Phased Array Steering Limits - Calculations

I wonder how the calculation for the beam steering in a phased array transducer goes:

For circular transducers the formula is 0.514 X (Wavelength)/ (Dia. of the transducer). How does that translate to finding the steering limits of a 5MHz, 16 element probe with a pitch of 0.6mm and a smallest possible element size of 0.2mm. The calculations do not make much sense when done but I know there was certain discussions before here about comparing rectangular and circular transducer.

    
 
 
Roger Duwe
NDT Inspector, API-510, 570, 653
MISTRAS, USA, Joined Jan 2009, 148

Roger Duwe

NDT Inspector, API-510, 570, 653
MISTRAS,
USA,
Joined Jan 2009
148
16:44 Jun-29-2009
Re: Phased Array Steering Limits - Calculations
In Reply to Vikram at 07:31 Jun-29-2009 (Opening).

Just a thought -- why not 'ask the transducer'. Try setting up a DAC curve using a 40 to 70 degree sectoral scan ['ducer on a wedge of approx 50 deg]. If it calibrates OK, try 35 to 75 or similar. If it cal's poorly, reduce the span of the scan angle.

My experience with 5L16's [5 MHz Linear 16-element] is that 50 to 70 degrees calibrates fine, 45 to 70 is problematic. Calculations make quite a few assumptions that may or may not hold true in the 'Real World'. Eventually, you will have to calibrate your system with that 'ducer. Why not now?

    
 
 
Vikram
NDT Inspector, - Phased Array Specialist
Mistras Group Inc., USA, Joined Jun 2009, 9

Vikram

NDT Inspector, - Phased Array Specialist
Mistras Group Inc.,
USA,
Joined Jun 2009
9
20:48 Jun-29-2009
Re: Phased Array Steering Limits - Calculations
In Reply to Roger Duwe at 16:44 Jun-29-2009 .

Funny that you mention it, I just did the calculations for a SDH of 0.0625" to check the beam steering capability for an L-wave. I am going to extend the same for shear wave and get the results.

    
 
 

Product Spotlight

Ultrasonic tomograph for imaging of concrete structures А1040 MIRA

Applicable for concrete inspection allowing imaging of the internal structure of objects from conc
...
rete, reinforced concrete, different stones. The operation applies pulse-echo technique at one-side access to the object. The instrument is feasible for concrete inspection for searching conduct ducts, conduits, detection of foreign inclusions, holes, honeycombing, cracks and other concrete defects.
>

IRIS 9000Plus - Introducing the next generation of heat exchanger inspection.

Representing the seventh generation of the IRIS system, the IRIS 9000 Plus has nearly 200 years of c
...
ombined field inspection experience incorporated in its design. This experience combined with a strong commitment to quality and a history of innovation has made Iris Inspection Services® the undisputed leader in IRIS technology.
>

Research and Applications Development For NDT

The Research and Applications Development (RAD) group is a newly formed team within Acuren dedicat
...
ed to tackling challenging inspection problems. Our focus is the development of novel, field deployable, advanced inspection techniques for use in cases where standard NDT methods are ineffective. We don't wait for new innovations, we engineer them. From concept to commissioning.
>

Swift & Scorpion2 - a state-of-the-art remote-control ultrasonic crawler

The Swift and Scorpion2 dry-coupled, remote-access ultrasonic crawler bring major efficiency and d
...
ata improvements to tank shell inspections and other structures such as vessels and offshore installations.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window