Stan Banner
NDT Inspector,
Canada, Joined Jan 2009, 31

Stan Banner

NDT Inspector,
Canada,
Joined Jan 2009
31
21:21 Jan-12-2009
ASME V Article 4 Calibrations

Hi:
I have a few questions regarding how to choose a what part of the ASME V Article 4 is appropriate for a given weld examination, and the calibration block requirements:
1 - Do all of the rules for Flat ASME calibration blocks apply to Curved Nonpiping calibration blocks, e.g. the need for both slots and side drilled holes and how are circumferentially drilled holes handled, and do slot depths need to be 2% as stated for nonpiping blocks or 10% as stated for piping blocks. If 2% is the case, that means that a non piping block made of 2.5" s80 pipe would need slots <0.006" deep, which seems to me quite extreme, and the side drilled hole would be around 1/3 the pipe thickness.
2 - When examining a pipe welded to a vessel, e.g. a weld on the the drain pipe welded on the bottom of a tower, but not actually welded to the vessel envelope, is the appropriate cal block a piping cal block, or a curved nonpiping block.
3 - When examining a branch connection weld on a piping system, e.g. a 4" 'Cat D' weld onto a 10" pipe, which of the two cal block specification come into play.
To sum up the last two points, when examination is to ASME VIII, are only nonpiping cal blocks used, while when examination is to ASME B31.3, are only piping cal blocks used.

Stan Banner

    
 
 
Phil Herman
Phil Herman
21:44 Jan-12-2009
Re: ASME V Article 4 Calibrations
In Reply to Stan Banner at 21:21 Jan-12-2009 (Opening).

Stan,
The 2007 ASME Code, Section V Article 4 contains Fig. T-434.3 which will answer many of your questions from #1. In summary, there are no side-drilled holes that you are familiar with from the Flat Non-piping blocks. Also, the depths of the 4 EDM notches are nominally 9.5% deep (8 to 11% of nominal wall is specified), not the 2% called out on the flat block.
Phil

    
 
 
Roger Duwe
NDT Inspector, API-510, 570, 653
MISTRAS, USA, Joined Jan 2009, 148

Roger Duwe

NDT Inspector, API-510, 570, 653
MISTRAS,
USA,
Joined Jan 2009
148
02:47 Jan-13-2009
Re: ASME V Article 4 Calibrations
In Reply to Stan Banner at 21:21 Jan-12-2009 (Opening).

Re #2: The U-1 Vessel Code Data Report will show if this weld in part of the vessel, or outside the boundary of the vessel.

#3 By definition, Cat 'D' welds are inside the Sect VIII code boundary, so test per ASME VIII.

Pipe examines per B31.3 [or whatever is appropriate], vessel by vessel code, Code Data Report shows the boundary.

    
 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1252

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1252
14:52 Jan-15-2009
Re: ASME V Article 4 Calibrations
In Reply to Roger Duwe at 02:47 Jan-13-2009 .

Stan:
Just a couple of observations on ASME Section V and piping versus non-piping weld examinations.
It is interesting to note that the function of the notches on the piping calibration blocks is in fact to set sensitivity. These are nominally 10% of the wall with some tolerance actually noted as a range from 8-11%. Setup using a phased-array S-scan using this notch configuration is problematic at best due to the reflectivity effects from a vertical face.
When we move to plate inspections (including curved plate) it is worthwhile noting that although the Figure of the Basic Calibration Block in T-434.2.1 includes a surface notch 2% of the wall thickness, its function is rarely used (and generally poorly understood). The function of the 2% notch in the Basic Calibration Block is described in the Non-mandatory Appendices. Appendix B in Sec. V Art. 4 describes the initial use of the notch in B-464.4. The first application is simply to “note” the position of the surface during the position calibration of the A-scan display. B-465 describes the second use of that 2% notch. As I noted above, the issue of calibrating sensitivity on a notch in piping is problematic. This problem is “described” in B-465 suggesting that the response from the 2% surface notch in a 45° beam may be 6dB more sensitive than the response from the SDH–based DAC. Yet the same notch may be 6dB less than the DAC for a 60° beam. ASME describes that “a correction is required in order to be equally critical of such an imperfection”. However, this is not compatible with the treatment of responses in the normal workmanship-style acceptance criteria used in ASME referencing Sections. I.e., although this Appendix acknowledges that the response from a corner reflector will be angle dependent it still requires reporting based solely on amplitude response so the instructions are (seemingly) useless as you are obliged to record all responses (regardless of beam angle and corner effects) exceeding a specified amplitude.
Ed

    
 
 

Product Spotlight

Semi-Automated Phased Array Immersion System for Small Composite Parts

Turn-key semi-automated system as an improved and affordable solution for inspection of small comp
...
osite parts. Includes support table, immersion tank, scanner, PA instrument, PC, Analysis software, database, wedge management and other options.
>

ISONIC 3510

Very Powerful Superior Performance Extremely Portable Smart Phased Array Ultrasonic Flaw Detector an
...
d Recorder with 2 Conventional UT and TOFD Channels uniquely combines PA, single- and multi-channel conventional UT, and TOFD modalities providing 100% raw data recording and imaging. Suitable for all kinds of every-day ultrasonic inspections
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

TOPAZ® 64: Fully Integrated Portable 64 Channel PAUT Instrument with FMC & TFM

Introducing TOPAZ64, the industry’s most intelligent fully integrated, portable 64-channel phased
...
array ultrasound (UT) instrument. TOPAZ64 combines code-compliant phased array UT with the industry’s most advanced full matrix capture (FMC) and total focusing method (TFM) capabilities. Featuring the highest acquisition frequency in its class, high resolution FMC and a 12” multi-touchscreen, users can easily visualize even the smallest flaws. TOPAZ64 can generate a bipolar pulse that provides more acoustic energy versus previous models for punching through thick components. The result is a portable tool that delivers increased inspection coverage, more accurate signals, and the ability to handle all UT inspections in one package. TOPAZ64 is ideal for challenging applications in transportation, oil and gas, manufacturing and power generation.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window