00:20 Aug-09-1999
TCG

Dear Sir,
I am very interesting to see your nice homepage.
Please help me Sir, I want to know information about the Time Corrected Gain Ultrasonic (TCG-theoritical)).

Thanks a lot for this matter.

Best Regards,
Mudi Haryanto
NDT Inspector


 
 Reply 
 
Rolf Diederichs
Director,
NDT.net, Germany, Joined Nov 1998, 609

Rolf Diederichs

Director,
NDT.net,
Germany,
Joined Nov 1998
609
02:29 Aug-09-1999
Re: TCG
Defination of "TCG"

Time-dependent gain (TDG)

Time dependent or swept gain function fitted to some ultrasonic instruments to
correct for the distance related reduction in reflected amplitude.
Editor comment: Usually called DAC 'distance amplitude correction' or TCG 'time controlled gain'

This definitions is used in the standard EN 12688-1 "Characterization and verification of ultrasonic examination -
Part 1: Instruments" For other definations the EN 1330 shall apply.
http://www.ndt.net/article/ut_az/ut_u/info/def_eq.htm

----------------
: Dear Sir,
: I am very interesting to see your nice homepage.
: Please help me Sir, I want to know information about the Time Corrected Gain Ultrasonic (TCG-theoritical)).

: Thanks a lot for this matter.

: Best Regards,
: Mudi Haryanto
: NDT Inspector




 
 Reply 
 
mudi haryanto
mudi haryanto
03:56 Aug-10-1999
Re: TCG
: Dear Sir,
Thanks for response soon,
I have an USD 10 (krautkramer), and for reject of assumption I used TCG. My question is how to applied and procedure of TCG ?
I can operate TCG corresponding to manual USD 10 but I don't know to apply for reject.
is the TCG include ASME, DGS, or AWS standard ?
I want to change information, may be everbody do you have experience to use USD 10 ?


Thanks

Mudi haryanto


 
 Reply 
 
E. Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1286

E. Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1286
03:48 Aug-10-1999
Re: TCG
Mudi:
ASME Section V Article 4 Appendix B Gives instructions on how to construct a Distance Amplitude Correction curve to set your primary sensitivity.
These instructions are for the "traditional" curve drawing marked directly on the screen. However, the CURVE is one option for the amplitude correction that is required with increasing soundpath.
When you go to the main body of ASME Section V Article 5 T-542.4.3 (as well as Article 4 Appendix B) it states that other techniques for calibrating angle beams may be used.
Since TCG (Time Corrected Gain) is merely a convenient method of increasing gain as distance along the timebase increases, this effectively allows you to provide the same sort of compensation to sensitivity as is provided with the DAC curve.
The advantage of the TCG is that for greater distances the signals to evaluate are a smaller percentage of the total screen height when using a DAC curve.
But when TCG is used the signals evaluated for 20% of the Reference sensitivity are the same screen height at 90% of the screen range (soundpath) as they are at 20% of the screen range.

I do not know the requirements for AWS but the wording in ASME obviously permits TCG.

Ed

: : Dear Sir,
: Thanks for response soon,
: I have an USD 10 (krautkramer), and for reject of assumption I used TCG. My question is how to applied and procedure of TCG ?
: I can operate TCG corresponding to manual USD 10 but I don't know to apply for reject.
: is the TCG include ASME, DGS, or AWS standard ?
: I want to change information, may be everbody do you have experience to use USD 10 ?

:
: Thanks

: Mudi haryanto




 
 Reply 
 
joseph
joseph
00:44 Mar-01-2000
Re: TCG
: Dear Sir,
: I am very interesting to see your nice homepage.
: Please help me Sir, I want to know information about the Time Corrected Gain Ultrasonic (TCG-theoritical)).

: Thanks a lot for this matter.

: Best Regards,
: Mudi Haryanto
: NDT Inspector




 
 Reply 
 

Product Spotlight

Teletest Focus+

Teletest Focus+ electronics have superior capabilities than rivals on the market. Beyond the usual
...
test features, Focus+ has 24 transmit channels and 24 receive channels, with an additional on-board switching capabilities. The instrument's frequency range is 10–300 kHz.
>

iProbe - USB Phased Array Probe

Turn your PC, Laptop or Tablet into a powerful 32:64 Phased Array system with our USB powered Phas
...
ed Array iProbe and inspection software package. Perform affordable weld inspection and corrosion mapping inspections with this powerful but small system. The probe has a 100V pulser, 8K PRF, dual axis encoder input and achieves outstanding signal quality.
>

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Combination of Digital Image Correlation and Thermographic Measurements

The combination of measuring results from the digital image correlation (ARAMIS, DIC) and temperat
...
ure measuring data from infrared cameras permits the simultaneous analysis of the thermal and mechanical behavior of test specimens in the materials and components testing field.
>