Porosity on TOFD inspection

Hello everybody,
Please I need a help concerning the Porosity on TOFD inspection. I couldn't find a specific criteria for porosity cluster. Can anybody help me?
Thanx!
Paulo


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1303

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1303
03:38 Jun-02-2008
Re: Porosity on TOFD inspection
Paulo:
Acceptance criteria will depend on the Code you work to. Unfortunately too many codes are still trying to make ultrasonic results look like radiography. If you can, with great certainty, declare that a TOFD indication obtained from a single line scan (D-scan) is a cluster of porosity, then your assessment skills and confidence is greater than mine.
An upper "tip" signal will be possible with a cluster of pores (it is likely to look like an irregular set of arrival times). The lower boundary of the indication from such a flaw will be VERY difficult to determine. Inter-pore scatter will generate several paths that will be unresolvable from the diffraction off the lowest portion of the flaw.

It is permissable in some Codes to assess pores as if they were planar flaws then a length needs to be determined.

Ed

----------- Start Original Message -----------
: Hello everybody,
: Please I need a help concerning the Porosity on TOFD inspection. I couldn't find a specific criteria for porosity cluster. Can anybody help me?
: Thanx!
: Paulo
------------ End Original Message ------------




 
 Reply 
 
Paulo
Paulo
07:44 Jun-03-2008
Re: Porosity on TOFD inspection
Ed,
First of all, thanks for your help.
So, what you are saying is that is quite impossible to assess porosity using only TOFD method? Cause it's impossible to know the correct position of the flaw on the scan axis, we just know the depth!
Thanks

Paulo

----------- Start Original Message -----------
: Paulo:
: Acceptance criteria will depend on the Code you work to. Unfortunately too many codes are still trying to make ultrasonic results look like radiography. If you can, with great certainty, declare that a TOFD indication obtained from a single line scan (D-scan) is a cluster of porosity, then your assessment skills and confidence is greater than mine.
: An upper "tip" signal will be possible with a cluster of pores (it is likely to look like an irregular set of arrival times). The lower boundary of the indication from such a flaw will be VERY difficult to determine. Inter-pore scatter will generate several paths that will be unresolvable from the diffraction off the lowest portion of the flaw.
: It is permissable in some Codes to assess pores as if they were planar flaws then a length needs to be determined.
: Ed
: : Hello everybody,
: : Please I need a help concerning the Porosity on TOFD inspection. I couldn't find a specific criteria for porosity cluster. Can anybody help me?
: : Thanx!
: : Paulo
------------ End Original Message ------------




 
 Reply 
 
mustafa GONULAL
Engineering,
CIMTAS / ENKA, Turkey, Joined Jun 2008, 8

mustafa GONULAL

Engineering,
CIMTAS / ENKA,
Turkey,
Joined Jun 2008
8
05:30 Jun-04-2008
Re: Porosity on TOFD inspection
Hi, I am a TOFD expert, I have inspected 20km weld with TOFD method. I have seen many clusters of porosity in a weld. The only thing you should do is to measure the distance between the porosities and to look up their homogenities.


 
 Reply 
 

Product Spotlight

Sci Aps Z-Series Portable Handheld Analysers

The world’s only handheld analyzer that measures carbon content in stainless (yes even L-grades),s
...
teels, and cast irons. Also accepted for low Si analysis for sulfidic corrosion analysis, and is widely used in the power industry for Cr analysis, for flow accelerated corrosion applications.
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

On Demand Webinar: Using Eddy Current Array for Large Areas in Place of Magnetic Particle and Liquid Penetrant Testing

New on-demand webinar on the use of eddy current array (ECA) technology in lieu of liquid penetrant
...
testing (PT) and magnetic particle testing (MT) for large area screenings. The webinar includes an introduction to the technologies with advantages and disadvantages of each. Side by side examples are discussed including aircraft stringers, engine blades, forgings, weld plates & more. View it today.
>

32:128PR PAUT & 2-ch TOFD Flaw Detector: SyncScan 2

●32:128PR PAUT with optional 2-ch TOFD to maximize your efficiency ●Support PR mode for corrosi
...
on inspection ●Multiple solutions for welds & corrosion ●Faster scan speed (Approximately 3 meters/minute) ●Light weight: 4kg only including battery
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window