Corrections calculation

Dear experts:

Is it possible, to determine the transference corrections (surface and attenuation) from a reference block with endless wall (V1/5) to an unknown flaw (smaller than the beam) in a certain piece with different acoustic characteristics and a different surface, which does not present backwall, through the correspondent AVG (DGS) diagram?
An example would be welcome.

Best regards,

Mario Belo



    
 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 174

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
174
09:35 Nov-14-2000
Re: Corrections calculation
reply by Udo Schlengermann

If you calibrate on a block which has different material and surface than the test object, this is not possible, for the following reasons:
Four different losses determine the signal amplitude:
1. divergence - depending on the frequency, the roughness and the size of the contact area.
2. attenuation - in reference block and test object.
3. transmission losses - if the materials are different for both objects (acoustical impedances).
4. reflectivity of reflectors - calibration reflector and detected discontinuity.

The DGS diagram is designed for a full size contact (effective transducer diameter) and a material without attenuation.
Most standards therefore allow DGS method only if the curvature compared to transducer size is below a limit.
Transfer losses on curved, rough objects therefore are only valid for a given contour and roughness, and a defined distance of reflectors, i.e. they change with distance, if attenuation is not comparably equal on both objects.
Before using transfer losses, they have to be determined on a test block having the same curvature and the same roughness as the test object, containing the same reflector (e.g. backwall) at the same distance than your calibration block. Then losses no 4. and 3. are constant in both cases and the remaining difference to the DGS-curve is the transfer loss (attenutation and transmission losses).
It is therefore necessary to have at least one reflector of known size and distance in the test object. You cannot determine transfer losses using an unknown reflector.

Kind regards

Udo Schlengermann
++++++
: Dear experts:

: Is it possible, to determine the transference corrections (surface and attenuation) from a reference block with endless wall (V1/5) to an unknown flaw (smaller than the beam) in a certain piece with different acoustic characteristics and a different surface, which does not present backwall, through the correspondent AVG (DGS) diagram?
: An example would be welcome.

: Best regards,

: Mario Belo




    
 
 

Product Spotlight

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

35 years of VOGT Ultrasonics!

This year we are celebrating a company anniversary! 35 years have passed since we first took our t
...
est device and swung the probe on behalf of safety. But our ultrasonic know-how goes back many years..
>

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window
s