UT waveform preference

Was wondering what is the common preference for people out there to view there UT waveforms in an A-Scan; full waveform, positive rectification, negative rectification, or full rectification. On my Krautkramer field unit, it defaults to negative rec. and on lab the instruments its full waveform. Is it just a personal preference for what one is use to looking at, or is there a benefit to one or the other? Thanks


    
 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
01:06 Feb-12-2008
Re: UT waveform preference
Some of it has to do with customary practice, and some with test requirements.

Full wave rectification is most people's choice for common flaw detection applications with portable instruments and narrowband transducers (like most angle beam assemblies), in part because for many years in the pre-digital age that was all that portable flaw detectors offered and it became customary, and in part because it's a simpler presentation. However RF provides a more detailed picture of the wave shape that is usually required for interpretation in TOFD testing or phase-sensitive bond testing, as well as in precision thickness gaging applications with broadband transducers where consistent echo lobe detection is required for maximum accuracy.

Oscilloscopes and PC-based waveform digitizers typically offer RF waveforms because historically a lot of users of that type of equipment were doing precision time measurements and/or more advanced wave analysis.

As with most ultrasonic tests, the rule is to choose the approach that meets your needs!




    
 
 
Joe Buckley
Consultant, ASNT L-III, Honorary Secretary of BINDT
Level X NDT, BINDT, United Kingdom, Joined Oct 1999, 520

Joe Buckley

Consultant, ASNT L-III, Honorary Secretary of BINDT
Level X NDT, BINDT,
United Kingdom,
Joined Oct 1999
520
01:34 Feb-12-2008
Re: UT waveform preference
I would agree with Tom's comments entirely, in addition some people prefer to use one or other of the half wave modes for precision thickness and similar applications

I suspect that this is mostly historic - I've never found an application where it gives better results than RF, but a lot of operators seem to feel more comfortable with the unipolar presentation.

Joe

----------- Start Original Message -----------
: Some of it has to do with customary practice, and some with test requirements.
: Full wave rectification is most people's choice for common flaw detection applications with portable instruments and narrowband transducers (like most angle beam assemblies), in part because for many years in the pre-digital age that was all that portable flaw detectors offered and it became customary, and in part because it's a simpler presentation. However RF provides a more detailed picture of the wave shape that is usually required for interpretation in TOFD testing or phase-sensitive bond testing, as well as in precision thickness gaging applications with broadband transducers where consistent echo lobe detection is required for maximum accuracy.
: Oscilloscopes and PC-based waveform digitizers typically offer RF waveforms because historically a lot of users of that type of equipment were doing precision time measurements and/or more advanced wave analysis.
: As with most ultrasonic tests, the rule is to choose the approach that meets your needs!
------------ End Original Message ------------




    
 
 

Product Spotlight

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

Sci Aps Z-Series Portable Handheld Analysers

The world’s only handheld analyzer that measures carbon content in stainless (yes even L-grades),s
...
teels, and cast irons. Also accepted for low Si analysis for sulfidic corrosion analysis, and is widely used in the power industry for Cr analysis, for flow accelerated corrosion applications.
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

Aerospace Systems - Automated Ultrasonic Inspection

USL are specialists in the design and manufacture of turnkey ultrasonic inspection systems for aer
...
ospace applications. From monolithic composites to complex honeycomb structures. This video shows just a few examples of what is possible, find out more at: www.ultrasonic-sciences.co.uk
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window