05:55 Oct-11-2000
temperature effects on plastic wedges

When performing angle beam testing of metals, there is a change in beam angle with temperature. This happens because the velocity change in plastic wedge materials for a given temperature change is larger than the change in the metal.
Beam angle increases with increasing temperature. The effect is most likely to be noticed when testing at high temperatures with a 70 degree wedge. Surface waves can drive you crazy. It has to be almost too cold to work to get
longitudinal waves with a 45 degree wedge. The more common problem is to be wrong about the location of the reflector because of the change in beam angle. I don't have the data with me and won't trust my memory for the numbers; but
I wrote an article about this for Materials Evaluation about 1987. I noticed this problem a couple of years earlier working outside in Wyoming in the winter. Different plastics used for wedges have different temperature coefficients.
If you keep your angle beam calibration reference at the same temperature as the test object, and keep track of the actual angle and attenuation as the temperature changes, that should take care of most problems unless you really have
to have a specific beam angle for a good reason. Attenuation changes can also be due to effects on the couplants.


 
 Reply 
 
William Friedman
R & D,
Lockheed Martin, USA, Joined Nov 1998, 6

William Friedman

R & D,
Lockheed Martin,
USA,
Joined Nov 1998
6
05:12 Oct-12-2000
Re: temperature effects on plastic wedges
: When performing angle beam testing of metals, there is a change in beam angle with temperature. This happens because the velocity change in plastic wedge materials for a given temperature change is larger than the change in the metal.
:
The problem I described in attenuation change occurs just as much when angle beams are not used so I don't think we are seeing an effect that occurs because of a change in angle. The difficulty is that even with a straight beam you might undercall the size of a flaw due when the temperature of the acrylic holder used to couple the transducer to the part.

William Friedman


 
 Reply 
 

Product Spotlight

Conformable wedge transducer

The conformability is obtained with a flexible membrane filled with water between the transducer and
...
the inspected component. The coupling between the membrane and the component requires a small quantity of water or couplant. The conformable wedge combines the acoustic performance of immersion technique with good coupling and low attenuation.
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

Silverwing RMS PA - Phased-array corrosion mapping

Eddyfi Technologies integrated two market leading solutions, Silverwing’s RMS and M2M’s instru
...
ments to provide a high-speed, remote access ultrasonic phased-array inspection system with live total focusing method (TFM).
>

NEW! The PragmaPro Instrument Platform

The PragmaPro is based on a modular cartridge technology and supports various NDT instrument modal
...
ities such as UT, PAUT, ECT and many more. This new platform is based on a machined, powder-coated aluminum frame for shock-proofness, best sealing qualities and maximum heat dissipation. This is practical to extend the outdoor temperature range and/or to extend the power injected in the transducers. The PragmaPro is aiming at a very wide range of applications, such as weld scanning, corrosion mapping and composite testing.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window