03:00 Oct-24-2000
Doubt

Dear experts:
I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
does it cause some anormality to manipulate it?

Kind regards,

Mario Belo


 
 Reply 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 181

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
181
09:25 Oct-25-2000
Re: Doubt
reply by Udo Schlengermann

Hello,

We are proud to see an instrument which is more than 20 years out of production in operation:
The USM2 had a calibrated decibel-attenuator (gain) stepped in 2-dB-steps.
To set an echo from a reference reflector exactly to a reference line an additional uncalibrated gain is necessary, if the 2-dB-steps do not fit exactly.
This uncalibrated gain is changed by the screw inside the dB-knob.
It's right position gives the maximum uncalibrated gain, it's left position the minimum uncalibrated gain.
Usually this uncalibrated gain should be set to its max-position.

Kind regards

Udo Schlengermann
+++++++

: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo




 
 Reply 
 
Bill Waldron
Sales
GE Inspection Technologies, USA, Joined Feb 2001, 2

Bill Waldron

Sales
GE Inspection Technologies,
USA,
Joined Feb 2001
2
02:48 Oct-25-2000
Re: Doubt
: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo

Mario,
This screw gives 10 dB of gain adjustment and does not affect linearity. Its intended use is to provide a means to fine adjust an echo to a reference level.
(As the USM2 attenuator is in 2 dB steps, it may not be possible to get an echo to an exact level. In this case you would use the small screw adjustment to fine tune the gain level.
Bill Waldron


 
 Reply 
 

Product Spotlight

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

NDTkit RT

NDTkit RT, TESTIA's Digital Radiography software The NDTkit product line software for X-ray analysi
...
s. NDTkit RT is a software benefiting from the Ultis kernel which is dedicated to radiographic image analysis. It offers a set of tools and filtering processes to assist RT operators in finding relevant flaws.
>

Ultrasonic Squirter/Gantry System

TecScan’s NDT Ultrasonic Gantry Systems are industrial Squirter scanners designed for non-destru
...
ctive quality testing and raster scanning of large structures and parts. The Gantry Systems are usually composed of an Industrial Mechanical Scanner and a Control Room with System Workstation.
>

YXLON Cougar EVO

Scalable small footprint X-ray inspection systems for assembly and laboratory applications. The
...
YXLON Cougar EVO series was designed to provide the "best-in- class" inspection solutions for SMT, semiconductor, and laboratory assembly applications, while maintaining a small system footprint for maximum convenience. With optimized software and hardware, these systems produce higher quality and more consistent results than other electronics inspection systems currently on the market.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window