00:00 Apr-13-2006
Calibration rule for digital UT instruments

Current ASME (and other) rules for UT scope calibration - e.g. linearity, cal. check frequency, duration etc. are out of date and (I judge) inadequate for current digital instruments with cal storage capabilites. The rules do not address the real issues - verification of cal when switching between, real need to fully re-cal etc.

Also I feel today's instruments stability is far greater than when the Codes were written, so . . . has anyone written or found a good set of practical "rules" that fit today's need. Are there any efforts to codify these rules?


    
 
 
Ed
Ed
09:34 Apr-13-2006
Re: Calibration rule for digital UT instruments
----------- Start Original Message -----------
: Current ASME (and other) rules for UT scope calibration - e.g. linearity, cal. check frequency, duration etc. are out of date and (I judge) inadequate for current digital instruments with cal storage capabilites. The rules do not address the real issues - verification of cal when switching between, real need to fully re-cal etc.
: Also I feel today's instruments stability is far greater than when the Codes were written, so . . . has anyone written or found a good set of practical "rules" that fit today's need. Are there any efforts to codify these rules?
------------ End Original Message ------------

That's a very good question. One that the ASME V Code Committee should probably address. I wonder if anyone has submitted a Technical Inquiry.


    
 
 
LAWRENCE MULLINS
Consultant
NxtNdT, USA, Joined Apr 2006, 7

LAWRENCE MULLINS

Consultant
NxtNdT,
USA,
Joined Apr 2006
7
00:24 Apr-13-2006
Re: Calibration rule for digital UT instruments
----------- Start Original Message -----------
: : Current ASME (and other) rules for UT scope calibration - e.g. linearity, cal. check frequency, duration etc. are out of date and (I judge) inadequate for current digital instruments with cal storage capabilites. The rules do not address the real issues - verification of cal when switching between, real need to fully re-cal etc.
: : Also I feel today's instruments stability is far greater than when the Codes were written, so . . . has anyone written or found a good set of practical "rules" that fit today's need. Are there any efforts to codify these rules?
: That's a very good question. One that the ASME V Code Committee should probably address. I wonder if anyone has submitted a Technical Inquiry.
------------ End Original Message ----------
I think not. I sat w/ SecV last month. I'm writing my own. Does anyone care to co-author and submit as a Code Case?


    
 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1252

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1252
03:25 Apr-13-2006
Re: Calibration rule for digital UT instruments
----------- Start Original Message -----------
: : : Current ASME (and other) rules for UT scope calibration - e.g. linearity, cal. check frequency, duration etc. are out of date and (I judge) inadequate for current digital instruments with cal storage capabilites. The rules do not address the real issues - verification of cal when switching between, real need to fully re-cal etc.
: : : Also I feel today's instruments stability is far greater than when the Codes were written, so . . . has anyone written or found a good set of practical "rules" that fit today's need. Are there any efforts to codify these rules?
: : That's a very good question. One that the ASME V Code Committee should probably address. I wonder if anyone has submitted a Technical Inquiry.
: ------------ End Original Message ----------
: I think not. I sat w/ SecV last month. I'm writing my own. Does anyone care to co-author and submit as a Code Case?
------------ End Original Message ------------

ASTM E07-06 committee has already formed a working group on this and made some revisions to E-317. Similar tests can be run on analogue and digital units but old terminology often gets in the way. e.g. a CRT (cathode ray tube) is now rarely used and the A-scan display is usually on a computer-like monitor.
Larry's comments on stability of instruments is another matter and very valid! Most Codes have "required" calibration intervals that are quite short. For most digital units these could be longer.
Ed



    
 
 
Thomas E Pedersen
Thomas E Pedersen
01:57 Jun-02-2006
Re: Calibration rule for digital UT instruments
----------- Start Original Message -----------
: : : : Current ASME (and other) rules for UT scope calibration - e.g. linearity, cal. check frequency, duration etc. are out of date and (I judge) inadequate for current digital instruments with cal storage capabilites. The rules do not address the real issues - verification of cal when switching between, real need to fully re-cal etc.
: : : : Also I feel today's instruments stability is far greater than when the Codes were written, so . . . has anyone written or found a good set of practical "rules" that fit today's need. Are there any efforts to codify these rules?
: : : That's a very good question. One that the ASME V Code Committee should probably address. I wonder if anyone has submitted a Technical Inquiry.
: : ------------ End Original Message ----------
: : I think not. I sat w/ SecV last month. I'm writing my own. Does anyone care to co-author and submit as a Code Case?
: ASTM E07-06 committee has already formed a working group on this and made some revisions to E-317. Similar tests can be run on analogue and digital units but old terminology often gets in the way. e.g. a CRT (cathode ray tube) is now rarely used and the A-scan display is usually on a computer-like monitor.
: Larry's comments on stability of instruments is another matter and very valid! Most Codes have "required" calibration intervals that are quite short. For most digital units these could be longer.
: Ed
-T----------- End Original Message ------------




