08:26 May-20-1998
NDT Newsgroup topic: Manual UT instrument "calibration" frequency

BACKGROUND:
In the currrent NDTnet issue May 1998 we summarize the new European Standard
'EN 126681-1 Part 1:
Characterization and Verification of Ultrasonic Examination - Instruments.'
http://www.ndt.net/article/0598/en126681/en126681.htm

This standard gives methods and acceptance criteria for verifying the
electronic performance of ultrasonic instruments. with A-scan displays,
employed for manual ultrasonic non-destructive examination.
A specific group of these methods and acceptance criteria shall be used
at least once every twelve months to verify the fitness for purpose of an
ultrasonic instrument during it's lifetime. Other two parts of this standard
consists of Probes (2) and Combined Equipment (3).

DISCUSSION:
A discussion on this topic have been hold on the NDT newsgroup.
Find below the intire message followed by its 3 replies.

One main question was: Is it really necessary to return the instruments
for calibration every year?
According to EN 126681-1 (PART 1 !) it is necessary to return the instrument
to the manufactorer (or his agent).
However, notice also the PART 3 'Combined Equipment'.
The Part three of this standard gives methods and acceptance criteria for a
complete system, with the probe, cable and ultrasonic instrument combined.
Initially these shall be used to verify the compatibility of a probe, cable and
ultrasonic instrument and to characterise the system's overall performance.
On site these methods and acceptance criteria shall be used to routinely check
the probe, cable and ultrasonic instrument during their lifetime.

Part 3 shall be performed at regular intervals on site!

You may continue the discussion on the NDT Newsgroup or on the UTonline Forum.
I will try to keep both sites on prospective messages informed.

Rolf Diederichs

-----------------------
Date: Tue, 12 May 1998 22:52:51 GMT
From: yoyak@my-dejanews.com
Subject: Manual UT instrument 'calibration' frequency

I am looking for some input on UT instrument 'calibration'.
Most instruments come with a manufacturer's calibration that expires one year
from the initial date. Is it really necessary to return the instruments
for calibration every year? Please give me your opinions for the following
'types' of instruments:
1.Digital thickness gauges (with or without A-scan feature)
2.Digitally controlled CRT instruments (such as Staveley 136)
3.Digital flaw detectors (such as Panametrics Epoch series of K-B USN series)

All the instruments are set up and calibrated before and after each use. The
flaw detectors have vertical, horizontal and gain control linearities
performed before and after use.

Thanks for your input - you can respond to the address above or post to the
news group.

-----------------------
Date: Wed, 13 May 1998 13:37:54 BST
From: John Alldred

It is worth noting that there are (at least) three different meanings/usages
of the term 'calibration' (of any type of instrument):

(1) Adjustments to bring the indicated values to within the permitted
tolerance of true values (this may be a 'factory only' adjustment);

(2) Verification (without adjustment) that the indicated values are within
tolerance of the true values;

(3) Preparation of a table or graph to allow conversion from indicated
values to true values.

Items (2) and (3) should be able to be performed by the user, although
they may require reference 'standards' or special test equipment;
item (2) is particularly relevant as it may be used to determine whether
an instrument requires a 're-calibration' as per item (1).

There are situations where a user is /competent/ to perform calibration
checks, but is not "accredited" to do so (or believe that to be the case);
this sort of confusion is likely to arise when "QA schemes" are in
operation which are more concerned with weight of paperwork than with
actual practicalities ;-)

John Alldred

----------------
Date: Wed, 13 May 1998 09:31:24 -0400
From: NelliganT@panametrics.com (Tom Nelligan)

The one year expiration period for calibration certification indeed
seems to be almost universal industry practice. We follow that
practice here at Panametrics, as do most other manufacturers. People
seem to expect it, if nothing else.

But to answer your question, it is entirely up to you how often you
return instruments to the manufacturer for calibration
recertification. This in turn will depend on how they're being used,
and what code requirements you may have to follow in your job. The
fact that you are verifying performance before use is an excellent
practice -- I always recommend that to anyone using UT equipment in
critical applications, since the fact that someone put a calibration
sticker on a piece of equipment a few months ago doesn't actually
prove that it's working properly (or more importantly, has been set up
properly by the user) today.

