Penetrameter in Radigraphy

As per code penetrameter shall be kept in souce side.What is the reason for same?In some cases we are unable to keep in source side (kpet in film side).What will be the disadvantage?

    
 
 Reply 
 
tj
tj
15:24 May-14-2009
Re: Penetrameter in Radigraphy
In Reply to Manish at 11:54 May-14-2009 (Opening).

Manish,

If the penetrameter is placed on film side it is not susceptible to the same unsharpness variables that are present with the specimen itself (see geometric unsharpness in your code/standard). Some codes will allow this placement in certain cases if the penetrameter is marked correctly (with an F).

    
 
 Reply 
 
Roger Duwe
NDT Inspector, API-510, 570, 653
MISTRAS, USA, Joined Jan 2009, 148

Roger Duwe

NDT Inspector, API-510, 570, 653
MISTRAS,
USA,
Joined Jan 2009
148
19:31 May-14-2009
Re: Penetrameter in Radigraphy
In Reply to tj at 15:24 May-14-2009 .

Manish, look at AWS D1.1 or ASME Sect V. They both have penetrameter charts based on the material thickness. When the penetrameter is placed Film-Side, a smaller one is mandated. Thus the requirement for the lead letter 'F' to be placed on the penetrameter. As TJ noted, a Film-Side penetrameter will be much clearer in the RT image, so a smaller one is required to prove the sensitivity of the image.

    
 
 Reply 
 
Neil Burleigh
Sales
Krautkramer Australia Pty Ltd, Australia, Joined Dec 2002, 157

Neil Burleigh

Sales
Krautkramer Australia Pty Ltd,
Australia,
Joined Dec 2002
157
00:10 May-15-2009
Re: Penetrameter in Radigraphy
In Reply to Manish at 11:54 May-14-2009 (Opening).

Hello Manish,
I have seen the posts in response to your request they have all provided the correct information but I think they have not answered your question.
1. Penetrameter what is the other name for this item? It is IQI or image quality indicator.
What is a radiographic image? It is an image of a part that has been exposed to radiation showing density variations. Therefore a penetrameter confirms the minimum resolvable density variation.
From this you should be able to determine why the IQI is on the source side or as Roger Duwe has said a smaller penetrameter is placed on the film side to prove sensitivity.

Regards

    
 
 Reply 
 
tortoro
tortoro
00:08 May-16-2009
Re: Penetrameter in Radigraphy
In Reply to Neil Burleigh at 00:10 May-15-2009 .

u did not answer his question u only explained what an IQI is. the question was what is the disadvantage.

    
 
 Reply 
 
Nilesh Pathare
Nilesh Pathare
14:28 May-16-2009
Re: Penetrameter in Radigraphy
In Reply to Manish at 11:54 May-14-2009 (Opening).

Dear Manish,
As per the code requirement the Penetrameter shall be kept at source side whenever practicable.
This is required for -
1. To check the Ug unsharpness is not exceeding the acceptable limit. If Ug exceeds then you will not able to see the required wire or hole for that Thickness.
2. It ensures that it will show the minimum discontinuity throughout the job thickness equivalent to wire dia. or hole dia. when you can see the particular wire or hole when the Penetrameter is kept on Source side.
3. Whenever you are keeping the Penetrameter on film side code asks to view smaller diameter wire /hole to compensate the absorption through the job thickness.

    
 
 Reply 
 
tj
tj
02:36 May-18-2009
Re: Penetrameter in Radigraphy
In Reply to Nilesh Pathare at 14:28 May-16-2009 .

1. Just because Ug exceeds acceptable limits, this does not mean that you will not see the penetrameter holes.

2. I would not refer to the penetrameter holes as guidlines for minimum discontinuity or in any other way related to acceptance criteria. It is to determine sensitivity of the radiograph. This can confuse new technicians.

3. Compensating for absorption has nothing to do with film/source side placement. It is simply the geometric effects related to forming the image.

    
 
 Reply 
 
NDTIII
NDTIII
07:03 May-18-2009
Re: Penetrameter in Radigraphy
In Reply to tj at 02:36 May-18-2009 .

Thank you NDT Inspector. I'd just like to elaborate a little on the subject.

I believe the intent of source side placement is merely to place the IQI in the least favorable location under the least favorable conditions, where practicable.

Remember the IQI is the only means we have to judge the quality of the image or technique. Nothing more.

    
 
 Reply 
 
tortoro
tortoro
13:57 May-18-2009
Re: Penetrameter in Radigraphy
In Reply to NDTIII at 07:03 May-18-2009 .

now that answers the question

    
 
 Reply 
 

Product Spotlight

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

Panther

M2M PANTHER is a phased-array equipment designed for both desktop and industrial applications, offer
...
ing unparalleled performance in a compact unit. It combines the speed required for industrial integrated Phased-Array Ultrasound (PAUT) solutions, with the most complete set of total focusing method (TFM) imaging techniques, making it the ultimate tool for R&D and procedure qualification.
>

Sci Aps Z-Series Portable Handheld Analysers

The world’s only handheld analyzer that measures carbon content in stainless (yes even L-grades),s
...
teels, and cast irons. Also accepted for low Si analysis for sulfidic corrosion analysis, and is widely used in the power industry for Cr analysis, for flow accelerated corrosion applications.
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window