Helium-Ammonia mixture for leak test

Helium Leak Test & Ammonia Leak Tests are two different Tests for sensitive leakage detection. We are thinking of using a mixture of both gases to save time. As recommended soaking time in case of Ammonia is 08 hours and that four Helium is 01 hour. So by mixing the two gases we can attempt Helium Leak test during soaking time of Ammonia. If some one has any expiernec or any apprehensions kindly please share.


 
 Reply 
 
alberto
Engineering, Leak Detection Service - Thermography Service -
ETS Sistemi Industriali, Italy, Joined Aug 2002, 17

alberto

Engineering, Leak Detection Service - Thermography Service -
ETS Sistemi Industriali,
Italy,
Joined Aug 2002
17
08:50 May-12-2008
Re: Helium-Ammonia mixture for leak test
Which type of tests you need? what is admitted leak value and pressure? It is a Bombing Test?
In any case, please consider that ammonia can damage helium leak detector, expecially if you utilise it in Vacuum Mode.
Ammonia soak during helium test can change the helium signal.
Please send me further information if you need assistance.

----------- Start Original Message -----------
: Helium Leak Test & Ammonia Leak Tests are two different Tests for sensitive leakage detection. We are thinking of using a mixture of both gases to save time. As recommended soaking time in case of Ammonia is 08 hours and that four Helium is 01 hour. So by mixing the two gases we can attempt Helium Leak test during soaking time of Ammonia. If some one has any expiernec or any apprehensions kindly please share.
------------ End Original Message ------------




 
 Reply 
 
Faran Hussain
Faran Hussain
02:11 May-13-2008
Re: Helium-Ammonia mixture for leak test
Thanks for your reply. We are experiencing a very minute leakage in our Urea Stripper
(carbamate on Tube side & steam on shell side). Here are the details of the two leakage tests we can perform separately.

Ammonia Leakage Test:
Method used: Pressure Method
Tracer Gas Used: Ammonia
Detection Method: Ammonia Sensitive paper
Total time required for the test is about 12 hours.

A mixture of Nitrogen & 15% by volume of Ammonia is used to pressurize the shell side. A soaking time of approx 8 hours is required. Nitrogen is used to increase the pressure & thus increasing the sensitivity.

Helium Leak Test
Method : Sniffer Method
Tracer Gas: Helium
Detection Method: Mass spectrometry
Total time required for the test is about 8-10 hours.

A mixture of Helium & Nitrogen is used to pressurize the shell. The soaking time is 1 hour. Nitrogen is used to increase the pressure & thus increasing the sensitivity.

Now to save time, why not use a mixture of helium & ammonia to pressurize theshell side. This way after 1 hour of Helium soaking we would be able to start performing Helium Leak Test and by the time we would complete our helium leak test hopefully soaking time of ammonia (i.e 8 hours) would be over and we would apply ammonia sensitive paper on both tube sheets.

As helium test is not recommended in moist environments. The molecule of Ammonia & that of water are similar in terms of hydrogen bonding. So one of our apprehension is that would ammonia gas would behave like moisture and adversely effect the sensitivity of helium leak test.

More details can also be provided. we would definetly love to have your assistance. Helium Leak tester being used by us is MKS sniffer method using mass spectropetry.


 
 Reply 
 
sridhar jena
sridhar jena
04:43 Oct-01-2008
Leak test using helium and nitrogen
dear Sir
for leak test in a metal which medium i.e. helium or nitrogen is better?
sridhar jena
sr. manager. ntpc


 
 Reply 
 
S.V.Swamy
Engineering, - Material Testing Inspection & Quality Control
Retired from Nuclear Fuel Complex , India, Joined Feb 2001, 787

S.V.Swamy

Engineering, - Material Testing Inspection & Quality Control
Retired from Nuclear Fuel Complex ,
India,
Joined Feb 2001
787
09:30 Oct-02-2008
Re: Leak test using helium and nitrogen
Dear Sridhar,

From considerations of sensitivity, Helium is the first choice, because of two reasons:

Firstly, Helium is much lighter than nitrogen and hence diffusion rates through the leak are faster.

Secondly, the helium content in the atmosphere is very less and so the background effect is much less.

Nitrogen has been used in the leak testing in some special cases. If your needs of sensitivity are met with nitrogen based leak testing equipment, the economies could be an important factor in your choice of the medium.


----------- Start Original Message -----------
: dear Sir
: for leak test in a metal which medium i.e. helium or nitrogen is better?
: sridhar jena
: sr. manager. ntpc
------------ End Original Message ------------




 
 Reply 
 
Pavankumar Jha
Pavankumar Jha
07:09 Feb-21-2014
Re: Leak test using helium and nitrogen
In Reply to S.V.Swamy at 09:30 Oct-02-2008 .

Pls suggest: (helium Leak test)
What is the time of soaking in pressurizing method (Shell side 10% He +Nitrogen) Pressure 1kg/Cm2.
Volume of shell side is Approx- 2.341E+08
cm3 & Tube side will be enveloped in sex sector (Volume of tube side: 2.806E+08 Cm3)
How to calculate?

 
 Reply 
 

Product Spotlight

Ultrasonic Probe Recharacterization Service

NDT Systems offers a comprehensive Aftercare and Recharacterization Service for all our ultrasonic
...
probes. The Recharacterization Service is fully compliant with International ASTM E1065 Standard Guide (and other applicable standards) and offers complete documentation, traceable to the ASTM E1065 Standard. For more details or to schedule Recharacterization Services contact ndtsales@ndtsystems.com
>

NDT Master Lecturer

In the program both university professors and practitioners will give lectures, which guarantees the
...
oretical depth and practical inside. Academic Director: Prof. Dr. Christian Boller
The following lecturers are not complete: Prof. Tadeusz Stepinski, Prof. Wieslaw Staszewski, Prof. Frank Walther, Prof. Giovanni Bruno, Prof. Gerd Dobmann, Prof. Philippe Guy
>

Scatter Correction for Industrial Cone-Beam Computed Tomography (CBCT)

Download our white paper on Scatter Correction for Industrial Cone-Beam Computed Tomography (CBCT)
...
Using 3D VSHARP, a fast GPU-Based Linear Boltzmann Transport Equation Solver here: http://bit.ly/2Y0t78P
>

Echomac® PA Bar Tester

The Echomac® PA BT provides 100% inspection of round bars for core and surface defects, at speeds u
...
p to 2 m/sec, depending on the specific application. The unit allows producers to handle a much larger range of diameters in one installation, and is available in 3 models to test 10 – 75mm diameter (small model); 20 – 130mm (medium model) or 50 – 254mm (large model) bars.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window