UT after MT. Surface preparation

Does it enough to perform UT after MT with manual cleaning on surface after MT? What i mean, UT still can be performed eventhough the weld surface is not cleaned by power brush.


 
 Reply 
 
chris s
chris s
08:00 Jul-06-2007
Re: UT after MT. Surface preparation
----------- Start Original Message -----------
: Does it enough to perform UT after MT with manual cleaning on surface after MT? What i mean, UT still can be performed eventhough the weld surface is not cleaned by power brush.
------------ End Original Message ------------

that should be fine. you will have problems if you ut and then mag. the ut gel will need to be completely removed.


 
 Reply 
 
Nigel Armstrong
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel Armstrong

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
06:33 Jul-11-2007
Re: UT after MT. Surface preparation
It depends on the overall condition of the scan surface(s). If such surfaces were very clean, smooth and free of spatter and other surface irregularities, then the MPI technician will only have needed to apply a very fine and even thickness contrast coating prior to MPI. This would make UT on the painted surface possible if the contast coat is tightly adhering and very thin. However if the surface condition was poor at the time of MPI, then you can expect a much thicker, less even thickness coating - MOPI techs tend to spray contrast coating until it looks a good smooth finish. My general advice would be to power brush the scanning surfaces prior to MPI because this allows the technician(s) more control over both the MPI (thinner more even coating) and the UT ( better probe coupling and manipulation) leading to a higher probability of detection of defects.


----------- Start Original Message -----------
: : Does it enough to perform UT after MT with manual cleaning on surface after MT? What i mean, UT still can be performed eventhough the weld surface is not cleaned by power brush.
: that should be fine. you will have problems if you ut and then mag. the ut gel will need to be completely removed.
------------ End Original Message ------------




 
 Reply 
 
chris s
chris s
00:18 Jul-12-2007
Re: UT after MT. Surface preparation
----------- Start Original Message -----------
: It depends on the overall condition of the scan surface(s). If such surfaces were very clean, smooth and free of spatter and other surface irregularities, then the MPI technician will only have needed to apply a very fine and even thickness contrast coating prior to MPI. This would make UT on the painted surface possible if the contast coat is tightly adhering and very thin. However if the surface condition was poor at the time of MPI, then you can expect a much thicker, less even thickness coating - MOPI techs tend to spray contrast coating until it looks a good smooth finish. My general advice would be to power brush the scanning surfaces prior to MPI because this allows the technician(s) more control over both the MPI (thinner more even coating) and the UT ( better probe coupling and manipulation) leading to a higher probability of detection of defects.
:
: : : Does it enough to perform UT after MT with manual cleaning on surface after MT? What imean, UT still can be performed eventhough the weld surface is not cleaned by power brush.
: : that should be fine. you will have problems if you ut and then mag. the ut gel will need to be completely removed.
------------ End Original Message ------------

i did not see any mention of a paint coating



 
 Reply 
 
Nigel Armstrong
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel Armstrong

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
05:33 Jul-13-2007
Re: UT after MT. Surface preparation
chris

If a fluorescent technique not requiring a contrast paint coating is required, then surface condition is even more critical. In nearly all other cases a contrast paint coat is required.

----------- Start Original Message -----------
: : It depends on the overall condition of the scan surface(s). If such surfaces were very clean, smooth and free of spatter and other surface irregularities, then the MPI technician will only have needed to apply a very fine and even thickness contrast coating prior to MPI. This would make UT on the painted surface possible if the contast coat is tightly adhering and very thin. However if the surface condition was poor at the time of MPI, then you can expect a much thicker, less even thickness coating - MOPI techs tend to spray contrast coating until it looks a good smooth finish. My general advice would be to power brush the scanning surfaces prior to MPI because this allows the technician(s) more control over both the MPI (thinner more even coating) and the UT ( better probe coupling and manipulation) leading to a higher probability of detection of defects.
: :
: : : : Does it enough to perform UT after MT with manual cleaning on surface after MT? What i mean, UT still can be performed eventhough the weld surface is not cleaned by power brush.
: : : that should be fine. you will have problems if you ut and then mag. the ut gel will need to be completely removed.
: i did not see any mention of a paint coating
------------ End Original Message ------------




 
 Reply 
 
chris s
chris s
08:09 Jul-17-2007
Re: UT after MT. Surface preparation
----------- Start Original Message -----------
are you talking about a developer?

: chris
: If a fluorescent technique not requiring a contrast paint coating is required, then surface condition is even more critical. In nearly all other cases a contrast paint coat is required.
: : : It depends on the overall condition of the scan surface(s). If such surfaces were very clean, smooth and free of spatter and other surface irregularities, then the MPI technician will only have needed to apply a very fine and even thickness contrast coating prior to MPI. This would make UT on the painted surface possible if the contast coat is tightly adhering and very thin. However if the surface condition was poor at the time of MPI, then you can expect a much thicker, less even thickness coating - MOPI techs tend to spray contrast coating until it looks a good smooth finish. My general advice would be to power brush the scanning surfaces prior to MPI because this allows the technician(s) more control over both the MPI (thinner more even coating) and the UT ( better probe coupling and manipulation) leading to a higher probability of detection of defects.
: : :
: : : : : Does it enough to perform UT after MT with manual cleaning on surface after MT? What i mean, UT still can be performed eventhough the weld surface is not cleaned by power brush.
: : : : that should be fine. you will have problems if you ut and then mag. the ut gel will need to be completely removed.
: : i did not see any mention of a paint coating
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

ISONIC 3510

Very Powerful Superior Performance Extremely Portable Smart Phased Array Ultrasonic Flaw Detector an
...
d Recorder with 2 Conventional UT and TOFD Channels uniquely combines PA, single- and multi-channel conventional UT, and TOFD modalities providing 100% raw data recording and imaging. Suitable for all kinds of every-day ultrasonic inspections
>

NEW - TD Focus-ScanRX

The NEW Next Generation Advanced UT platform, TD Focus ScanRX - Also available as a card stack solut
...
ion. Key Improvements 1. Data acquisition is significantly faster than current design 2. Better aesthetic – closely aligns with HandyScan RX 3. Improved IP rating (Target IP66) 4. Ruggedized housing 5. Connectors are protected from impact and ingress 6. Integrated stand and separate retractable handle easy to keep clean) 7. Touchscreen with ruggedized display glass 8. 3-Axis encoder input
>

HARDNESS TESTER TKM-459CE combi

TKM-459CE combi applies 2 methods of hardness control: UCI and Leeb. It provides high-accuracy tes
...
ting of metals and alloys as well as items of different sizes and configurations, their hardened layers and galvanic coatings. Device represents results in HB, HRC, HV and others. Shock-, dust- and water-proof housing with intuitive software make this gauge easy to use in all working conditions.
>

MIZ®-21C: Truly Affordable Eddy Current Handheld with Surface Array Capability

Introducing MIZ-21C, the first truly affordable handheld eddy current instrument with surface array
...
capabilities. MIZ-21C can deliver fast, accurate inspections in demanding NDT applications including aerospace, oil and gas, manufacturing, and power generation. The surface array solution can reduce inspection time by up to 95% compared to traditional handheld pencil probes. The ergonomic design, long battery life, and intuitive touchscreen mean you can inspect more areas faster than ever without fatigue.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window