Stan
NDT Inspector,
Canada, Joined Jan 2009, 31

Stan

NDT Inspector,
Canada,
Joined Jan 2009
31
16:44 Jul-07-2009
PA probe focusing

Hi:
I am hoping that some one will help me out here.
I was taught that with a traditional probe, if focusing is applied to it, the focal point must be less than the distance to the near zone, that is, focusing can only bring the focal point closer to the probe, you cannot focus the probe farther away from the probe than the near zone.
I was also taught that a phased array probe of a given virtual aperture has the same properties as a conventional probe of the same size. To me this implies that the probes focusing properties should be similar to those of a conventional probe, however, perhaps when using the phased array focusing this is not the case. Is it possible to focus a phased array probe farther from the probe than the end of the near zone? If so, what effect does this have on the size of the focal spot? Generally, the farther the focal distance is from the probe, the larger the focal spot size.
Will someone enlighten me here.
Thanks
Stan

 
 Reply 
 
Roger Duwe
NDT Inspector, API-510, 570, 653
MISTRAS, USA, Joined Jan 2009, 148

Roger Duwe

NDT Inspector, API-510, 570, 653
MISTRAS,
USA,
Joined Jan 2009
148
20:21 Jul-07-2009
Re: PA probe focusing
In Reply to Stan at 16:44 Jul-07-2009 (Opening).

A PA probe dows not have the 'same properties' as a conventional 'ducer. The focus depth of the PA 'virtual probe' can be varied based on material thickness, and should be refocused for actual indications found [assuming that the defect was not readily apparant and easily 'diagnosed' during the initial scan]. Additionally, PA is at its best in the nearfield, where conventional shearwave works best past nearfield zone.

The conventional probes that are focused are typically duals or straight-beam. Using PA as a straight scan, without any sectoral sweeping, seems to me like using an elephant gun to hunt mice. It works just fine, but simpler & cheaper methods work just as well.

Where PA 'shines' is with a 32- or 64-element probe, running a sectoral sweep, focused at [or just beyond] anticipated defects and simultaniously scanning for laminations and pitting [or crack tips, scanning at 45 deg's] by using a series of parallel beams generated by a series of 8-to-16 element 'virtual' probes. A linear 64-element transducer can generate up to 49 16-element 'virtual' probes all performing a 'ladder shaped' series of linear scans [64-16, +1]. It is all in "the power of the number of elements"

 
 Reply 
 
TomNelligan
Engineering,
retired, USA, Joined Nov 1998, 390

TomNelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
22:30 Jul-07-2009
Re: PA probe focusing
In Reply to Stan at 16:44 Jul-07-2009 (Opening).

A phased array transducer cannot be focused beyond its near field, and in the case of common square and rectangular probes, focusing occurs in the steering direction only. You can find some further information in the focusing section of our phased array tutorial at

http://www.olympus-ims.com/en/ndt-tutorials/transducers/focusing/

 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1285

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1285
23:52 Jul-07-2009
Re: PA probe focusing
In Reply to Stan at 16:44 Jul-07-2009 (Opening).

Stan:
There was a thread last month about a similar topic.
http://www.ndt.net/forum/thread.php?admin=&msgID=31403&rootID=8596#31403
The discussion related to where the near field was on a rectangular probe (which most linear phased-array probes used today are duplicating).
The principles involved in focusing are still valid (i.e. only in the near field and the focal spot gets smaller as the focal point is brought closer to the probe). There have been some really nice results (especially for zonal discrimination techniques) whereby the passive aperture axis is curved and the active aperture focused in the normal fashion for a phased-array.
See http://www.ndt.net/article/ecndt2006/doc/Fr.1.1.1.pdf
and http://www.ndt.net/article/panndt2007/papers/146.pdf

 
 Reply 
 

Product Spotlight

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

Echomac® Small

Available with up to eight channels of electronics to detect and evaluate thickness, flaws and eccen
...
tricity, this UT tester is housed in a convenient, smaller cabinet. This instrument can be used in conjunction with bubbler or immersion tank systems, or with a test bench or in laboratory applications. The Echomac® Small is available in the FD4, FD6 or FD6A versions.
>

Ultrasonic Testing Immersion Tanks with Unmatched Scanning Features

TecScan’s non-destructive testing Ultrasonic Immersion Tanks & scanners are designed for high pe
...
rformance and demanding NDT testing applications. Our Scan3D™ line of High Precision Immersion Tanks are specifically designed for automated ultrasonic testing of complex composites parts used in aerospace and industrial applications.
>

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window