EMI Flaw Report

Dear all,

Can anyone explain me about elecromagnetic inspection

When i calibrate EMI unit for inspecting seamless pipe with 10% notches both ID & OD
for longitudinal & transverse orientations, all peak signals approxcimate 50% FSH.Thus running for inspection, the unit catch 21% fsh on the chart but i interested to evaluate the indication. During evaluation proven that indication from OD Pitting with depth 12% of nominal thickness. My question are:

~Why indication appears on ID?
~Why indication of OD pitting less than Notches, whereas deeper than ??

Regards
Kusnadie

    
 
 Reply 
 
Andreas
Andreas
22:42 Jun-02-2009
Re: EMI Flaw Report
In Reply to kusnadie at 11:24 Jun-01-2009 (Opening).

Hallo ,
it looks like you operating a four function emi unit.
Separation of flaw for od or id location is based on some unit about the frequence of the flaw signal, but does not worked everytime correctly.
Anyway round bottom pitting on the external surface hardly produce a fluxleakage.Normally you see that better on the longitudinal system.
For used pipe normally a 1/16" drilled through hole is used.You did not mention what kind of pipe or which spec you used.
A notch produce is higher flux leakage and therefor mechical defects like couges ,cut etc will overn evaluted and pitting under estimated.
A good indication is the frequency of your flaw indication.

Regards

Andreas

    
 
 Reply 
 
S.Senthilkumar
Engineering, QA/QC/NDT
NOV (National Oilwell Varco), Angola, Joined Mar 2006, 41

S.Senthilkumar

Engineering, QA/QC/NDT
NOV (National Oilwell Varco),
Angola,
Joined Mar 2006
41
22:01 Jun-03-2009
Re: EMI Flaw Report
In Reply to Andreas at 22:42 Jun-02-2009 .

As Andreas,
I am using 1/16" hole in a pipe helically which give the signal pefectly in pipes. 6 to 8 holes which gives the reference signals.

Regards,
S.Senthilkumar

    
 
 Reply 
 
Kusnadie
Kusnadie
15:30 Jun-04-2009
Re: EMI Flaw Report
In Reply to S.Senthilkumar at 22:01 Jun-03-2009 .

Hi Andreas,

Kind of pipe is seamless pipe with OD 9 5/8", 0.380" nominal wall thickness. I am using 10 % notches depth, 0.040" width and 2" length as API 5CT Spec.

Regards
Kusnadie

    
 
 Reply 
 
Andreas
Andreas
17:21 Jun-04-2009
Re: EMI Flaw Report
In Reply to Kusnadie at 15:30 Jun-04-2009 .

Make up ypur mind ,you using 1/16" holes or notches or both ?
Or using the notches for the longitudinal system and holes for the transversal system.
On standard unit there should be 8 holes to check trail and lead sensors in one go.
As i sayes on the first replay OD pitting have hardly a flux leakage and therefore a smaller amplitude.
OD/ID flaw separation on your unit seems to be by frequency of the flaw response.

Normally a 10% notch is used for new casing acc API Standard API5CT.On new casing you looking for defeccts from the manufacturing process lile seams,overlap etc.

On used pipe you looking mainly for mechanical defects/flaws and corrossion.The flaw response from this defects are differrent and you need to establish your on reference level according to the defect to be found.

If you inspecting used casing you should use the API RP 5C1.


Regards

Andreas

    
 
 Reply 
 
Kusnadie
NDT Inspector, -
PT SEAMLESS PIPE INDONESIA JAYA, Indonesia, Joined Apr 2007, 5

Kusnadie

NDT Inspector, -
PT SEAMLESS PIPE INDONESIA JAYA,
Indonesia,
Joined Apr 2007
5
16:09 Jun-05-2009
Re: EMI Flaw Report
In Reply to Andreas at 17:21 Jun-04-2009 .

Hello Andreas,

I would like to thank for your suggests, opinions and supports.
Your suggests supported me to establishes any manner in inspection better.

Regards
Kusnadie

    
 
 Reply 
 

Product Spotlight

ISONIC 3510

Very Powerful Superior Performance Extremely Portable Smart Phased Array Ultrasonic Flaw Detector an
...
d Recorder with 2 Conventional UT and TOFD Channels uniquely combines PA, single- and multi-channel conventional UT, and TOFD modalities providing 100% raw data recording and imaging. Suitable for all kinds of every-day ultrasonic inspections
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

AMIGO2

TSC Amigo2 - ACFM technology has developed a solid reputation for accurately detecting and sizing
...
surface-breaking cracks through paint and coatings. As the industry demands increased performance in speed, signal quality, and portability, it’s time for an evolution. It’s time for Amigo2.
>

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window