coverage for TOFD

I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.


    
 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1228

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1228
03:11 Jul-25-2007
Re: coverage for TOFD
----------- Start Original Message -----------
: I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.
------------ End Original Message ------------
The "missed data" would be the dead zone. This is not actually a function of the wall thickness but instead, the ring time equivalent depth of the pulse. Frequency, probe spacing and angle used can all be factors. A simple calculator is provided as a download on NDT.net to estimate the Dead Zone heights associated with TOFD setups. Go to http://www.ndt.net/article/v10n06/ginzel/ginzel.htm
For it to be 1/3 wall thickness the specimen you are working with is probably very thin.


    
 
 
sentho
sentho
05:54 Jul-29-2007
Re: coverage for TOFD
- The "missed data" would be the dead zone.-his is from your replay. How it happents? We are using one probe as transmitter and the other is a receiver.--------- Start Original Message -----------
: : I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.
: The "missed data" would be the dead zone. This is not actually a function of the wall thickness but instead, the ring time equivalent depth of the pulse. Frequency, probe spacing and angle used can all be factors. A simple calculator is provided as a download on NDT.net to estimate the Dead Zone heights associated with TOFD setups. Go to http://www.ndt.net/article/v10n06/ginzel/ginzel.htm
: For it to be 1/3 wall thickness the specimen you are working with is probably very thin.
------------ End Original Message ------------




    
 
 
Nigel Armstrong
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1094

Nigel Armstrong

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1094
02:31 Jul-29-2007
Re: coverage for TOFD
Sentho

The dead zone which Ed is refering to is not of the same origin as the dead zone with the single crystal pulse-echo technique which is a characteristic of the transducer vibration. The TOFD dead zone is caused by the surface or lateral wave travelling from the transmitting to the receiving probe. Any near-surface diffracted signals will be swamped by the lateral wave.

Regards

Nigel


---------- Start Original Message -----------
: - The "missed data" would be the dead zone.-his is from your replay. How it happents? We are using one probe as transmitter and the other is a receiver.--------- Start Original Message -----------
: : : I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.
: : The "missed data" would be the dead zone. This is not actually a function of the wall thickness but instead, the ring time equivalent depth of the pulse. Frequency, probe spacing andangle used can all be factors. A simple calculator is provided as a download on NDT.net to estimate the Dead Zone heights associated with TOFD setups. Go to http://www.ndt.net/article/v10n06/ginzel/ginzel.htm
: : For it to be 1/3 wall thickness the specimen you are working with is probably very thin.
------------ End Original Message ------------




    
 
 

Product Spotlight

SONOAIR - air-coupled Phased Array Ultrasonic Inspection System

For highly attenuating materials, the performance of the system is critical. The ultrasonic sensors,
...
the scanning area and the system settings should be flexibly adapted to the test task and the material. These high expectations are met with the new and modular testing system SONOAIR. With the world’s first air-coupled phased-array UT inspection system SONOAIR we developed a technology that works with up to 4 transmitter and receiver channels with freely configurable square wave burst transmitters as wells as low noise receiving amplifiers.
>

NEOS III

NEOS III is Logos Imagings lightest DR system. With a built in \r\nbattery and internal wireless c
...
ommunication, the NEOS III is perfect \r\nfor users that want to quickly assess an item.
>

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

MUSE Mobile Ultrasonic Equipment

The MUSE, a portable ultrasonic imaging system, was developed for in-field inspections of light-weig
...
ht structures. The MUSE consists of a motor-driven manipulator, a water circulation system for the acoustic coupling and a portable ultrasonic flaw detector (USPC 3010). The MUSE provides images of internal defects (A-, B-,C- and D-scan).
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window