A problem with Magnetic Testing

Hi,
On welding of tube, 5ox5mm with two differents materials (SA213T91 and SA213T22)we have a linear indication (longitudinal),from Magnetic Testing.
Can you give us an explanation, about the origin of this problem?
Best regards, waiting for solution to this problem.


 
 Reply 
 
Michael Trinidad
Consultant,
LMATS Pty Ltd , Australia, Joined Jan 2003, 138

Michael Trinidad

Consultant,
LMATS Pty Ltd ,
Australia,
Joined Jan 2003
138
09:41 Jul-09-1999
Re: A problem with Magnetic Testing
Hi,
On welding of tube, 5ox5mm with two different materials (SA213T91 and SA213T22)we have a linear indication (longitudinal),from Magnetic Testing.
Can you give us an explanation, about the origin of this problem?
Best regards, waiting for solution to this problem.


This can be quite a common problem if there is a significant permeability difference between the two steels. What you are most likely seeing is the flux leakage occurring at the permeability change

Michael Trinidad
ETRS NDT Project Supervisor
KBJV Worsley Expansion Project
Direct Tel: 61-8-9781-6232
Fax: 61-8-9781-6233
Email: etrs@kbjv.com




 
 Reply 
 
David Hermanutz
Consultant,
Hbndt.com, China, Joined Jul 2012, 85

David Hermanutz

Consultant,
Hbndt.com,
China,
Joined Jul 2012
85
04:13 Jul-11-1999
Re: A problem with Magnetic Testing
: Hi,
: On welding of tube, 5ox5mm with two different materials (SA213T91 and SA213T22)we have a linear indication (longitudinal),from Magnetic Testing.
: Can you give us an explanation, about the origin of this problem?
: Best regards, waiting for solution to this problem.

:
: This can be quite a common problem if there is a significant permeability difference between the two steels. What you are most likely seeing is the flux leakage occurring at the permeability change

: Michael Trinidad
: ETRS NDT Project Supervisor
: KBJV Worsley Expansion Project
: Direct Tel: 61-8-9781-6232
: Fax: 61-8-9781-6233
: Email: etrs@kbjv.com

It is best to follow up such indications with a dye penetrant procedure to determine the absence of a weld defect.
Or switch to a non magnetic surface test procedure.
In one instance I came across a faulty weld procedure after doing a PMI out of curiosity of such an indication.
148 cut outs and 26 pups later the problem was corrected... (six day overrun on a shutdown due to one typo)

David Hermanutz
Pilbara NDT
Melbourne, Australia





 
 Reply 
 

Product Spotlight

SONOAIR - air-coupled Phased Array Ultrasonic Inspection System

For highly attenuating materials, the performance of the system is critical. The ultrasonic sensors,
...
the scanning area and the system settings should be flexibly adapted to the test task and the material. These high expectations are met with the new and modular testing system SONOAIR. With the world’s first air-coupled phased-array UT inspection system SONOAIR we developed a technology that works with up to 4 transmitter and receiver channels with freely configurable square wave burst transmitters as wells as low noise receiving amplifiers.
>

NEOS III

NEOS III is Logos Imagings lightest DR system. With a built-in battery and internal wireless commu
...
nication, the NEOS III is perfect for users that want to quickly assess an item.
>

GUL QSR1® Scanning

How do you measure pipe wall thickness without direct access to the area? QSR® Scanning - Guide Wav
...
e Quantitative Short Range Scanning.
>

Magnetic X-Ray Pipeline Crawler

Zhong Yi brand pipeline crawler is magnetic/micro-video controlled crawler with DC X ray unit inst
...
alled for checking welding of pipeline. Move steadily inside the pipeline 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometers and provide the efficient inspection of the pipeline.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window