Double wall double image

What is the reasoning behind when taking 3 shots at 120 degrees to each other rather than 0deg,60 & 120


 
 Reply 
 
Uli Mletzko
R & D, Retired
Germany, Joined Nov 1998, 89

Uli Mletzko

R & D, Retired
Germany,
Joined Nov 1998
89
08:18 May-09-2000
Re: Double wall double image
: What is the reasoning behind when taking 3 shots at 120 degrees to each other rather than 0deg,60 & 120

If your term 'double image' is referring to the situation, where you have a 'small' diameter of the tube (> 100 mm) and a 'large' source distance (some 100 mm away from the tube), there is no difference. Either 0/60/120 or 0/120/240 deg. interleave are possible (and allowed due to the European or former German standards).

Only in the situation of 'Double wall single image' (large tube diameter, source near the outer surface of the tube), only 0/120/240 deg are possible, of course.

OK ?

Best regards
Uli Mletzko, NDT Group,
State Materials Testing Institute (MPA),
University of Stuttgart, Germany




 
 Reply 
 
Uli Mletzko
R & D, Retired
Germany, Joined Nov 1998, 89

Uli Mletzko

R & D, Retired
Germany,
Joined Nov 1998
89
08:24 May-09-2000
Re: Double wall double image

: diameter of the tube (> 100 mm)

Sorry for my mistyping. Of course it should be

diameter of the tube (< 100 mm) (less than 100 mm)

Best regards
Uli Mletzko, NDT Group,
State Materials Testing Institute (MPA),
University of Stuttgart, Germany




 
 Reply 
 
Jacques Parent
Jacques Parent
08:49 May-09-2000
Re: Double wall double image
: What is the reasoning behind when taking 3 shots at 120 degrees to each other rather than 0deg,60 & 120

Double wall exposure and superimposed double wall viewing only
0-120-240
or
0-60-120

The 240 and 60 shots provide the same coverage with opposite top and bottom walls.

I believe the 60 shot is prefered because it is easier to set up from a shooting point of view. All three shots are taken from the same quadrant.

Regards
jp




 
 Reply 
 
Nick Welland
Nick Welland
01:12 May-11-2000
Re: Double wall double image
: : What is the reasoning behind when taking 3 shots at 120 degrees to each other rather than 0deg,60 & 120

: Double wall exposure and superimposed double wall viewing only
: 0-120-240
: or
: 0-60-120

: The 240 and 60 shots provide the same coverage with opposite top and bottom walls.

: I believe the 60 shot is prefered because it is easier to set up from a shooting point of view. All three shots are taken from the same quadrant.

: Regards
: jp

I agree that on site the 0, 60, 120 is usually easier to set up and provides the same coverage. But when considering the loss of sensitivity with double-wall viewing, especially superimposed, the 0-120-240 scheme would seem to ensure the favourable near-film location is available in all quadrants.
Nick


 
 Reply 
 
Jacques Parent
Jacques Parent
04:51 May-11-2000
Re: Double wall double image
M. Welland;

I know what you are saying and I do not disagree.
But:
1) 0, 120, 240 shooting on small diameter (<100mm) does not ensure complete coverage on the film side. One still need to evaluate the source side image.

2) Source side IQI's (as required by Code) ensure equivalent and acceptable sensitivity.

In my opinion both techniques are acceptable and the choice should be left to the "shooter"

Regards
jp




 
 Reply 
 
Nick Welland
Nick Welland
02:11 May-12-2000
Re: Double wall double image
: M. Welland;

: I know what you are saying and I do not disagree.
: But:
: 1) 0, 120, 240 shooting on small diameter (<100mm) does not ensure complete coverage on the film side. One still need to evaluate the source side image.

: 2) Source side IQI's (as required by Code) ensure equivalent and acceptable sensitivity.

: In my opinion both techniques are acceptable and the choice should be left to the "shooter"

: Regards
: jp

OK Jacques, c'est entendu, either technique is acceptable. I was trying to see whether there would in fact be any justification for preferring 0-120-240.

Regards
Nick Welland



 
 Reply 
 
john williams
john williams
04:17 May-18-2000
Re: Double wall double image
: What is the reasoning behind when taking 3 shots at 120 degrees to each other rather than 0deg,60 & 120
What is the best method to x-ray two plates of different thicknesses (12mm to 40mm). Also is it true that if you increase the kv by 20 your exposure time is reduced by 50%.





 
 Reply 
 
Uli Mletzko
R & D, Retired
Germany, Joined Nov 1998, 89

Uli Mletzko

R & D, Retired
Germany,
Joined Nov 1998
89
07:07 May-19-2000
Re: Double wall double image
: : What is the reasoning behind when taking 3 shots at 120 degrees to each other rather than 0deg,60 & 120

Unfortunately you have posted your question as follow-up to a wrong subject. Posting of answers therefore may be difficult.

: What is the best method to x-ray two plates of different thicknesses (12mm to 40mm).

You have to know the exposure chart of the x-ray tube (milliamps depending on wall thickness, with voltage as parameter, for a certain film) and the exposure chart (with gamma information) of that film. Then you have the choice for the maximum density (thin wall) and the minimum density (thick wall). The output should be one single voltage, where the task could be solved.
To have a wallthickness difference of 22 mm within a density range of 2 (2.0 to 4.0), you will need high energy like Cobalt 60 or a linear accelerator. The task can't be solved with standard x-ray tubes (you need a flat inclination of exposure chart line).

: Also is it true that if you increase the kv by 20 your exposure time is reduced by 50%.

This is depending not only from the increase, but also from the voltage itself. The output (intensity) of a x-ray tube is depending roughly from the square of the actual peak voltage (integral of or area below the spectral intensity curve). The exposure is depending roughly on the inverse of the intensity.

Regards
Uli Mletzko
MPA, University of Stuttgart, Germany



 
 Reply 
 
saeed payam
NDT Inspector
Iran, Joined May 2002, 1

saeed payam

NDT Inspector
Iran,
Joined May 2002
1
04:39 May-22-2002
Re:film
can you lead me to a site wich it able to show by,online,any welds defect on radiographic film or send it by e-mail to me please.


 
 Reply 
 

Product Spotlight

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

Surf-X® Array Probe

Introducing the Surf-X family of flexible Eddy Current Array (ECA) probes. Featuring unique multiple
...
coil sets and proprietary X-PROBE technology, Surf-X array probes can quickly and accurately handle a range of inspection applications, from inspecting corrosion or cracking in pipes, pressure vessels, or tanks, to assessing and sizing cracks in raised welds and friction stir welds.
>

SONOAIR - air-coupled Phased Array Ultrasonic Inspection System

For highly attenuating materials, the performance of the system is critical. The ultrasonic sensors,
...
the scanning area and the system settings should be flexibly adapted to the test task and the material. These high expectations are met with the new and modular testing system SONOAIR. With the world’s first air-coupled phased-array UT inspection system SONOAIR we developed a technology that works with up to 4 transmitter and receiver channels with freely configurable square wave burst transmitters as wells as low noise receiving amplifiers.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window