oxie layer measurment in boiler tube

pls inform proceduer to measure thickness of oxide layer through A-Scan instrument thickness gauge step by step for above measurment please also inform what display and where cursor to be set/gate


    
 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
09:11 Aug-12-2008
Re: oxie layer measurment in boiler tube
The specific answer to your question depends on which gage and transducer you are using. Please provide that additional information. If you're using Panametrics-NDT products I can help with the setup, and if you're using something from one of our competitors perhaps someone from that company can step in.

In general, you will always measure from the echo representing the steel/oxide boundary to the echo representing the inside of the oxide layer, so gates are set for detection of those echoes. Some typical waveforms can be found in the application note at the link below.


----------- Start Original Message -----------
: pls inform proceduer to measure thickness of oxide layer through A-Scan instrument thickness gauge step by step for above measurment please also inform what display and where cursor to be set/gate
------------ End Original Message ------------




    
 
 
RickyOng88
Director, -
ASNT, Singapore, Joined Aug 2008, 12

RickyOng88

Director, -
ASNT,
Singapore,
Joined Aug 2008
12
05:34 Aug-14-2008
Re: oxie layer measurment in boiler tube
----------- Start Original Message -----------
: pls inform proceduer to measure thickness of oxide layer through A-Scan instrument thickness gauge step by step for above measurment please also inform what display and where cursor to be set/gate
------------ End Original Message ------------
Whether the Boiler is Fired Tube or Water Tube the best method to be used for measuring the whole length of tube (except bends) is by Internal Rotary Inspection System (IRIS). Spot ultrasonic thickness measurement not good method for boiler tubes only area you could access.



    
 
 
David Hermanutz
Consultant,
Hbndt.com, China, Joined Jul 2012, 85

David Hermanutz

Consultant,
Hbndt.com,
China,
Joined Jul 2012
85
02:50 Aug-21-2008
Re: oxie layer measurment in boiler tube
----------- Start Original Message -----------
: : pls inform proceduer to measure thickness of oxide layer through A-Scan instrument thickness gauge step by step for above measurment please also inform what display and where cursor to be set/gate
: Whether the Boiler is Fired Tube or Water Tube the best method to be used for measuring the whole length of tube (except bends) is by Internal Rotary Inspection System (IRIS). Spot ultrasonic thickness measurement not good method for boiler tubes only area you could access.
------------ End Original Message ------------

The Olympus NDT / Panametrics 37DL Plus has a feature that handles this, please contact your local sales rep and they can show you the way out of your dilemma.

Regards,

David Hermanutz
www.HBNDT.com




    
 
 
RickyOng88
Director, -
ASNT, Singapore, Joined Aug 2008, 12

RickyOng88

Director, -
ASNT,
Singapore,
Joined Aug 2008
12
07:11 Aug-22-2008
Re: oxie layer measurment in boiler tube
----------- Start Original Message -----------
: : pls inform proceduer to measure thickness of oxide layer through A-Scan instrument thickness gauge step by step for above measurment please also inform what display and where cursor to be set/gate

: If you only want to measure thickness of oxide layer try using coating thickness gauge, we use it on stainless steel cladded to steel plate for production oil tanker. We tried using "A" Scan but failed.
------------ End Original Message ------------

Whether the Boiler is Fired Tube or Water Tube the best method to be used for measuring the whole length of tube (except bends) is by Internal Rotary Inspection System (IRIS) an ultrasonic method, as well.

Maybe I misunderstood you for wanted to measure boiler tube thickness.


    
 
 

Product Spotlight

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

Semi-Automated Phased Array Immersion System for Small Composite Parts

Turn-key semi-automated system as an improved and affordable solution for inspection of small comp
...
osite parts. Includes support table, immersion tank, scanner, PA instrument, PC, Analysis software, database, wedge management and other options.
>

Compact NDT inspection-heads for measurements with active thermography

The compact inspection head is suitable for thermographic ndt tasks. The uncooled infrared camera
...
is specially developed for NDI-tasks and offers a thermal sensitivity until now known only from thermal imagers with cooled detector. All required components and functions are integrated into the inspection-head. You will only need an ethernet cable to connect the sensor with the evaluation system.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window