Root Pass Evaluation S.S. Pipes

Hi everybody,

Is there any NDT method (or idea) for evaluation of the root pass of austenitic stainless steel pipes which are suspected to use of the wrong electrode (i.e. E6010)?
The thickness range is from 5 to 20 mm. with variable diameter.

Regards,


B. Hosseini



    
 
 
Ryan Burns
Ryan Burns
03:30 Jul-10-2007
Re: Root Pass Evaluation S.S. Pipes
If weld prep angle allows, use of a XRF (Positive Material Analysis ie Niton) unit should work. For thickness closer to the 5mm mark, a smaller than standard head would likely have to be used. If you truly suspect E6010 was used (or another ferromagnetic rod) why not just use a small magnet?

Regards

Ryan.


----------- Start Original Message -----------
: Hi everybody,
: Is there any NDT method (or idea) for evaluation of the root pass of austenitic stainless steel pipes which are suspected to use of the wrong electrode (i.e. E6010)?
: The thickness range is from 5 to 20 mm. with variable diameter.
: Regards,
:
: B. Hosseini
------------ End Original Message ------------




    
 
 
Ryan Burns
Ryan Burns
05:14 Jul-11-2007
Re: Root Pass Evaluation S.S. Pipes
My preceding remarks are assuming just root pass is welded. If fill and cap are complete as well, then I am not sure. I found someone else with a similar problem, the thread can be viewed here.

Ryan.

----------- Start Original Message -----------
: If weld prep angle allows, use of a XRF (Positive Material Analysis ie Niton) unit should work. For thickness closer to the 5mm mark, a smaller than standard head would likely have to be used. If you truly suspect E6010 was used (or another ferromagnetic rod) why not just use a small magnet?
: Regards
: Ryan.
:
: : Hi everybody,
: : Is there any NDT method (or idea) for evaluation of the root pass of austenitic stainless steel pipes which are suspected to use of the wrong electrode (i.e. E6010)?
: : The thickness range is from 5 to 20 mm. with variable diameter.
: : Regards,
: :
: : B. Hosseini
------------ End Original Message ------------




    
 
 
Ryan Burns
Ryan Burns
05:17 Jul-11-2007
Re: Root Pass Evaluation S.S. Pipes
http://www.aws.org/cgi-bin/mwf/topic_show.pl?pid=65571;hl=


----------- Start Original Message -----------
: My preceding remarks are assuming just root pass is welded. If fill and cap are complete as well, then I am not sure. I found someone else with a similar problem, the thread can be viewed here.
: Ryan.
: : If weld prep angle allows, use of a XRF (Positive Material Analysis ie Niton) unit should work. For thickness closer to the 5mm mark, a smaller than standard head would likely have to be used. If you truly suspect E6010 was used (or another ferromagnetic rod) why not just use a small magnet?
: : Regards
: : Ryan.
: :
: : : Hi everybody,
: : : Is there any NDT method (or idea) for evaluation of the root pass of austenitic stainless steel pipes which are suspected to use of the wrong electrode (i.e. E6010)?
: : : The thickness range is from 5 to 20 mm. with variable diameter.
: : : Regards,
: : :
: : : B. Hosseini
------------ End Original Message ------------




    
 
 
Godfrey Hands
Consultant,
PRI Nadcap, United Kingdom, Joined Nov 1998, 293

Godfrey Hands

Consultant,
PRI Nadcap,
United Kingdom,
Joined Nov 1998
293
08:43 Jul-11-2007
Re: Root Pass Evaluation S.S. Pipes
----------- Start Original Message -----------
: Hi everybody,
: Is there any NDT method (or idea) for evaluation of the root pass of austenitic stainless steel pipes which are suspected to use of the wrong electrode (i.e. E6010)?
: The thickness range is from 5 to 20 mm. with variable diameter.
: Regards,
:
: B. Hosseini
------------ End Original Message ------------
Hi Barham (and Ryan).
I have come accross this problem a couple of times in the past, and we have been able to get a satisfactory solution.

Let us assume that the problem is caused mostly on closer welds and repairs where it is difficult for the welder to purde the line before welding.
If you can agree with this, then you only have these welds to inspect, otherwise its probably a 100% inspection regime.

