UV Light Intensity - Labino Lamp

Many MT and Pt standards requires that UV Light intensity be 800 or 1000 micro-Watts / square centimeters minimum at the surface of the part with the light at 15 inches from the part. With the very high intensity of the Labino lamps how do people in the field measure the light intensity without exceedig the working range of their light meter? Do you increase the distance between the light and the meter and find the intensity using the inverse square law formula?


 
 Reply 
 
David
David
02:38 Oct-23-2007
Re: UV Light Intensity - Labino Lamp
Michael,
Why do you need the actual measurement? The spec says minimum, if you keep a log just enter .... above whatever is the highest range of the meter. We have a Labino lamp and had to go to metres away to get in range, makes no sense to enter that distance as you don't inspect at that distance and why bother with the calculation, you just end up with some massive number that is well above spec anyway.
Why complicate it for the guys in the field, they have enough things to worry about?

----------- Start Original Message -----------
: Many MT and Pt standards requires that UV Light intensity be 800 or 1000 micro-Watts / square centimeters minimum at the surface of the part with the light at 15 inches from the part. With the very high intensity of the Labino lamps how do people in the field measure the light intensity without exceedig the working range of their light meter? Do you increase the distance between the light and the meter and find the intensity using the inverse square law formula?
------------ End Original Message ------------




 
 Reply 
 
Michel
NDT Inspector,
consultant, Canada, Joined Sep 2006, 868

Michel

NDT Inspector,
consultant,
Canada,
Joined Sep 2006
868
08:32 Oct-23-2007
Re: UV Light Intensity - Labino Lamp
----------- Start Original Message -----------
: Michael,
: Why do you need the actual measurement? The spec says minimum, if you keep a log just enter .... above whatever is the highest range of the meter. We have a Labino lamp and had to go to metres away to get in range, makes no sense to enter that distance as you don't inspect at that distance and why bother with the calculation, you just end up with some massive number that is well above spec anyway.
: Why complicate it for the guys in the field, they have enough things to worry about?
: : Many MT and Pt standards requires that UV Light intensity be 800 or 1000 micro-Watts / square centimeters minimum at the surface of the part with the light at 15 inches from the part. With the very high intensity of the Labino lamps how do people in the field measure the light intensity without exceedig the working range of their light meter? Do you increase the distance between the light and the meter and find the intensity using the inverse square law formula?
------------ End Original Message ------------

David,

I do agree with you. I was just curious as to what people are doing in this situation. On the other hand, auditors in the aircraft industry very seldom accept "above minimum" as a measurement of intensity. You have to be able to duplicate your results when challenged.


 
 Reply 
 
David
David
03:51 Oct-24-2007
Re: UV Light Intensity - Labino Lamp
----------- Start Original Message -----------
: : Michael,
: : Why do you need the actual measurement? The spec says minimum, if you keep a log just enter .... above whatever is the highest range of the meter. We have a Labino lamp and had to go to metres away to get in range, makes no sense to enter that distance as you don't inspect at that distance and why bother with the calculation, you just end up with some massive number that is well above spec anyway.
: : Why complicate it for the guys in the field, they have enough things to worry about?
: : : Many MT and Pt standards requires that UV Light intensity be 800 or 1000 micro-Watts / square centimeters minimum at the surface of the part with the light at 15 inches from the part. With the very high intensity of the Labino lamps how do people in the field measure the light intensity without exceedig the working range of their light meter? Do you increase the distance between the light and the meter and find the intensity using the inverse square law formula?
: David,
: I do agree with you. I was just curious as to what people are doing in this situation. On the other hand, auditors in the aircraft industry very seldom accept "above minimum" as a measurement of intensity. You have to be able to duplicate your results when challenged.
------------ End Original Message ------------

Michael,
Yes sometimes auditor check lists don't keep up with improvements in the real world. Instead of distance and calculation how about trying a grid to cut out a %age of light? if you calibrate it before hand and record the reduction in energy maybe you can bring the light back into range (15"), take a reading and apply a multiplication factor? That way you could have a reasonably easy daily check and a documented correction factor. Or maybe source a UV filter to place over the meter sensor and apply corrections that way. That way you could record actual numbers and have repeatablity.


 
 Reply 
 
Nigel Armstrong
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel Armstrong

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
05:18 Oct-25-2007
Re: UV Light Intensity - Labino Lamp
Never heard of Labino lamps. If they are so powerful, do they run hot?

Any problems viewing properly-processed parts under such high intansity UV-A provided proper safety precautions are taken - such as not looking at the source, covering bare skin, not touching the lamp?


