01:49 Sep-10-1996
Is the GE specification good and still relevant?

The GE spec describes the probe's and equipment's capability
for its near to surface resolution.
See also GE Spec in UT A-Z.

For my opinion this specification let to much parameters open,
for instant there is not any dB value mentioned about the
signal noise ratio at the flaw's echo. Is this true?

Are there better standards available?


    
 
 Reply 
 
Robert A. Dau
Engineering
Milky Way Jewels, USA, Joined Nov 1998, 40

Robert A. Dau

Engineering
Milky Way Jewels,
USA,
Joined Nov 1998
40
09:49 Sep-11-1996
Re: Is the GE specification good and still relevant?
Rolf -

Near surface resolution remains a difficult area to
specify because the actual parameters you need depend on
what size defect you need to detect and on the signal-to-
noise you need to reliably detect it. Certainly the GE
specification will not be adequate in some cases, it
will be in others. So my answer is it depends.

One very good way to make a specification for a
particular test is to pin everything down. Specify the
near surface resolution in terms of a reference FBH a
certain distance from the surface and the dB of signal-
to-noise you need. The GE specification is of this type
but not completely rigorous because it doesn't specify
how you measure SN. If you want to detect 0.4 mm FBH 1.2
mm into steel. The GE spec. is good although putting a
more "scientific" definition on the two wave amplitudes
is desirable.

If you want to apply the GE spec. to another material
that is more attenuative for example, then the GE spec.
will be inappropriate. Specifying the near surface
independently from the application is as far as I can
see beyond our current knowledge. The "do everything in
steel and translate to other materials method" is of
limited use to most ultrasonic users because there are
no "translation" codes available to do that. Indeed
there is not much data on attenuation as a function of
frequency out "there."

I might also add that near surface resolution is more
complex than most ultrasonic measurements because it is
a function of almost every parameter you can think off.
Center frequency, band width (damping), material
attenuation, element diameter, focal length, instrument
setting, and water path all change near surface
resolution. So I think you are stuck with using the GE
Spec. and "translating" it experimentally to your needs
or of creating a spec. specific to your own needs.

Regards,
Rocky

: The GE spec describes the probe's and equipment's capability
:for its near to surface resolution.
: See also GE Spec in UT A-Z.

: For my opinion this specification let to much parameters open,
: for instant there is not any dB value mentioned about the
: signal noise ratio at the flaw's echo. Is this true?

: Are there better standards available?




    
 
 Reply 
 

Product Spotlight

Silverwing RMS PA - Phased-array corrosion mapping

Eddyfi Technologies integrated two market leading solutions, Silverwing’s RMS and M2M’s instru
...
ments to provide a high-speed, remote access ultrasonic phased-array inspection system with live total focusing method (TFM).
>

YXLON Cougar EVO

Scalable small footprint X-ray inspection systems for assembly and laboratory applications. The
...
YXLON Cougar EVO series was designed to provide the "best-in- class" inspection solutions for SMT, semiconductor, and laboratory assembly applications, while maintaining a small system footprint for maximum convenience. With optimized software and hardware, these systems produce higher quality and more consistent results than other electronics inspection systems currently on the market.
>

Research and Applications Development For NDT

The Research and Applications Development (RAD) group is a newly formed team within Acuren dedicat
...
ed to tackling challenging inspection problems. Our focus is the development of novel, field deployable, advanced inspection techniques for use in cases where standard NDT methods are ineffective. We don't wait for new innovations, we engineer them. From concept to commissioning.
>

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window