My company has experienced a recent spate of ultrasonic transducer failures due to blistering of the epoxy facing of cylindrically focused immersion transducers, 3 in the last two months. One transducer was not quite one year old. The other two were approximately 5 years old. Does anyone know of a possible cause of transducer blistering? Perhaps an additive or change in chemical composition of the water? Are there any known chemicals which could attack the epoxy faces of transducers? Could sodium sulfite be a cause?


    
 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
08:16 Oct-17-2007
Re: Transducer failures
Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.

In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.


    
 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1247

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1247
03:06 Oct-18-2007
Re: Transducer failures
I would like to first concur with Tom and then just add that if the water temperature has been elevated (e.g. 25-35C) it may accelerate the situation

----------- Start Original Message -----------
: Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.
: In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.
------------ End Original Message ------------




    
 
 

Product Spotlight

Panther

M2M PANTHER is a phased-array equipment designed for both desktop and industrial applications, offer
...
ing unparalleled performance in a compact unit. It combines the speed required for industrial integrated Phased-Array Ultrasound (PAUT) solutions, with the most complete set of total focusing method (TFM) imaging techniques, making it the ultimate tool for R&D and procedure qualification.
>

Navic - Steerable Modular Automated Scanner

The Navic is a modular, motorized, steerable scanner designed to carry multiple attachments used
...
in various scanning and inspection applications. The Navic is capable of weld scanning (girth welds and long seam welds), automated corrosion mapping, and tank scanning.
>

Mentor UT – Ultrasonic Phased Array Flaw Detector

With Mentor UT, you get an accessible and efficient inspection experience. Create user-defined menus
...
and workflows (“apps”) with GE’s desktop software Mentor Create to ensure consistency, even for the most complex inspections, every time. Mentor UT is the first UT device to easily allow wireless connectivity and live streaming, so you get second opinions when you need them most—in real-time.
>

High Resolution Industrial CT System for Small/Medium Size Parts Inspection

The YXLON FF35 CT computed tomography system is designed to achieve extremely precise inspection res
...
ults for a wide range of applications. Available in a single or dual tube configuration, it is perfect for very small to medium size parts inspection in the automotive, electronics, aviation, and material science industries.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window