Calibration certification really just shows that the instrument was
working according to manufacturer's specifications on a particular
day. If you're doing a critical test, it's up to you to make sure it's
working properly *today*. If you're doing a non-critical test and
you're satisfied that the instrument is working properly, you may not
need up-to-date cal certification at all. We have some customers who
return instruments for manufacturer's re-certification more than once
a year because they are working to procedures that require it. We have
other customers who re-certify equipment rarely if at all because
their procedures do not require it.

Modern digital instruments tend to fail catastrophically if they go.
Many have self-test protocols that will detect any problems on
start-up or during operation. They are not subject to the sort of
horizontal drift that could affect analog instruments, and vertical
(gain) failures often tend to be obvious too. Most digital gages,
scopes, and flaw detectors will essentially maintain calibration
forever, barring a major (and usually obvious) failure.

Two potential problems that are *much* more common than calibration
drift or failure in digital instruments are incorrect setup by the
operator, or use of an inappropriate or failing transducer. These
problems can be avoided through the implementation of daily (or at
least regular) test procedures of the sort you're probably using.

Good luck!

--Tom Nelligan
Senior Applications Engineer
Panametrics, Inc.

--------------------------
Date: Wed, 13 May 1998 10:46:08 -0700
From: rocky

Calibration is a complex issue and in general each user must make up his
own mind. In some cases calibration is regulated by standards of practice.
The ASME Code for example requires calibration to ASTM E-317, every three
months for Section III. You may wish to look at
http://www.xdcr.com/calibration.html where I discuss calibration at greater
length.

It is necessary to understand that the calibration you do before and after
a shift is not calibration but an only a calibration check or reference
standard calibration. The calibration furnished by the vendor is usually a
certification with traceable standards meeting some standard like ASTM
E-317 or the manufacturers own requirements. The reference standards run
before a test do not meet these requirements and do not meet the
requirements in standards for periodic calibration of measurement & test
equipment.

The basic function of the more stringent requirements is to test for
nonlinearity that might go undetected when running the reference standards.
Such an event is unlikely, although I have seen it occasionally in 25 years
in ultrasonics, so the interval to do the calibration is essentially a
question of risk management. Just how much money or time are you willing
to spend to guard against an unlikely event. Or how much risk your
customers are willing to take if you want to meet particular standards that
specify calibration.

My advise is to examine the standards you must meet and see what they say
about calibration of measurement & test equipment. Even without such
standards I would not ignore the manufacturers recommendation. If their
cal expires in one year I would renew at that time and ask for another
certification and a report on how much the instrument deviated from
specifications. The latter is important because if there are no deviations
from specifications that is reasonable justification to recalibrate on a
longer interval, assuming no standards require an interval.

One caveat is necessary and that is what you will do if the instrument does
deviate from specifications when recalibrated. That is when it does
deviate what effect will that have on all the NDT done with that instrument
since the last valid cal? Only you can answer that question and only you
can determine if a deviation is significant. Because the instruments are
very good compared to what is usually needed you may be able to easily
justify not repeating all inspections done since the last calibration. The
chances of an invalid cal of course increase as you lengthen the
calibration interval. If you have to repeat two or three years of NDT done
with that instrument it can be expensive.

Regards,
Robert (Rocky) A. Day
Second Sound
Ultrasonic Transducers
904 Cortland Avenue
San Francisco, CA 94110 - 5633
(415) 641-4947
Fax: (415) 641-5502




    
 
 

Product Spotlight

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

NDTkit RT

NDTkit RT, TESTIA's Digital Radiography software The NDTkit product line software for X-ray analysi
...
s. NDTkit RT is a software benefiting from the Ultis kernel which is dedicated to radiographic image analysis. It offers a set of tools and filtering processes to assist RT operators in finding relevant flaws.
>

NEW - TD Focus-ScanRX

The NEW Next Generation Advanced UT platform, TD Focus ScanRX - Also available as a card stack solut
...
ion. Key Improvements 1. Data acquisition is significantly faster than current design 2. Better aesthetic – closely aligns with HandyScan RX 3. Improved IP rating (Target IP66) 4. Ruggedized housing 5. Connectors are protected from impact and ingress 6. Integrated stand and separate retractable handle easy to keep clean) 7. Touchscreen with ruggedized display glass 8. 3-Axis encoder input
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window
s