08:56 Oct-02-2008
adnan
sonatest sitescan130

Hi guys.. i got sonatest sitescan 130 with no manual available.. just wonder if any guys out there know how to change the ref dB at box dB's .. inside the box, got 'ref dB', then got 'step dB' and the 'actual dB/current dB'... please advise how to change ref dB or set it back to 0 ... currently set at ref 20 .. and the scanning db like 20 is out of the calibration ie... 1.5 side drielled hole with 80 FSH detect the at normal ut set around 40-50 db but this unit like 70-80 dB . so in my point is this ref dB need to turn to 0 // please advise.. thank you in advance with your helps guys...


 
03:39 Oct-02-2008

Justin Lehmann

Consultant, NDT Instructor
Quality Testing Services, Inc,
USA,
Joined Sep 2007
22
Re: sonatest sitescan130 We have a Sitescan 230 with manual. Ours is also set at 20 Ref dB. I could not locate how to reset in the manual or on the instrument itself. I do not, however, believe there is 20 dB added gain to the instrument. Sorry I could not be of more help.

Justin

----------- Start Original Message -----------
: Hi guys.. i got sonatest sitescan 130 with no manual available.. just wonder if any guys out there know how to change the ref dB at box dB's .. inside the box, got 'ref dB', then got 'step dB' and the 'actual dB/current dB'... please advise how to change ref dB or set it back to 0 ... currently set at ref 20 .. and the scanning db like 20 is out of the calibration ie... 1.5 side drielled hole with 80 FSH detect the at normal ut set around 40-50 db but this unit like 70-80 dB . so in my point is this ref dB need to turn to 0 // please advise.. thank you in advance with your helps guys...
------------ End Original Message ------------




 
01:20 Oct-03-2008

collin maloney

NDT Inspector, - Plant Inspector
Applus RTD,
Australia,
Joined Nov 2000
147
Re: sonatest sitescan130 Not familiar with the Sitescan unit, but have you looked to see any of the following:
1. any damping is on
2. most digitals have selectivity of probe frequency, incorrect selection will affect gain required
3. Suppresion
Hope this helps.

----------- Start Original Message -----------
: Hi guys.. i got sonatest sitescan 130 with no manual available.. just wonder if any guys out there know how to change the ref dB at box dB's .. inside the box, got 'ref dB', then got 'step dB' and the 'actual dB/current dB'... please advise how to change ref dB or set it back to 0 ... currently set at ref 20 .. and the scanning db like 20 is out of the calibration ie... 1.5 side drielled hole with 80 FSH detect the at normal ut set around 40-50 db but this unit like 70-80 dB . so in my point is this ref dB need to turn to 0 // please advise.. thank you in advance with your helps guys...
------------ End Original Message ------------




 
04:43 Jan-30-2009
Joel
Re: sonatest sitescan130 In Reply to Justin Lehmann at 03:39 Oct-02-2008 .

Hi guys! Justin - you wouldn't happen to have an electronic copy of your Sitescan 230 manual?!?!

 
15:39 Jan-30-2009

Justin Lehmann

Consultant, NDT Instructor
Quality Testing Services, Inc,
USA,
Joined Sep 2007
22
Re: sonatest sitescan130 In Reply to Joel at 04:43 Jan-30-2009 .

Joel,

Sorry but I do not have the manual electronically.

Justin

 