07:21 Dec-10-1999
Piezo-Composite UT Elements

I am currently using a piezo-composite element with a positive going square wave pulser. The units have been failing in a slowly, within hours, deteriorating fashion. I have spoke with various xdcr manufacturers and found that the polarization of the element requires a negative going pulse. I have always been able to just pick up a transducer and not worry about pulse polarity. Questions I have: can the elements be selectively polarized upon manufacture? Is there an industry standard stating all pulsers be negative going? Is their an industry standard for the polarization of elements?
Any response will be greatly appreciated.


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1300

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1300
08:49 Dec-10-1999
Re: Piezo-Composite UT Elements
A positive going spike for a pulser is a bit non-standard but it is still
odd that it should cause a depolarisation (especialy so fast).
Is the voltage of the pulse especially high?
Is the duration of the pulse very long?
Are there a lot of cycles in the pulse?
Usually a fairly high temperature would need to be associated with a "depolarising voltage".
Are you running the experiment in a heat experiment?

Ed
: I am currently using a piezo-composite element with a positive going square wave pulser. The units have been failing in a slowly, within hours, deteriorating fashion. I have spoke with various xdcr manufacturers and found that the polarization of the element requires a negative going pulse. I have always been able to just pick up a transducer and not worry about pulse polarity. Questions I have: can the elements be selectively polarized upon manufacture? Is there an industry standard stating all pulsers be negative going? Is their an industry standard for the polarization of elements?
: Any response will be greatly appreciated.




 
 Reply 
 
Paul Meyer
R & D,
GE Inspection Technologies, USA, Joined Nov 1998, 47

Paul Meyer

R & D,
GE Inspection Technologies,
USA,
Joined Nov 1998
47
09:16 Dec-10-1999
Re: Piezo-Composite UT Elements
Hi Tim,
The factors causing dpolarization of a piezoelectric material are electric field, temperature, and time. There are many different piezoelectric materials used in transducer construction and some are more capable of tolerating reverse fields than others. As Ed points out high reverse fields at high temperatures are especially damaging. It's also possible that high rep rates can increase the internal temperature of the probe even though the environment isn't hot. Higher frequency probes use thinner piezoelectric elements which increase the field for the same pulse voltage.
If you can't reverse the pulsed field, you can probably have your probe supplier reverse the piezoelectric element to avoid this effect.
Regards,
Paul

: I am currently using a piezo-composite element with a positive going square wave pulser. The units have been failing in a slowly, within hours, deteriorating fashion. I have spoke with various xdcr manufacturers and found that the polarization of the element requiresa negative going pulse. I have always been able to just pick up a transducer and not worry about pulse polarity. Questions I have: can the elements be selectively polarized upon manufacture? Is there an industry standard stating all pulsers be negative going? Is their an industry standard for the polarization of elements?
: Any response will be greatly appreciated.




 
 Reply 
 

Product Spotlight

Magnetic X-Ray Pipeline Crawler 6''-60''

Suitable to work in 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometer
...
s and provide the efficient inspection of the pipeline.
>

On-Demand Webinar: Advanced Ultrasonic Techniques for HTHA Detection & Assessment

In the Oil and Gas industry, High Temperature Hydrogen Attack (HTHA) can occur in steels operating a
...
t high temperature, usually above 205°C (400°F). Early stages of HTHA are very difficult to detect. Join us for a webinar presented by Guy Maes, Zetec Sales Engineer Director. The webinar will demonstrate how advanced ultrasonic techniques are deployed for efficient and reliable HTHA inspections.
>

Cygnus 6+ PRO Multi-Mode Ultrasonic Thickness Gauge

The Cygnus 6+ PRO thickness gauge is the most advance gauge within the Cygnus range with key featu
...
res including: comprehensive data logging; A-scan and B-scan display; manual gain control; Bluetooth connectivity; and much more. With its unique dual display and three measuring modes (Multiple-Echo, Echo-Echo and Single-Echo), this surface thickness gauge offers maximum versatility for inspections.
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window