Excitation energy effect on shear wave transducers.

Is there any effect on beam angle when changing the Excitation pulse from 100v to 400v as found on the Epoch3?

How would a typical sound pulse react to this increase in energy? Could different lobes be produced resulting in odd echo patterns from a known defect?


    
 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
05:24 May-12-2000
Re: Excitation energy effect on shear wave transducers.
: Is there any effect on beam angle when changing the Excitation pulse from 100v to 400v as found on the Epoch3?

: How would a typical sound pulse react to this increase in energy? Could different lobes be produced resulting in odd echo patterns from a known defect?

There is no effect on beam angle. Increasing the excitation voltage will increase pulse amplitude, but it doesn't change the refracted angle from the wedge, which is based on the Snell's Law relationship between the sound velocities of the wedge insert material and the metal you're testing.

As for the shape of the pulse, it will get bigger, but generally won't change significantly in shape if you're using common transducers designed for angle beam inspection. We normally recommend using the highest excitation voltage (400v) for best signal-to-noise ration in most applications. The 100v setting is most useful when working with high frequency delay line transducers rather than wedges.

Damping changes, on the other hand, MAY alter the shape and phase of the waveform, so you should always do your initial calibration at whatever damping setting you're going to use in your actual inspection.

--Tom Nelligan
Panametrics, Inc.



    
 
 

Product Spotlight

IRIS 9000Plus - Introducing the next generation of heat exchanger inspection.

Representing the seventh generation of the IRIS system, the IRIS 9000 Plus has nearly 200 years of c
...
ombined field inspection experience incorporated in its design. This experience combined with a strong commitment to quality and a history of innovation has made Iris Inspection Services® the undisputed leader in IRIS technology.
>

Swift & Scorpion2 - a state-of-the-art remote-control ultrasonic crawler

The Swift and Scorpion2 dry-coupled, remote-access ultrasonic crawler bring major efficiency and d
...
ata improvements to tank shell inspections and other structures such as vessels and offshore installations.
>

Sci Aps Z-Series Portable Handheld Analysers

The world’s only handheld analyzer that measures carbon content in stainless (yes even L-grades),s
...
teels, and cast irons. Also accepted for low Si analysis for sulfidic corrosion analysis, and is widely used in the power industry for Cr analysis, for flow accelerated corrosion applications.
>

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window