09:18 Jun-12-2008
Ultrasonic A-Scan interpretation

Dear colleagues

There are 2 approaches for setting the range on an A-Scan:
1) at the position of the peak amplitude.
2) at 10% of the peak amplitude on the leading edge (flank).

Historically (for very narrow bandwidth probes) with long ring down times, this 10% story makes a bit of sence, as the peak is not always evident. However in this day and age, this is no longer an issue, and the peak is the correct mathematical location of the reflector (can be proven with broadband probes and CSSP, as all frequencies are in phase at the peak). Since beam skew will give longer measured ranges, an error in the shorter direction is also beneficial (statistically) if the operator does not have the correct wedge angle or does not maximise as he should.
If the gain is then changed, the range for the '10%' setting will change as the leading edge will come closer.

Now comes the question:
In phased array images, the overlay is nearly always placed at the 10% location. To me this is incorrect...
Why isthis done?

Best regards




 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1307

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1307
00:09 Jun-12-2008
Re: Ultrasonic A-Scan interpretation
Philippe:
Good question. I suspect that the assessment of phased array pulses is a carry-over from the tradition of manual single element A-scan assessments. It might be imagined that the "leading edge" of the pulse is the first point of contact and this rationalised the use of "flank". But some of the blame must be put on our trend (over the past 40-50 years) for a preference for rectified A-scans. I suspect that if we were to revert to the real data (unrectified waveforms) we would be more inclined to look for the first well defined peak. On a broadband pulse this will be easier to locate than on a "ringy" narrowband pulse. In fact we seem to have rediscovered the advantage of unrectified waveforms and most of us use a peak for TOFD assessments (phased array generated or single element TOFD is treated the same).
Cheers, Ed


----------- Start Original Message -----------
: Dear colleagues
: There are 2 approaches for setting the range on an A-Scan:
: 1) at the position of the peakamplitude.
: 2) at 10% of the peak amplitude on the leading edge (flank).
: Historically (for very narrow bandwidth probes) with long ring down times, this 10% story makes a bit of sence, as the peak is not always evident. However in this day and age, this is no longer an issue, and the peak is the correct mathematical location of the reflector (can be proven with broadband probes and CSSP, as all frequencies are in phase at the peak). Since beam skew will give longer measured ranges, an error in the shorter direction is also beneficial (statistically) if the operator does not have the correct wedge angle or does not maximise as he should.
: If the gain is then changed, the range for the '10%' setting will change as the leading edge will come closer.
: Now comes the question:
: In phased array images, the overlay is nearly always placed at the 10% location. To me this is incorrect...
: Why is this done?
: Best regards
------------ End Original Message ------------




 
 Reply 
 
mustafa GONULAL
Engineering,
CIMTAS / ENKA, Turkey, Joined Jun 2008, 8

mustafa GONULAL

Engineering,
CIMTAS / ENKA,
Turkey,
Joined Jun 2008
8
01:10 Jun-12-2008
Re: Ultrasonic A-Scan interpretation
----------- Start Original Message -----------
: Philippe:
: Good question. I suspect that the assessment of phased array pulses is a carry-over from the tradition of manual single element A-scan assessments. It might be imagined that the "leading edge" of the pulse is the first point of contact and this rationalised the use of "flank". But some of the blame must be put on our trend (over the past 40-50 years) for a preference for rectified A-scans. I suspect that if we were to revert to the real data (unrectified waveforms) we would be more inclined to look for the first well defined peak. On a broadband pulse this will be easier to locate than on a "ringy" narrowband pulse. In fact we seem to have rediscovered the advantage of unrectified waveforms and most of us use a peak for TOFD assessments (phased array generated or single element TOFD is treated the same).
: Cheers, Ed
:
: : Dear colleagues
: : There are 2 approaches for setting the range on an A-Scan:
: : 1) at the position of the peak amplitude.
: : 2) at 10% of the peak amplitude on the leading edge (flank).
: : Historically (for very narrow bandwidth probes) with long ring down times, this 10% story makes a bit of sence, as the peak is not always evident. However in this day and age, this is no longer an issue, and the peak is the correct mathematical location of the reflector (can be proven with broadband probes and CSSP, as all frequencies are in phase at the peak). Since beam skew will give longer measured ranges, an error in the shorter direction is also beneficial (statistically) if the operator does not have the correct wedge angle or does not maximise as he should.
: : If the gain is then changed, the range for the '10%' setting will change as the leading edge will come closer.
: : Now comes the question:
: : In phased array images, the overlay is nearly always placed at the 10% location. To me this is incorrect...
: : Why is this done?
: : Best regards
------------ End Original Message ------------

First of all I would like to thank you for such a discussion,
My concern is mostly for sizing of defects in Phased Array.
For planar defects like crak, interpreting the Phased Array tomographic view resembles to the TOFD image and you can use tip diffraction signals.
However for other type of defects standards saying 6dB drop method for flaw sizing.
For Phased Array my opinion is different because, you are using a fan of beams coming in different angles with different metal paths. So using 6dB method is too old fashion.


 
 Reply 
 

Product Spotlight

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

NDTkit RT

NDTkit RT, TESTIA's Digital Radiography software The NDTkit product line software for X-ray analysi
...
s. NDTkit RT is a software benefiting from the Ultis kernel which is dedicated to radiographic image analysis. It offers a set of tools and filtering processes to assist RT operators in finding relevant flaws.
>

FD800 Bench Top Flaw Detectors

The bench-top FD800 flaw detector range combines state-of-the-art flaw detection with advanced mater
...
ial thickness capabilities. Designed for use in the laboratory these gauges are the tool you need for all your flaw detecting needs.
>

HD-CR 35 NDT Computed Radiography System

Portable high-resolution CR scanner for all radiography applications - weld testing, profile images
...
and aerospace. No matter what type of radiographic testing you are performing, the unique TreFoc Technology of the HD-CR 35 NDT imaging plate scanner always guarantees the highest image quality.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window