03:00 Oct-24-2000
Doubt

Dear experts:
I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
does it cause some anormality to manipulate it?

Kind regards,

Mario Belo


    
 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 176

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
176
09:25 Oct-25-2000
Re: Doubt
reply by Udo Schlengermann

Hello,

We are proud to see an instrument which is more than 20 years out of production in operation:
The USM2 had a calibrated decibel-attenuator (gain) stepped in 2-dB-steps.
To set an echo from a reference reflector exactly to a reference line an additional uncalibrated gain is necessary, if the 2-dB-steps do not fit exactly.
This uncalibrated gain is changed by the screw inside the dB-knob.
It's right position gives the maximum uncalibrated gain, it's left position the minimum uncalibrated gain.
Usually this uncalibrated gain should be set to its max-position.

Kind regards

Udo Schlengermann
+++++++

: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo




    
 
 
Bill Waldron
Sales
GE Inspection Technologies, USA, Joined Feb 2001, 2

Bill Waldron

Sales
GE Inspection Technologies,
USA,
Joined Feb 2001
2
02:48 Oct-25-2000
Re: Doubt
: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo

Mario,
This screw gives 10 dB of gain adjustment and does not affect linearity. Its intended use is to provide a means to fine adjust an echo to a reference level.
(As the USM2 attenuator is in 2 dB steps, it may not be possible to get an echo to an exact level. In this case you would use the small screw adjustment to fine tune the gain level.
Bill Waldron


    
 
 

Product Spotlight

Compact NDT inspection-heads for measurements with active thermography

The compact inspection head is suitable for thermographic ndt tasks. The uncooled infrared camera
...
is specially developed for NDI-tasks and offers a thermal sensitivity until now known only from thermal imagers with cooled detector. All required components and functions are integrated into the inspection-head. You will only need an ethernet cable to connect the sensor with the evaluation system.
>

Research Center IDEKO develops an ultrasonic train wheel inspection system for CAF

The Basque Research centre IK4-IDEKO has developed a state-of-the-art ultrasonic inspection system f
...
or the train wheel. This system secures a sound condition of train wheels and is thus a significant contribution to rail transport safety. The device was delivered to CAF recently and its use allows the manufacturer to become an approved supplier of rolling stock in Italy, as it meets the demanding homologation standards of this country. \\\\r\\\\n
>

NEW! The PragmaPro Instrument Platform

The PragmaPro is based on a modular cartridge technology and supports various NDT instrument modal
...
ities such as UT, PAUT, ECT and many more. This new platform is based on a machined, powder-coated aluminum frame for shock-proofness, best sealing qualities and maximum heat dissipation. This is practical to extend the outdoor temperature range and/or to extend the power injected in the transducers. The PragmaPro is aiming at a very wide range of applications, such as weld scanning, corrosion mapping and composite testing.
>

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window