CALIBRATION OF SHEAR WAVES ON A USM 25

PLEASE HELP TO TO LEARN HOW TO CALIBRATE THE USM 25


 
 Reply 
 
David
David
02:13 Jul-14-2006
Re: CALIBRATION OF SHEAR WAVES ON A USM 25
----------- Start Original Message -----------
: PLEASE HELP TO TO LEARN HOW TO CALIBRATE THE USM 25
------------ End Original Message ------------

Try reading the manual, they usually have a section on basic calibration.
Calibration specifics depend on the type of probe and blocks you have and the intended application.
Best way is to attend training course as calibration is only the start, if you intend doing some useful work you need to get some guidance.
Thats why certification schemes call for record of training and experience under guidance.
Like most things, its not really as simple as experienced technicians make it look.


 
 Reply 
 
Roberto Falconio
Engineering, NDT/Metrology SUPERVISOR
El Salvador, Joined Feb 2000, 13

Roberto Falconio

Engineering, NDT/Metrology SUPERVISOR
El Salvador,
Joined Feb 2000
13
03:01 Jul-14-2006
Re: CALIBRATION OF SHEAR WAVES ON A USM 25
----------- Start Original Message -----------
: PLEASE HELP TO TO LEARN HOW TO CALIBRATE THE USM 25
------------ End Original Message ------------

Hi Cameron, USM 25 is a very nice piece of equipment, go to CAL functions, If you are using staight beam, introduce Reference echo 1 and echo 2 as the distance matching 2 consecutives back wall reflection on you standard block, you may use the first back wall and the second back wall reflections, lets do it simple, imagine you have a 10 mm thick cal block if it is steel go to the base menu and introduce a velocity similar to the steel normally this equipment will handle 5920 m/s, then go to CAL and with the transduer on the block identify both fist and second back wall reflections, introduce REF 1 as 10 mm and REF 2 as 20 mm position he gate on echo 1 (first back wall reflection) then press copy and a message echo 1 is recorded will be displyed and then on do the same on echo 2 (second back wall reflection), the equipmente will automatically calculate the velocity and the P delay.
you can do the same with an angle beam block but keep in mind that velocity should be around the half of 5920 m/s and echoes to select should be 25 mm and 100 mm with a range of 250 mm. once the equipment is calibrated you may change the range without disturbing the calibratio
if you have more question write me
good luck and best regards!
Roberto Falconio
ASNT NDT level III Reg 105293
AEROMAN TACA


 
 Reply 
 

Product Spotlight

FAAST-PA! OEM Patented phased Array for high speed UT inspection

Multiangle, Multifocus, Multifrequency, Multibeam. Instead of stacking UT electronics and having m
...
any PA probes, FAAST-PA is able to transmit all delay laws within ONE single shot in Real time.
>

AIS229 - Multipurpose Real Time System

Latest standard & automatic real time system developed by Balteau. The AIS229 has been designed to
...
do series inspection in a wide variety of industry. Composed of a shielded cabinet, 5 axis manipulator, x-ray generator and tubehead from 160kV to 225kV, a fl at panel & much more, the AIS229 is most certainly one of the most multipurpose RTR system available on the market.
>

Ultrasonic Probe Recharacterization Service

NDT Systems offers a comprehensive Aftercare and Recharacterization Service for all our ultrasonic
...
probes. The Recharacterization Service is fully compliant with International ASTM E1065 Standard Guide (and other applicable standards) and offers complete documentation, traceable to the ASTM E1065 Standard. For more details or to schedule Recharacterization Services contact ndtsales@ndtsystems.com
>

MANTIS

MANTIS is a compact, powerful and affordable flaw detector (16:64PR), dedicated to field operators
...
. This portable equipment offers standard (electronic-, sector-, compound scan, conventional UT for pulse-echo, dual array and TOFD inspections), and advanced phased-array (real-time TFM -Total Focusing Method). MANTIS benefits from the same intuitive and user-friendly interface (CAPTURE) than Gekko for easier inspections.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window