0.5mm reflector on V1 and EN 12223 Cal blocks

Colleagues,
We had an opportunity to recalibrate a customer-supplied EN 12223 cal block no. 1 this week and noticed something interesting that I wanted to get some feedback on. The zero point slot (on the 100mm radius centerline) is specified as 4.0mm deep x 0.5mm wide x 30.0mm long. No end view of this slot is provided in the EN. BS 2704 includes an end view of this feature and clearly indicates that the slot depth is to be uniform over the length except for the inboard ends which are radiussed (to allow conventional machining using a slitting saw.) PH Tool opts to machine the slots this way. The block in question was clearly machined via wire EDM as was evidenced by a cast replica. As a result, the observed slot was angled or tapered from 0.0 to 4.0mm deep over its length. In addition, rather than having square-cornered features, the slot exhibited a corner radius of .25mm, typical of wire EDM. My question is this: Is the angled slot with radiussed corners in compliance with EN 12223 and would a reflector machined in this manner function as intended? Considerable time would be saved machining it via wire, but in my estimation, an inferior product will result. Any comments would be greatly appreciated.

    
 
 Reply 
 
Roger Duwe
NDT Inspector, API-510, 570, 653
MISTRAS, USA, Joined Jan 2009, 148

Roger Duwe

NDT Inspector, API-510, 570, 653
MISTRAS,
USA,
Joined Jan 2009
148
00:04 May-28-2009
Re: 0.5mm reflector on V1 and EN 12223 Cal blocks
In Reply to Phil Herman at 17:14 May-27-2009 (Opening).

Phil, please send me a copy on EN 12223. I suspect the answer will be right there.

    
 
 Reply 
 
Dave Davies
Dave Davies
10:18 May-28-2009
Re: 0.5mm reflector on V1 and EN 12223 Cal blocks
In Reply to Phil Herman at 17:14 May-27-2009 (Opening).

Phil,
As you had, presumably, both types of block to hand you could have answered your own question about functionality with some side by side comparisons.

As the purpose of the slot is to allow repeats from the 100mm radius I would guess that the actual form of the slot is not critical so long as it works.

As for the spec then a question to the relevant working group at ISO would clear that up.

I have had problems with these blocks before where units with imperial and metric dimensions exist side by side with no marking to differentiate. I had to go and sort them out and hard stamp myself to avoid mix ups with the inspectors.

    
 
 Reply 
 
Roger Duwe
NDT Inspector, API-510, 570, 653
MISTRAS, USA, Joined Jan 2009, 148

Roger Duwe

NDT Inspector, API-510, 570, 653
MISTRAS,
USA,
Joined Jan 2009
148
23:19 May-28-2009
Re: 0.5mm reflector on V1 and EN 12223 Cal blocks
In Reply to Dave Davies at 10:18 May-28-2009 .

Concurr with Dave D. -- the slots are just to lobtain a 200 mm reflection to go with the 100 mm reflection. The lack of parallelism [specifically, not parallel to the 100 mm radius] hurts the quantity & quality of sound to generate the 200 peak. The radius at the bottom of this slot on the corner closest to the 100 radius also decreases the available surface parallel to the radisused end, giving more loss of sound. The 'implied tolerance' is 4.0 +/- 0.1 mm of flat face parallel to the radiused end of the block. Since that tolerance is not expressed in EN 12223, it falls under 'workmanship', and is not mandatory.

In short, Phil is right. Poor workmanship to cut the block and leave a tapered draft and radiused bottom corners. However, it meets the strict interpretation of the EN 12223 dwg's and will work as a cal block. Second peak will just be smaller, broader, and 'fuzzier'.

And I will continue to buy my reference blocks from PH Tool. What's a little more money for a prime reference that will still be useable in 2-3 more generations? 'First Rate' always has cost a little more.

    
 
 Reply 
 

Product Spotlight

Semi-Automated Phased Array Immersion System for Small Composite Parts

Turn-key semi-automated system as an improved and affordable solution for inspection of small comp
...
osite parts. Includes support table, immersion tank, scanner, PA instrument, PC, Analysis software, database, wedge management and other options.
>

XRHRobotStar

In high volume industries like automotive the requirement for a hundred percent X-ray inspection c
...
reates a bottleneck in the production. The XRHRobotStar is a fully Automated Defect Recognition (ADR) capable robot-system that allows an ultra-fast in-line inspection.
>

NovaScope 6000

The all-digital Novascope 6000 is a portable, ultra-high precision thickness gauge for high-speed
...
thickness measurement. Novascope 6000 has unmatched capabilities and unique features including: •Superior Resolution with high contrast, high-speed color RF display •High pulser voltage •Real-time video output •Increased internal/external data storage •Programmable SetUp features •Battery & AC Powered
>

GEKKO - Portable Phased Array Testing with TFM in Real-Time

The portable phased array testing system GEKKO provides 64 parallel test channels. On creating testi
...
ng parameters the operator is assisted by the CIVA software. Due to its modular set-up the GEKKO instrument is suitable for operators of all skill levels.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window