    
 
 
Thomas E Pedersen
Thomas E Pedersen
02:03 Jun-02-2006
Re: Calibration rule for digital UT instruments
----------- Start Original Message -----------
: : : : Current ASME (and other) rules for UT scope calibration - e.g. linearity, cal. check frequency, duration etc. are out of date and (I judge) inadequate for current digital instruments with cal storage capabilites. The rules do not address the real issues - verification of cal when switching between, real need to fully re-cal etc.
: : : : Also I feel today's instruments stability is far greater than when the Codes were written, so . . . has anyone written or found a good set of practical "rules" that fit today's need. Are there any efforts to codify these rules?
: : : That's a very good question. One that the ASME V Code Committee should probably address. I wonder if anyone has submitted a Technical Inquiry.
: : ------------ End Original Message ----------
: : I think not. I sat w/ SecV last month. I'm writing my own. Does anyone care to co-author and submit as a Code Case?
: ASTM E07-06 committee has already formed a working group on this and made some revisions to E-317. Similar tests can be run on analogue and digital units but old terminology often gets in the way. e.g. a CRT (cathode ray tube) is now rarely used and the A-scan display is usually on a computer-like monitor.
: Larry's comments on stability of instruments is another matter and very valid! Most Codes have "required" calibration intervals that are quite short. For most digital units these could be longer.
: Ed
-T----------- End Original Message ------------
There is one major factor I see often with the use of digital calibrations,mainly with thickness measurements,People take for granted that the computer goes to "0" digitally but Hardware used cannot resolve the rejectable measurement due to nearfield of the transducer.



    
 
 
Martyn
Martyn
02:08 Jul-28-2006
Re: Calibration rule for digital UT instruments
----------- Start Original Message -----------
: : : : : Current ASME (and other) rules for UT scope calibration - e.g. linearity, cal. check frequency, duration etc. are out of date and (I judge) inadequate for current digital instruments with cal storage capabilites. The rules do not address the real issues - verification of cal when switching between, real need to fully re-cal etc.
: : : : : Also I feel today's instruments stability is far greater than when the Codes were written, so . . . has anyone written or found a good set of practical "rules" that fit today's need. Are there any efforts to codify these rules?
: : : : That's a very good question. One that the ASME V Code Committee should probably address. I wonder if anyone has submitted a Technical Inquiry.
: : : ------------ End Original Message ----------
: : : I think not. I sat w/ SecV last month. I'm writing my own. Does anyone care to co-author and submit as a Code Case?
: : ASTM E07-06 committee has already formed a working group on this and made some revisions to E-317. Similar tests can be run on analogue and digital units but old terminology often gets in the way. e.g. a CRT (cathode ray tube) is now rarely used and the A-scan display is usually on a computer-like monitor.
: : Larry's comments on stability of instruments is another matter and very valid! Most Codes have "required" calibration intervals that are quite short. For most digital units these could be longer.
: : Ed
: -T----------- End Original Message ------------
: There is one major factor I see often with the use of digital calibrations,mainly with thickness measurements,People take for granted that the computer goes to "0" digitally but Hardware used cannot resolve the rejectable measurement due to nearfield of the transducer.
------------ End Original Message ------------

I believe the current code does indeed address the need to verify the calibration when switching from one probe to another, as would be the case when changing cals via the digital scope. One should be very weary of assuming that a stored calibration is still valid, without a reverification. Items such as wear face deterioration, temperature, probe/wedge coupling etc. can all lead to a small, yet significant, variation in calibration. better to just perform the quick and dirty verification as called out in the code, in my opinion.



