My company has experienced a recent spate of ultrasonic transducer failures due to blistering of the epoxy facing of cylindrically focused immersion transducers, 3 in the last two months. One transducer was not quite one year old. The other two were approximately 5 years old. Does anyone know of a possible cause of transducer blistering? Perhaps an additive or change in chemical composition of the water? Are there any known chemicals which could attack the epoxy faces of transducers? Could sodium sulfite be a cause?


 
 Reply 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
08:16 Oct-17-2007
Re: Transducer failures
Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.

In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1307

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1307
03:06 Oct-18-2007
Re: Transducer failures
I would like to first concur with Tom and then just add that if the water temperature has been elevated (e.g. 25-35C) it may accelerate the situation

----------- Start Original Message -----------
: Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.
: In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

32:128PR PAUT & 2-ch TOFD Flaw Detector: SyncScan 2

●32:128PR PAUT with optional 2-ch TOFD to maximize your efficiency ●Support PR mode for corrosi
...
on inspection ●Multiple solutions for welds & corrosion ●Faster scan speed (Approximately 3 meters/minute) ●Light weight: 4kg only including battery
>

IRIS 9000Plus - Introducing the next generation of heat exchanger inspection.

Representing the seventh generation of the IRIS system, the IRIS 9000 Plus has nearly 200 years of c
...
ombined field inspection experience incorporated in its design. This experience combined with a strong commitment to quality and a history of innovation has made Iris Inspection Services® the undisputed leader in IRIS technology.
>

Extended Range Variable Wall® Piping Calibration Blocks for Longseam Inspection

This set of two blocks has been specially designed for situations where ultrasonic examination tak
...
es place circumferentially or “around the curve” for applications such as inspection of longseam welds in piping and vessels. Jointly designed by PH Tool and Holloway NDT, the blocks feature side-drilled hole reflectors positioned within novel compound-curves, each with multiple thickness steps. The unique design results in a light and compact form factor, with the largest block weighing just over 15 lbs and spanning just 16 ¾”. Coverage is provided for nominal pipe sizes from 6” to 10” in the small block and 12” to a whopping 56” in the large block. The compound OD curves and multiple wall thicknesses actually exceed ASME requirements, providing exceptional signal response and measurement accuracy. Properly managing beam divergence off the ID and stabilizing beam skew due to probe rocking is essential to maintaining a consistent calibration when inspecting lengthwise on curved surfaces.
>

Cygnus 6+ PRO Multi-Mode Ultrasonic Thickness Gauge

The Cygnus 6+ PRO thickness gauge is the most advance gauge within the Cygnus range with key featu
...
res including: comprehensive data logging; A-scan and B-scan display; manual gain control; Bluetooth connectivity; and much more. With its unique dual display and three measuring modes (Multiple-Echo, Echo-Echo and Single-Echo), this surface thickness gauge offers maximum versatility for inspections.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window