Elect.Impedance of PZT Pickup


Dear colleagues

For economical purpose I had construct a simple PZT pickup
From PZT -4 Material of Fundamental frequency 2MHZ
When I measured the Electrical Impedance of Pickup against frequency
(Over frequency range 10 KHz to 4 MHZ )
I was found three Peaks
1- first at 250 kHz
2- second at 850 kHz
3- third at 3300 kHz
I Haven't explanation for this result, may be any colleague
has an explanation
Thank you very much



 
 Reply 
 
Jim Barshinger
Engineering
General Electric Global Research Center, USA, Joined Aug 2000, 8

Jim Barshinger

Engineering
General Electric Global Research Center,
USA,
Joined Aug 2000
8
05:44 Sep-18-2000
Re: Elect.Impedance of PZT Pickup
:
: Dear colleagues

: For economical purpose I had construct a simple PZT pickup
: From PZT -4 Material of Fundamental frequency 2MHZ
: When I measured the Electrical Impedance of Pickup against frequency
: (Over frequency range 10 KHz to 4 MHZ )
: I was found three Peaks
: 1- first at 250 kHz
: 2- second at 850 kHz
: 3- third at 3300 kHz
: I Haven't explanation for this result, may be any colleague
: has an explanation
: Thank you very much

The first and second peaks are most likely lateral resonances in the material. If you only have one lateral dimension, for instance a circular or square ceramic, I would guess that the 250 kHz peak is the fundamental lateral resonance and the 850 kHz peak is the third harmonic of the lateral resonance. The 3300 kHz peak should be your fundamental thickness mode resonance. I'm guessing that you are measuring the parallel resonance which will be a point of maximum real impedance. The series resonance should occur slightly lower in frequency and is a point of minimum real impedance. This should occur at 2 MHz (if you have ground your ceramic to the correct thickness). The center frequency that the device will run at depends on whether your recieving electronics matches better with the series or parallel resonance.

Good luck,
Jim





 
 Reply 
 

Product Spotlight

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

iProbe - USB Phased Array Probe

Turn your PC, Laptop or Tablet into a powerful 32:64 Phased Array system with our USB powered Phas
...
ed Array iProbe and inspection software package. Perform affordable weld inspection and corrosion mapping inspections with this powerful but small system. The probe has a 100V pulser, 8K PRF, dual axis encoder input and achieves outstanding signal quality.
>

NEW - TD Focus-ScanRX

The NEW Next Generation Advanced UT platform, TD Focus ScanRX - Also available as a card stack solut
...
ion. Key Improvements 1. Data acquisition is significantly faster than current design 2. Better aesthetic – closely aligns with HandyScan RX 3. Improved IP rating (Target IP66) 4. Ruggedized housing 5. Connectors are protected from impact and ingress 6. Integrated stand and separate retractable handle easy to keep clean) 7. Touchscreen with ruggedized display glass 8. 3-Axis encoder input
>

Conformable wedge transducer

The conformability is obtained with a flexible membrane filled with water between the transducer and
...
the inspected component. The coupling between the membrane and the component requires a small quantity of water or couplant. The conformable wedge combines the acoustic performance of immersion technique with good coupling and low attenuation.
>