We used (electro)magnetic methods as follows:
For thicknesses below 5.5 mm we used a hand held small magnet. test pieces needed to calibrate against.

For thicknesses between 5mm and about 11mm we used a permeability meter (Forster sell one but it is very expensive. Other companies also sell them for much less money). Again calibration pieces needed.

For thicknesses above 11 mm we used a multi-frequency eddy current instrument (in our cases we used IBG Eddyliner with 25mm diameter probe, which tests simultaneously at up to 8 frequencies. Frome memory, frequencies around 150 Hz seemed to work best, but other frequencies also contribute to the result).
Again calibration pieces are needed.

Please note. Depending on the pipe orientation (horizontal or vertical pipe), there will be different amounts of free ferrite in the weld from the heating effect of the welding. This is also true for repairs where extra heat has been introduced, so free ferrite in the weld will be higher.
Also if the weld is TIG root and stick filled, this will be different to TIG root and TIG fill and cap.

Looks like you have a major job on your hands.

Success.

Regards,

Godfrey Hands




    
 
 
B. Hosseini
Director,
IQSecc, Iran, Joined Oct 2002, 4

B. Hosseini

Director,
IQSecc,
Iran,
Joined Oct 2002
4
00:40 Jul-11-2007
Re: Root Pass Evaluation S.S. Pipes
Hi Ryan.
for using the magnet, there is no access to pipes inside, all Tie-in welds have already done.

B. Hosseini
----------- Start Original Message -----------
: If weld prep angle allows, use of a XRF (Positive Material Analysis ie Niton) unit should work. For thickness closer to the 5mm mark, a smaller than standard head would likely have to be used. If you truly suspect E6010 was used (or another ferromagnetic rod) why not just use a small magnet?
: Regards
: Ryan.
:
: : Hi everybody,
: : Is there any NDT method (or idea) for evaluation of the root pass of austenitic stainless steel pipes which are suspected to use of the wrong electrode (i.e. E6010)?
: : The thickness range is from 5 to 20 mm. with variable diameter.
: : Regards,
: :
: : B. Hosseini
------------ End Original Message ------------




    
 
 
Ingo Becker
Engineering, NDT Inspector & Sales
ec-works GmbH, Germany, Joined Apr 2000, 29

Ingo Becker

Engineering, NDT Inspector & Sales
ec-works GmbH,
Germany,
Joined Apr 2000
29
08:57 Jul-12-2007
Re: Root Pass Evaluation S.S. Pipes
It should also work with the use of eddy current. The electrical conductivities of the different electrode materials are different, and the eddy current method requires these conductivity differences. We have done exactly these inspection or search task several times. Please contact me for detailed information.

Best regards

Ingo Becker

----------- Start Original Message -----------
: Hi everybody,
: Is there any NDT method (or idea) for evaluation of the root pass of austenitic stainless steel pipes which are suspected to use of the wrong electrode (i.e. E6010)?
: The thickness range is from 5 to 20 mm. with variable diameter.
: Regards,
:
: B. Hosseini
------------ End Original Message ------------




    
 
 

Product Spotlight

Varex Imaging Large Field of View (FOV) Digital Detector Arrays (DDAs)

A larger FOV DDA can reduce the space and volume of the X-ray inspection system on the factory floor
...
, enable faster scanning times, better throughput and better resolution images at a lower dose. Customers can also save time and money. With these benefits in mind, Varex Imaging has designed a family of large FOV detectors (4343HE, XRD 1611, 4343DX-I, 4343CT) for our industrial imaging customers.
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

FD800 Bench Top Flaw Detectors

The bench-top FD800 flaw detector range combines state-of-the-art flaw detection with advanced mater
...
ial thickness capabilities. Designed for use in the laboratory these gauges are the tool you need for all your flaw detecting needs.
>

FAAST-PA! OEM Patented phased Array for high speed UT inspection

Multiangle, Multifocus, Multifrequency, Multibeam. Instead of stacking UT electronics and having m
...
any PA probes, FAAST-PA is able to transmit all delay laws within ONE single shot in Real time.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window