Technical Discussions
Pengzhu Wang
Pengzhu Wang
06:51 Nov-14-2008
Maximum penetration of Ir-192 in steel

Hi,
We are using Ir-192 to examine the casting quality of a steel part with maximum thickness 95.25mm:
(1) Is Ir-192 OK for this application?
(2) Does the decay influence the maximum penetration thickness?
Many thanks,
Pengzhu



    
 
 
tj
tj
07:11 Nov-14-2008
Re: Maximum penetration of Ir-192 in steel
----------- Start Original Message -----------
: Hi,
: We are using Ir-192 to examine the casting quality of a steel part with maximum thickness 95.25mm:
: (1) Is Ir-192 OK for this application?
: (2) Does the decay influence the maximum penetration thickness?
: Many thanks,
: Pengzhu
------------ End Original Message ------------

This would be working at the upper practical limit of Ir192. The energy levels would remain the same with decay. However, with loss of activity comes increased shot time and at 3in. of steel it would be impractical for normal use. Better to use higher energy x-ray or cobalt60 gamma.


    
 
 
Juan Amado
Engineering, Inspection
Arco Industrial, S.A., Panama, Joined Nov 2001, 44

Juan Amado

Engineering, Inspection
Arco Industrial, S.A.,
Panama,
Joined Nov 2001
44
09:53 Nov-14-2008
Re: Maximum penetration of Ir-192 in steel
----------- Start Original Message -----------
: : Hi,
: : We are using Ir-192 to examine the casting quality of a steel part with maximum thickness 95.25mm:
: : (1) Is Ir-192 OK for this application?
: : (2) Does the decay influence the maximum penetration thickness?
: : Many thanks,
: : Pengzhu
: This would be working at the upper practical limit of Ir192. The energy levels would remain the same with decay. However, with loss of activity comes increased shot time and at 3in. of steel it would be impractical for normal use. Better to use higher energy x-ray or cobalt60 gamma.
------------ End Original Message ------------


If 95mm is the maximum thickness, but you have thinner sections, you could probably gain valuable information about the part using Ir192, however, if you are looking to get a good density reading at the thickest portion of the casting you probably will not get it with this source.
Hope this helps, regards,
Juan Amado



    
 
 
Peter Alfers
Peter Alfers
09:53 Nov-16-2008
Re: Maximum penetration of Ir-192 in steel
you can use Ir 192 in conjunction with two Films how will
exposure at once. Than you need the half time for your right density .For the interpretation use the films lying upon each other.


    
 
 

Product Spotlight

CIVA 2017 UT Module

CIVA NDE Simulation Software is the world leader of NDT Simulation. The UT simulation Module incl
...
udes: - "Beam computation": Beam propagation simulation - "Inspection Simulation": Beam interaction with flaws or specimens The user can simulate a whole inspection process (pulse echo, tandem or TOFD) with a wide range of probes (conventional, Phased- arrays or EMAT), components, and flaws.
>

HD-CR 35 NDT Computed Radiography System

Portable high-resolution CR scanner for all radiography applications - weld testing, profile images
...
and aerospace. No matter what type of radiographic testing you are performing, the unique TreFoc Technology of the HD-CR 35 NDT imaging plate scanner always guarantees the highest image quality.
>

Navic - Steerable Modular Automated Scanner

The Navic is a modular, motorized, steerable scanner designed to carry multiple attachments used
...
in various scanning and inspection applications. The Navic is capable of weld scanning (girth welds and long seam welds), automated corrosion mapping, and tank scanning.
>

NDT Master Lecturer

In the program both university professors and practitioners will give lectures, which guarantees the
...
oretical depth and practical inside. Academic Director: Prof. Dr. Christian Boller
The following lecturers are not complete: Prof. Tadeusz Stepinski, Prof. Wieslaw Staszewski, Prof. Frank Walther, Prof. Giovanni Bruno, Prof. Gerd Dobmann, Prof. Philippe Guy
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window
s