----------- Start Original Message -----------
: : : Michael,
: : : Why do you need the actual measurement? The spec says minimum, if you keep a log just enter .... above whatever is the highest range of the meter. We have a Labino lamp and had to go to metres away to get in range, makes no sense to enter that distance as you don't inspect at that distance and why bother with the calculation, you just end up with some massive number that is well above spec anyway.
: : : Why complicate it for the guys in the field, they have enough things to worry about?
: : : : Many MT and Pt standards requires that UV Light intensity be 800 or 1000 micro-Watts / square centimeters minimum at the surface of the part with the light at 15 inches from the part. With the very high intensity of the Labino lamps how do people in the field measure the light intensity without exceedig the working range of their light meter? Do you increase the distance between the light and the meter and find the intensity using the inverse square law formula?
: : David,
: : I do agree with you. I was just curious as to what people are doing in this situation. On the other hand, auditors in the aircraft industry very seldom accept "above minimum" as a measurement of intensity. You have to be able to duplicate your results when challenged.
: Michael,
: Yes sometimes auditor check lists don't keep up with improvements in the real world. Instead of distance and calculation how about trying a grid to cut out a %age of light? if you calibrate it before hand and record the reduction in energy maybe you can bring the light back into range (15"), take a reading and apply a multiplication factor? That way you could have a reasonably easy daily check and adocumented correction factor. Or maybe source a UV filter to place over the meter sensor and apply corrections that way. That way you could record actual numbers and have repeatablity.
------------ End Original Message ------------




 
 Reply 
 
Michel
NDT Inspector,
consultant, Canada, Joined Sep 2006, 868

Michel

NDT Inspector,
consultant,
Canada,
Joined Sep 2006
868
05:37 Oct-26-2007
Re: UV Light Intensity - Labino Lamp
----------- Start Original Message -----------
: Never heard of Labino lamps. If they are so powerful, do they run hot?
: Any problems viewing properly-processed parts under such high intansity UV-A provided proper safety precautions are taken - such as not looking at the source, covering bare skin, not touching the lamp?
:
Hi Nigel,

Labino Lamp have been on the market for about 5 years now. They are fairly expensive compare to others, but worth every penny. At about 4 feet, one can have a light intensity of about 2 - 3,000 microwatts. Some standards will allow you to have 10 foot-candle of background light if your black light have such an intensity. The operator doesn't have to stand so close to the work as well. They are bulkier than other lamps, but are also well insulated. I never had the chance to work much with them, but I was impressed with what I saw. One supplier of these lamp in Ontario, Canada is Avofix in Mississauga. Call Francis Foo he'll set you up.


 
 Reply 
 
Christoph Schnitger
Christoph Schnitger
07:16 Oct-26-2007
UV Light Intensity - Labino Lamp
Hello Nigel,

we have some Labino lamps in stock ( also some used one ) ! We sold Labino lamps to the german market now for more than 8 years with a lot of sucess. Unfortunally their is a german risk classifikation so nowerdays you need special add-on glass in front. The quality, handling and working with these lamps are really good, but please have in mind that the bulb lamps is really expensive ( and after 2000 hours ) you have to buy a new one !



 
 Reply 
 

Product Spotlight

CIVA 2020 UT Module

CIVA NDE Simulation Software is the world leader of NDT Simulation. The UT simulation Module incl
...
udes: - "Beam computation": Beam propagation simulation - "Inspection Simulation": Beam interaction with flaws or specimens The user can simulate a whole inspection process (pulse echo, tandem or TOFD) with a wide range of probes (conventional, Phased- arrays or EMAT), components, and flaws.
>

SONOAIR - air-coupled Phased Array Ultrasonic Inspection System

For highly attenuating materials, the performance of the system is critical. The ultrasonic sensors,
...
the scanning area and the system settings should be flexibly adapted to the test task and the material. These high expectations are met with the new and modular testing system SONOAIR. With the world’s first air-coupled phased-array UT inspection system SONOAIR we developed a technology that works with up to 4 transmitter and receiver channels with freely configurable square wave burst transmitters as wells as low noise receiving amplifiers.
>

Ultrasonic Squirter/Gantry System

TecScan’s NDT Ultrasonic Gantry Systems are industrial Squirter scanners designed for non-destru
...
ctive quality testing and raster scanning of large structures and parts. The Gantry Systems are usually composed of an Industrial Mechanical Scanner and a Control Room with System Workstation.
>

Echomac® Small

Available with up to eight channels of electronics to detect and evaluate thickness, flaws and eccen
...
tricity, this UT tester is housed in a convenient, smaller cabinet. This instrument can be used in conjunction with bubbler or immersion tank systems, or with a test bench or in laboratory applications. The Echomac® Small is available in the FD4, FD6 or FD6A versions.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window