Radiography interpratation Technique-IQI selection

hello guys...
i have some confusion.
1- when interprect elips film.. in order to calculate the sensitivity do we used single wall or 2t .. and is the thicness based on parent material or added with capping ..for example t =4 mm and capping 1.5 as percode... what is the requirement for IQI used.. will it be group A or group B for elips film

2- also on DWSI... for sensity calculation do we used single t or 2t... in order to make or used IQI selection.. please advise.. thank you in advance.


    
 
 
Collin Maloney
NDT Inspector, - Plant Inspector
Applus RTD, Australia, Joined Nov 2000, 147

Collin Maloney

NDT Inspector, - Plant Inspector
Applus RTD,
Australia,
Joined Nov 2000
147
05:55 Sep-20-2008
Re: Radiography interpratation Technique-IQI selection
----------- Start Original Message -----------
: hello guys...
: i have some confusion.
: 1- when interprect elips film.. in order to calculate the sensitivity do we used single wall or 2t .. and is the thicness based on parent material or added with capping ..for example t =4 mm and capping 1.5 as percode... what is the requirement for IQI used.. will it be group A or group B for elips film
: 2- also on DWSI... for sensity calculation do we used single t or 2t... in order to make or used IQI selection.. please advise.. thank you in advance.
------------ End Original Message ------------

Hi

Sensitivity is calculated on single wall thickness including weld cap against thinest wire visible in the weld. Subtract 1 wire if IQI positioned on film side.
Selection of IQI for DWSI is "t".


    
 
 
Juan Amado
Engineering, Inspection
Arco Industrial, S.A., Panama, Joined Nov 2001, 44

Juan Amado

Engineering, Inspection
Arco Industrial, S.A.,
Panama,
Joined Nov 2001
44
06:13 Sep-20-2008
Re: Radiography interpratation Technique-IQI selection
When using this technique (Double Side Double Exposure) it is important to remeber that even though youare looking at the front and back welds in a single film, you are evaluating the quality of a single thickness weld, and hence the required sensitivity level for the radiograph must be set according to that. This also means that in your Ug calculation you shall take into account the total diameter of the pipe, so that both sides of the weld (i.e. the one near the film and the one far away) will meet the Ug requirement.

----------- Start Original Message -----------
: : hello guys...
: : i have some confusion.
: : 1- when interprect elips film.. in order to calculate the sensitivity do we used single wall or 2t .. and is the thicness based on parent material or added with capping ..for example t =4 mm and capping 1.5 as percode... what is the requirement for IQI used.. will it be group A or group B for elips film
: : 2- also on DWSI... for sensity calculation do we used single t or 2t... in order to make or used IQI selection.. please advise.. thank you in advance.
: Hi
: Sensitivity is calculated on single wall thickness including weld cap against thinest wire visible in the weld. Subtract 1 wire if IQI positioned on film side.
: Selection of IQI for DWSI is "t".
------------ End Original Message ------------




    
 
 

Product Spotlight

NovaScope 6000

The all-digital Novascope 6000 is a portable, ultra-high precision thickness gauge for high-speed
...
thickness measurement. Novascope 6000 has unmatched capabilities and unique features including: •Superior Resolution with high contrast, high-speed color RF display •High pulser voltage •Real-time video output •Increased internal/external data storage •Programmable SetUp features •Battery & AC Powered
>

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

FlexoFORM™

The FlexoFORM™ solution integrates flexible phased array probe technology in a scanner to solv
...
e the challenges that come with inspecting pipe elbows. When combined with a water column, this solution can be used to collect easy-to-interpret data on elbows with diameters ranging from 4.5 in. OD up to flat and offers many benefits.
>

NEW - TD Focus-ScanRX

The NEW Next Generation Advanced UT platform, TD Focus ScanRX - Also available as a card stack solut
...
ion. Key Improvements 1. Data acquisition is significantly faster than current design 2. Better aesthetic – closely aligns with HandyScan RX 3. Improved IP rating (Target IP66) 4. Ruggedized housing 5. Connectors are protected from impact and ingress 6. Integrated stand and separate retractable handle easy to keep clean) 7. Touchscreen with ruggedized display glass 8. 3-Axis encoder input
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window