Magnetic Particle Inspection Unit. Its re-calibration is overdue and we want to re-calibrate it. Can anyone guide me how to re-calibrate (in-house) or help me in developing the re-calibration procedure.
: We have some other equipment as digital fiield strength indicator which is calibrated, may be it will help us.
: Thanks & Regards
------------ End Original Message ------------




 
 Reply 
 
mandar vinze
Engineering, manufacturer
magnafield controls, India, Joined May 2006, 3

mandar vinze

Engineering, manufacturer
magnafield controls,
India,
Joined May 2006
3
02:51 Jun-07-2006
Re: Re-calibration of EMCOL 6K Portable Current Flow Magnetic Particle Inspection Unit
we are manufacturers of M T equipment and would be glad to help you in any matter relating to equipment,accesories ,calliberation service etc.We are in India and are in this for more than 30 yrs----------- Start Original Message -----------
: Jamil
: Calibration of the Emcol unit will primarily be to verify the accuacy of the mounted ammeters, so you will need someone who can check their accuracy back to National Standards. For the magnetic flux function which ia a %age, no calibration is feasible, just a function test. Best bet is to find an approved calibration authority in your area.
: Hope this helps. Nigel
: :
: : We are using the Current Flow Magnetic Particle Inspection Unit. Its re-calibration is overdue and we want to re-calibrate it. Can anyone guide me how to re-calibrate (in-house) or help me in developing the re-calibration procedure.
: : We have some other equipment as digital fiield strength indicator which is calibrated, may be it will help us.
: : Thanks & Regards
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

AIS229 - Multipurpose Real Time System

Latest standard & automatic real time system developed by Balteau. The AIS229 has been designed to
...
do series inspection in a wide variety of industry. Composed of a shielded cabinet, 5 axis manipulator, x-ray generator and tubehead from 160kV to 225kV, a fl at panel & much more, the AIS229 is most certainly one of the most multipurpose RTR system available on the market.
>

Navic - Steerable Modular Automated Scanner

The Navic is a modular, motorized, steerable scanner designed to carry multiple attachments used
...
in various scanning and inspection applications. The Navic is capable of weld scanning (girth welds and long seam welds), automated corrosion mapping, and tank scanning.
>

EKOSCAN Phased Array

In order to always fit your needs, EKOSCAN can manufacture any type of UT transducer, either convent
...
ional or Phased Array. As an ISO 9001: 2015 certified company, EKOSCAN is extremely careful as far a material selection and manufacturing processes are concerned. Our probes guarantee our customers the benefits of latest innovations regarding piezo-composite, backing, impedance adaptation layer, etc. Specific probes for hostile environment: high temperature, high pressure, corrosive environment,etc. Specific probes designed to fit your specific application: optimization of every parameter to guarantee you the best detection.
>

CIVA 2020 UT Module

CIVA NDE Simulation Software is the world leader of NDT Simulation. The UT simulation Module incl
...
udes: - "Beam computation": Beam propagation simulation - "Inspection Simulation": Beam interaction with flaws or specimens The user can simulate a whole inspection process (pulse echo, tandem or TOFD) with a wide range of probes (conventional, Phased- arrays or EMAT), components, and flaws.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window