PT dwell time variations in reference to temperature extremes

Does anybody have....... or know where i can find.... a table outlining penetrant and developer dwell times with regards to temperature? I know certain codes and standards provide guidelines but I'm looking for something less vague.


    
 
 Reply 
 
S.V.Swamy
Engineering, - Material Testing Inspection & Quality Control
Retired from Nuclear Fuel Complex , India, Joined Feb 2001, 787

S.V.Swamy

Engineering, - Material Testing Inspection & Quality Control
Retired from Nuclear Fuel Complex ,
India,
Joined Feb 2001
787
00:50 Dec-27-2008
Re: PT dwell time variations in reference to temperature extremes
What is vague about them? And the manufacturers provide enough literature, if needed. If the temperature is too low, viscosity will be too high and penetration will not take place properly. If the temperature is too high, the penetrant will dry out and will again not penetrate. If in doubt, use test coupons at the same temperature and see whether you are able to pick up the needed indications.

Swamy
NDT Guru
Hyderabad, India

----------- Start Original Message -----------
: Does anybody have....... or know where i can find.... a table outlining penetrant and developer dwell times with regards to temperature? I know certain codes and standards provide guidelines but I'm looking for something less vague.
------------ End Original Message ------------




    
 
 Reply 
 
Michel Couture
NDT Inspector,
consultant, Canada, Joined Sep 2006, 848

Michel Couture

NDT Inspector,
consultant,
Canada,
Joined Sep 2006
848
09:59 Dec-27-2008
Re: PT dwell time variations in reference to temperature extremes
Hi Jim,

I totally agree with Swamy. Even more I do remember 20 years ago when I got into NDT and dealing mostly with Military Standard that we (in the Canadian Air Force) used MIL-I-6868 to select the penetrant and developer that was compatible with the job we had to do. I do know if it still exist today or if it has been superseded by another code. But, in the end, if you consult the manufacturer's spec. for a product and stay within the temperature range of the code your working to, you shouldn't have any problem. Remember keep it simple and don't try to reinvent the wheel. Many people have spent countless hours to come up with with is already establish and it work.

Cheerio's

Michel

----------- Start Original Message -----------
: What is vague about them? And the manufacturers provide enough literature, if needed. If the temperature is too low, viscosity will be too high and penetration will not take place properly. If the temperature is too high, the penetrant will dry out and will againnot penetrate. If in doubt, use test coupons at the same temperature and see whether you are able to pick up the needed indications.
: Swamy
: NDT Guru
: Hyderabad, India
: : Does anybody have....... or know where i can find.... a table outlining penetrant and developer dwell times with regards to temperature? I know certain codes and standards provide guidelines but I'm looking for something less vague.
------------ End Original Message ------------




    
 
 Reply 
 
Justin Lehmann
Consultant,
Quality Testing Services, Inc, A Premium Inspection and Testing Group Affiliate, USA, Joined Sep 2007, 22

Justin Lehmann

Consultant,
Quality Testing Services, Inc, A Premium Inspection and Testing Group Affiliate,
USA,
Joined Sep 2007
22
03:31 Jan-06-2009
Re: PT dwell time variations in reference to temperature extremes
Jim,

Take a look at ASTM E-1417. It states "The component,
penetrant, and ambient temperatures shall all be in the range from 40 to 125°F [4 to 52°C] unless otherwise specified." It goes on, "The dwell time, unless otherwise specified, shall be a minimum of 10 min. For temperatures between 40 and 50°F [4.4 and 10°C], dwell time shall be a minimum of 20 min."
The ASTM does not address temperatures outside of these ranges.

Justin Lehmann
Manager - Training and Consulting Services
Quality Testing Services, Inc
2305 Millpark Dr
St. Louis, MO 63043
314-770-0607
314-770-0103 fax
314-575-7485 cell
jcl@qualitytesting.net

----------- Start Original Message -----------
: Hi Jim,
: I totally agree with Swamy. Even more I do remember 20 years ago when I got into NDT and dealing mostly with Military Standard that we (in the Canadian Air Force) used MIL-I-6868 to select the penetrant and developer that was compatible with the job we had to do. I do know if it still exist today or if it has been superseded by another code. But, in the end, if you consult the manufacturer's spec. for a product and stay within the temperature range of the code your working to, you shouldn't have any problem. Remember keep it simple and don't try to reinvent the wheel. Many people have spent countless hours to come up with with is already establish and it work.
: Cheerio's
: Michel
: : What is vague about them? And the manufacturers provide enough literature, if needed. If the temperature is too low, viscosity will be too high and penetration will not take place properly. If the temperature is too high, the penetrant will dry out and will again not penetrate. If in doubt, use test coupons at the same temperature and see whether you are able to pick up the needed indications.
: : Swamy
: : NDT Guru
: : Hyderabad, India
: : : Does anybody have....... or know where i can find.... a table outlining penetrant and developer dwell times with regards to temperature? I know certaincodes and standards provide guidelines but I'm looking for something less vague.
------------ End Original Message ------------




    
 
 Reply 
 

Product Spotlight

AIS229 - Multipurpose Real Time System

Latest standard & automatic real time system developed by Balteau. The AIS229 has been designed to
...
do series inspection in a wide variety of industry. Composed of a shielded cabinet, 5 axis manipulator, x-ray generator and tubehead from 160kV to 225kV, a fl at panel & much more, the AIS229 is most certainly one of the most multipurpose RTR system available on the market.
>

Silverwing RMS PA - Phased-array corrosion mapping

Eddyfi Technologies integrated two market leading solutions, Silverwing’s RMS and M2M’s instru
...
ments to provide a high-speed, remote access ultrasonic phased-array inspection system with live total focusing method (TFM).
>

HARDNESS TESTER TKM-459CE combi

TKM-459CE combi applies 2 methods of hardness control: UCI and Leeb. It provides high-accuracy tes
...
ting of metals and alloys as well as items of different sizes and configurations, their hardened layers and galvanic coatings. Device represents results in HB, HRC, HV and others. Shock-, dust- and water-proof housing with intuitive software make this gauge easy to use in all working conditions.
>

FD800 Bench Top Flaw Detectors

The bench-top FD800 flaw detector range combines state-of-the-art flaw detection with advanced mater
...
ial thickness capabilities. Designed for use in the laboratory these gauges are the tool you need for all your flaw detecting needs.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window