ariestiawan,ronnie
ariestiawan,ronnie
04:23 Jan-16-2004
Pressure gauge calibrator

Any one have information about device for calibrating pressure gauge


    
 
 
frank davis
frank davis
05:07 Jan-19-2004
Re: Pressure gauge calibrator
----------- Start Original Message -----------
: Any one have information about device for calibrating pressure gauge
------------ End Original Message ------------

You will to find someone with a deadweight tester.
Most testers have a low pressure and high pressure
fixtures.


    
 
 
Brian Tolan
Brian Tolan
09:48 Feb-06-2004
Re: Pressure gauge calibrator
Yes, I agree with Mr. Davis concerning the dead weight tester at a calibration shop. I would take a moment; however, to consider the age and type of the gauge in question. I have, at times, calibrated units which were, in my opinion, due to be replaced. It's all about comfort and economy of scale. Would you, by chance, happen to have the specifications and approximate age of the unit in question? Please Advise. Thank You & Best Regards - Tolan


----------- Start Original Message -----------
: : Any one have information about device for calibrating pressure gauge
: You will to find someone with a deadweight tester.
: Most testers have a low pressure and high pressure
: fixtures.
------------ End Original Message ------------




    
 
 
Nagendran
Nagendran
03:41 Jun-23-2004
Pressure Gauge Calibrator
From my experience, to calibrate a pressure gauge, it would be really practical with use of a digital pressure calibrator. Its accurate, easier in calibration methodology and will be able to get the error percentage much better without the need to estimate, which most often done with a dwt. One excellent product is Beamex MC5 series, available at www.beamex.com. You can check your local supplier from its website.

I am not denying the use of a dwt, but there is a place and application which the use of dwt will be more beneficial compared to calibrating pressure gauges. I can forward you more information on calibration methodology using a digital pressure calibrator compared to a dwt, if needed.

You will primarily need to accertain that the pressure gauge is in good working condition. A pre-test of the pressure range indicated will be good. I will also recommend that you shall not over-pressure the gauge more than 90% of its working range so that you may not damage or overs-span the pressure gauge at test.

Should you need more info, reach me at my email.




    
 
 

Product Spotlight

Aerospace Systems - Automated Ultrasonic Inspection

USL are specialists in the design and manufacture of turnkey ultrasonic inspection systems for aer
...
ospace applications. From monolithic composites to complex honeycomb structures. This video shows just a few examples of what is possible, find out more at: www.ultrasonic-sciences.co.uk
>

HD-CR 35 NDT Computed Radiography System

Portable high-resolution CR scanner for all radiography applications - weld testing, profile images
...
and aerospace. No matter what type of radiographic testing you are performing, the unique TreFoc Technology of the HD-CR 35 NDT imaging plate scanner always guarantees the highest image quality.
>

Alpha Pro Digital Radiography System

The Alpha Pro digital X-ray system is the ideal system for your NDT \r\nneeds, as it offers amazin
...
g 16 bit images which enable the highest \r\nlevel of detection and identification of hairline cracks. While weighing only 3kg (6.6lbs), this panel has the largest imaging area: 35 X 43 cm (14X17 in).
>

Exertus Dual 120

The Exertus Dual 120 Projector has the ability to accept Iridium 192 sources or Selenium 75 source
...
s. This projector incorporates design and safety features that make it flexible, compact and lightweight. The Projector is lighter than most of its competitors. It incorporates an improved source channel, based on a new helicoidal design, which makes maintenance easier. The helicoidal design also allows smoother movement of the source assembly inside the device, making it easier for the operator and improving safety. The Projector also has a unique safety feature not found in competitive products. The source assembly locking mechanism is triggered by the source holder capsule at the front of the source assembly, thereby always assuring the operator that the source has returned to the safe position. The Exertus Dual 120 is ISO3999:2004 compliant.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window
s