suresh
suresh
02:54 Aug-15-2007
curved probes in ToFD

what is the criteria for choice of curved probes in TOFD?

suresh


    
 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1268

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1268
03:52 Aug-15-2007
Re: curved probes in ToFD
Suresh:
I am not aware of any of the existing TOFD Codes (EN, ASTM or BS) giving instruction on curvature matching of probes. The best advice I have seen is in the EN Standard for UT on Welded Joints, EN1714. There, in paragraph 6.3.3, instruction is made for Adaption of Probes to curved scanning surfaces. This requires that the gap between the test surface and bottom of the probe shoe be not greater than 0.5mm. They suggest that this is satisfied when the diameter of the component is 15 times the dimension of the probe shoe (in the direction of examination) for a flat probe....otherwise the probe SHALL be adapted to the curved surface. I assume this is for testing from the OD and not the ID.
For testing from the OD and using the 0.5mm gap as a guide I generally recommend that when the probe shoe is curved its radius should be greater than or equal to the test surface curvature (not to exceed the 0.5mm gap on the shoe edges).
Avoid making the probe shoe curvature LESS than the test part. This will result in a gap in the middle of the shoe over the test piece. Even if the gap is only 0.5mm at the middle of the shoe the amplitude loss can be between 6-12dB!!
Regards
Ed

----------- Start Original Message -----------
: what is the criteria for choice of curved probes in TOFD?
: suresh
------------ End Original Message ------------




    
 
 Reply 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 177

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
177
05:26 Aug-15-2007
Re: curved probes in ToFD
----------- Start Original Message -----------
: what is the criteria for choice of curved probes in TOFD?
: suresh
------------ End Original Message ------------

In addition to the reply of Ed Ginzel with respect to the referenced standards:

CEN/TS 14751:2005
Use of time-of-flight diffraction technique (TOFD) for testing of welds

in 7.2.2 asks for the matching of probes according to EN 1714.

EN 1714:1997 (2002)
Ultrasonic testing of welded joints

in 6.3.3 gives the equation referenced by Ed Ginzel.

But this equation is not correct.

EN 1417 is under revision now and will be published again soon.

The correct equation (when the max gap is smaller than 0.5 mm, which can be tolerated for 4 MHz), is

Dobj < (D)squared/ 2 mm,

where Dobj is the outer diameter of the object, and D is the dimension of the transducer parallel to the curvature.

The same correction has to be made to the respective equations in 3.4.1 and 3.4.2 of

EN 583-2:2001
Ultrasonic testing - Part 2: Sensitivity and range setting

(The above equation with squared transducer diameter is correct)

Best regards

Udo Schlengermann
GE Inspection Technologies GmbH
Huerth, Germany

Chairman of CEN/TC121/SC5/WG2 (UT of welds)
Chairman of CEN/TC138/WG2 (UT)




    
 
 Reply 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 177

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
177
05:38 Aug-15-2007
Re: curved probes in ToFD
Correction:

Before someone starts to think about my last comment,
I have to apologize for two mistypings:

- The European Standard of course is EN 1714 (not 1417).


- The correct equation is

Dobj > (D)squared/ 2 mm

( Dobj bigger not smaller than - all dimensions in mm)

Sorry
Udo Schlengermann


----------- Start Original Message -----------
: what is the criteria for choice of curved probes in TOFD?
: suresh
------------ End Original Message ------------




    
 
 Reply 
 

Product Spotlight

AMIGO2

TSC Amigo2 - ACFM technology has developed a solid reputation for accurately detecting and sizing
...
surface-breaking cracks through paint and coatings. As the industry demands increased performance in speed, signal quality, and portability, it’s time for an evolution. It’s time for Amigo2.
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

NEW! The PragmaPro Instrument Platform

The PragmaPro is based on a modular cartridge technology and supports various NDT instrument modal
...
ities such as UT, PAUT, ECT and many more. This new platform is based on a machined, powder-coated aluminum frame for shock-proofness, best sealing qualities and maximum heat dissipation. This is practical to extend the outdoor temperature range and/or to extend the power injected in the transducers. The PragmaPro is aiming at a very wide range of applications, such as weld scanning, corrosion mapping and composite testing.
>

NEW - TD Focus-ScanRX

The NEW Next Generation Advanced UT platform, TD Focus ScanRX - Also available as a card stack solut
...
ion. Key Improvements 1. Data acquisition is significantly faster than current design 2. Better aesthetic – closely aligns with HandyScan RX 3. Improved IP rating (Target IP66) 4. Ruggedized housing 5. Connectors are protected from impact and ingress 6. Integrated stand and separate retractable handle easy to keep clean) 7. Touchscreen with ruggedized display glass 8. 3-Axis encoder input
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window