Dear experts:
I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
does it cause some anormality to manipulate it?

Kind regards,

Mario Belo


 
 Reply 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 183

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
183
09:25 Oct-25-2000
Re: Doubt
reply by Udo Schlengermann

Hello,

We are proud to see an instrument which is more than 20 years out of production in operation:
The USM2 had a calibrated decibel-attenuator (gain) stepped in 2-dB-steps.
To set an echo from a reference reflector exactly to a reference line an additional uncalibrated gain is necessary, if the 2-dB-steps do not fit exactly.
This uncalibrated gain is changed by the screw inside the dB-knob.
It's right position gives the maximum uncalibrated gain, it's left position the minimum uncalibrated gain.
Usually this uncalibrated gain should be set to its max-position.

Kind regards

Udo Schlengermann
+++++++

: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo




 
 Reply 
 
Bill Waldron
Sales
GE Inspection Technologies, USA, Joined Feb 2001, 2

Bill Waldron

Sales
GE Inspection Technologies,
USA,
Joined Feb 2001
2
02:48 Oct-25-2000
Re: Doubt
: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo

Mario,
This screw gives 10 dB of gain adjustment and does not affect linearity. Its intended use is to provide a means to fine adjust an echo to a reference level.
(As the USM2 attenuator is in 2 dB steps, it may not be possible to get an echo to an exact level. In this case you would use the small screw adjustment to fine tune the gain level.
Bill Waldron


 
 Reply 
 

Product Spotlight

A1525 SOLO

A1525 Solo – the most compact and affordable TMF unit with two phased array transducers and 3D v
...
isualization and analysis software in standard delivery set. A compact, ergonomic and easy to handheld Phased Array unit based on Total Focusing Method for easy-going imaging of inspection objects with two-dimensional and three-dimensional visualization and evaluation of inspection results.
>

Extended Range Variable Wall® Piping Calibration Blocks for Longseam Inspection

This set of two blocks has been specially designed for situations where ultrasonic examination tak
...
es place circumferentially or “around the curve” for applications such as inspection of longseam welds in piping and vessels. Jointly designed by PH Tool and Holloway NDT, the blocks feature side-drilled hole reflectors positioned within novel compound-curves, each with multiple thickness steps. The unique design results in a light and compact form factor, with the largest block weighing just over 15 lbs and spanning just 16 ¾”. Coverage is provided for nominal pipe sizes from 6” to 10” in the small block and 12” to a whopping 56” in the large block. The compound OD curves and multiple wall thicknesses actually exceed ASME requirements, providing exceptional signal response and measurement accuracy. Properly managing beam divergence off the ID and stabilizing beam skew due to probe rocking is essential to maintaining a consistent calibration when inspecting lengthwise on curved surfaces.
>

Conformable wedge transducer

The conformability is obtained with a flexible membrane filled with water between the transducer and
...
the inspected component. The coupling between the membrane and the component requires a small quantity of water or couplant. The conformable wedge combines the acoustic performance of immersion technique with good coupling and low attenuation.
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window