03:01 May-12-2000
Excitation energy effect on shear wave transducers.

Is there any effect on beam angle when changing the Excitation pulse from 100v to 400v as found on the Epoch3?

How would a typical sound pulse react to this increase in energy? Could different lobes be produced resulting in odd echo patterns from a known defect?


 
 Reply 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
05:24 May-12-2000
Re: Excitation energy effect on shear wave transducers.
: Is there any effect on beam angle when changing the Excitation pulse from 100v to 400v as found on the Epoch3?

: How would a typical sound pulse react to this increase in energy? Could different lobes be produced resulting in odd echo patterns from a known defect?

There is no effect on beam angle. Increasing the excitation voltage will increase pulse amplitude, but it doesn't change the refracted angle from the wedge, which is based on the Snell's Law relationship between the sound velocities of the wedge insert material and the metal you're testing.

As for the shape of the pulse, it will get bigger, but generally won't change significantly in shape if you're using common transducers designed for angle beam inspection. We normally recommend using the highest excitation voltage (400v) for best signal-to-noise ration in most applications. The 100v setting is most useful when working with high frequency delay line transducers rather than wedges.

Damping changes, on the other hand, MAY alter the shape and phase of the waveform, so you should always do your initial calibration at whatever damping setting you're going to use in your actual inspection.

--Tom Nelligan
Panametrics, Inc.



 
 Reply 
 

Product Spotlight

Wireless TOFD scanner

Quick, accurate and highly reproducible welds testing. The System operates wirelessly and is compat
...
ible with any type of Windows based Laptop, Desktop or Tablet.
>

Combination of Digital Image Correlation and Thermographic Measurements

The combination of measuring results from the digital image correlation (ARAMIS, DIC) and temperat
...
ure measuring data from infrared cameras permits the simultaneous analysis of the thermal and mechanical behavior of test specimens in the materials and components testing field.
>

HD-CR 35 NDT Computed Radiography System

Portable high-resolution CR scanner for all radiography applications - weld testing, profile images
...
and aerospace. No matter what type of radiographic testing you are performing, the unique TreFoc Technology of the HD-CR 35 NDT imaging plate scanner always guarantees the highest image quality.
>

Varex Imaging Large Field of View (FOV) Digital Detector Arrays (DDAs)

A larger FOV DDA can reduce the space and volume of the X-ray inspection system on the factory floor
...
, enable faster scanning times, better throughput and better resolution images at a lower dose. Customers can also save time and money. With these benefits in mind, Varex Imaging has designed a family of large FOV detectors (4343HE, XRD 1611, 4343DX-I, 4343CT) for our industrial imaging customers.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window