    
 
 
Csaba Hollo
Csaba Hollo
01:13 Dec-15-2006
Re: Calibration rule for digital UT instruments
----------- Start Original Message -----------
: : : : : : Current ASME (and other) rules for UT scope calibration - e.g. linearity, cal. check frequency, duration etc. are out of date and (I judge) inadequate for current digital instruments with cal storage capabilites. The rules do not address the real issues - verification of cal when switching between, real need to fully re-cal etc.
: : : : : : Also I feel today's instruments stability is far greater than when the Codes were written, so . . . has anyone written or found a good set of practical "rules" that fit today's need. Are there any efforts to codify these rules?
: : : : : That's a very good question. One that the ASME V Code Committee should probably address. I wonder if anyone has submitted a Technical Inquiry.
: : : : ------------ End Original Message ----------
: : : : I think not. I sat w/ SecV last month. I'm writing my own. Does anyone care to co-author and submit as a Code Case?
: : : ASTM E07-06 committee has alreadyformed a working group on this and made some revisions to E-317. Similar tests can be run on analogue and digital units but old terminology often gets in the way. e.g. a CRT (cathode ray tube) is now rarely used and the A-scan display is usually on a computer-like monitor.
: : : Larry's comments on stability of instruments is another matter and very valid! Most Codes have "required" calibration intervals that are quite short. For most digital units these could be longer.
: : : Ed
: : -T----------- End Original Message ------------
: : There is one major factor I see often with the use of digital calibrations,mainly with thickness measurements,People take for granted that the computer goes to "0" digitally but Hardware used cannot resolve the rejectable measurement due to nearfield of the transducer.
: I believe the current code does indeed address the need to verify the calibration when switching from one probe to another, as would be the case when changing cals via the digital scope. One should be very weary of assuming that a stored calibration is still valid, without a reverification. Items such as wear face deterioration, temperature, probe/wedge coupling etc. can all lead to a small, yet significant, variation in calibration. better to just perform the quick and dirty verification as called out in the code, in my opinion.
------------ End Original Message ------------

Martyn,

I believe the discussion involved the instrument performance calibration methods and intervals, such as the 3 monthly requirement in ASME, or the annual or bi-annual outlined in other codes and internal QC documentation.
The inspection calibrations dealing with specific testing parameters including transducers, cables, etc. are most certainly a requirement each and every time the test is performed, or when an inspection variable such as probe wear, different operator, reset from another calibration, etc. occurs. I hope that anyone simply recalling a stored instrument setting and proceeding to test prior to verifying the parameters with the appropriate blocks has thier insurance paid up.

Csaba


    
 
 

Product Spotlight

NDT Master Lecturer

In the program both university professors and practitioners will give lectures, which guarantees the
...
oretical depth and practical inside. Academic Director: Prof. Dr. Christian Boller
The following lecturers are not complete: Prof. Tadeusz Stepinski, Prof. Wieslaw Staszewski, Prof. Frank Walther, Prof. Giovanni Bruno, Prof. Gerd Dobmann, Prof. Philippe Guy
>

Navic - Steerable Modular Automated Scanner

The Navic is a modular, motorized, steerable scanner designed to carry multiple attachments used
...
in various scanning and inspection applications. The Navic is capable of weld scanning (girth welds and long seam welds), automated corrosion mapping, and tank scanning.
>

TOPAZ® 64: Fully Integrated Portable 64 Channel PAUT Instrument with FMC & TFM

Introducing TOPAZ64, the industry’s most intelligent fully integrated, portable 64-channel phased
...
array ultrasound (UT) instrument. TOPAZ64 combines code-compliant phased array UT with the industry’s most advanced full matrix capture (FMC) and total focusing method (TFM) capabilities. Featuring the highest acquisition frequency in its class, high resolution FMC and a 12” multi-touchscreen, users can easily visualize even the smallest flaws. TOPAZ64 can generate a bipolar pulse that provides more acoustic energy versus previous models for punching through thick components. The result is a portable tool that delivers increased inspection coverage, more accurate signals, and the ability to handle all UT inspections in one package. TOPAZ64 is ideal for challenging applications in transportation, oil and gas, manufacturing and power generation.
>

AIS229 - Multipurpose Real Time System

Latest standard & automatic real time system developed by Balteau. The AIS229 has been designed to
...
do series inspection in a wide variety of industry. Composed of a shielded cabinet, 5 axis manipulator, x-ray generator and tubehead from 160kV to 225kV, a fl at panel & much more, the AIS229 is most certainly one of the most multipurpose RTR system available on the market.
>