| 14:05 Sep-28-2009 Sensitivity comparison between Shear & Longitudinal wavee Could any one clarify whether there is any sensitivity difference between Shear & Longitudinal wave ,if so Why?Why is Longitudinal wave used in ToFD.Please reply. |
|
|
|
| Nigel Armstrong Engineering, - Specialist services United Kingdom, Joined Oct 2000, 1094 |
| 04:41 Sep-29-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Rupesh Nair at 14:05 Sep-28-2009 (Opening).
Rupesh
One needs to be cautious using the term "sensitivity". The ndt.net encyclopaedia defines sensitivity as "The characteristics of an ultrasonic instrument and probe combination which determine the ability to detect small signals limited by the signal-to-noise ratio. In order to perform ultrasonic systems for flaw detection to the requirements of various products it is necessary to adjust the sensitivity level."
ASTM E-1316 defines senstivity as, "a measure of the smallest ultrasonic signal which
will produce a discernible indication on the display of an ultrasonic system."
The ASNT handbook on ultrasonic testing 3rd edition defines sensitivity as "Ability of signal to change with small changes of measured quantity"
For the 1st part of your question. Using compression and shear waves of equal wavelength (i.e. different frequency crystals) and equal soundpath distance and a side-drilled hole as the reflector, I can only imagine that, without mode conversion (i.e. compression - compression or shear - shear cases), the reflected sound energy difference will be minimal unless the mode of propagation of the incident wave results in more sound energy being diffracted or otherwise diverted away from direct reflection.
For the 2nd part of your question, use of longitudinal waves for TOFD is dictated by the greater sound velocity thus meaning that all reflected shear waves return to the receiving probes after the reflected longitudinal wave from the backwall giving a more easily interpretable area of interest. |
|
|
|
| Rupesh Nair NDT Inspector, Instructor Jubail Industrial College, India, Joined Jul 2008, 5 |
 Rupesh Nair NDT Inspector, Instructor Jubail Industrial College, India, Joined Jul 2008
5
| 09:42 Sep-29-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Nigel Armstrong at 04:41 Sep-29-2009 .
Dear Nigel
Thank you for the reply.
For the first part of the question if the type of defect was a transverse notch will it make a difference in the sensitivity?,as the diffraction & mode conversion phenonmenon are more dominant. |
|
|
|
| Nigel Armstrong Engineering, - Specialist services United Kingdom, Joined Oct 2000, 1094 |
| 11:12 Sep-29-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Rupesh Nair at 09:42 Sep-29-2009 .
Rupesh
May I suggest a literature search using such names as JA Ogilvie, GA Georgiou, JM Coffey, RK Chapman amongst many others who have written papers on the complex subject of soundfield - defect interaction and scattering phenomenon.
Good luck and best regards |
|
|
|
| Slawomir Mackiewicz
 |
Slawomir Mackiewicz
 | 13:08 Sep-29-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Rupesh Nair at 14:05 Sep-28-2009 (Opening).
The sensitivity definitions cited by Nigel are too general to give a definite answer to the Rupesh question. But if we define the sensitivity of an ultrasonic system as a relative echo amplitude from the reflector of a given shape, size and orientation as compared to the BW echo received from the same distance we can say something about relative sensitivities of L and T waves.
The general answer following from the elasticity theory is Yes, the amplitude of reflected/back scattered wave depends on the mode (L or T) of incident wave. The strict solutions for cylinders and spheres are available in advanced books on elastic wave propagation theory. I could also risk a statement that generally (for most reflectors) the share wave is more sensitive than longitudinal one assuming both wave have the same wavelengths.
The explanation is simple and follows directly from the Snells Law. Just consider that L wave undergoes partial transformation (energy loss) for every incident angle on the reflector surface except 0 deg. The T wave undergoes transformation only for incidence angles less than critical angle, 33 deg for steel-air interface. So statistically energy losses on reflections due to transformation are grater for L waves than for T waves.
Having this in mind you can predict that relative sensitivity of T waves will be better than L waves when detecting SDH and notches but comparable to L waves when detecting FBH.
Best Regard |
|
|
|
| Nigel Armstrong Engineering, - Specialist services United Kingdom, Joined Oct 2000, 1094 |
| 16:25 Sep-29-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Slawomir Mackiewicz at 13:08 Sep-29-2009 .
Rupesh
Are your two questions connected? It reads very much as if you wish to present a case for use of shear waves for TOFD inspections. |
|
|
|
| Rupesh Nair NDT Inspector, Instructor Jubail Industrial College, India, Joined Jul 2008, 5 |
 Rupesh Nair NDT Inspector, Instructor Jubail Industrial College, India, Joined Jul 2008
5
| 19:28 Sep-29-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Nigel Armstrong at 16:25 Sep-29-2009 .
Nigel
Your guess is right. As with slawomir"s explanation if the shear component is more sensitive than the longitudinal wave(with respect to the energy transformation loss ) then why cant we use the shear part of the wave in ToFD for resolution & sizing. It is possible to view the shear part even after the L-component wave.
Best regards |
|
|
|
| Slawomir Mackiewicz
 |
Slawomir Mackiewicz
 | 12:51 Sep-30-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Rupesh Nair at 19:28 Sep-29-2009 .
Rupesh
Your question - Why cant we use the shear wave for TOFD inspection is quite reasonable and worth of detailed discussion. The Nigels argument that the L-wave is faster does not prohibit the use of T-wave. You can use, for example, the standard T60 shear wave probe and have no problem with interfering L-wave, because it doesnt exist in this case.
The main problem with T-wave in TOFD is that the so called lateral wave is closely followed by interfering Rayleigh surface wave traveling with slightly lower speed (about 7% for steel) than the share wave. Due to this effect you have unwanted waveforms trailing the actual lateral wave and pretending or obscuring relevant indications from the volume of tested material. Of course there are some other problems connected with directivity patterns of diffracted waves but in my opinion they are not critical.
Best Regards |
|
|
|
| Rupesh Nair NDT Inspector, Instructor Jubail Industrial College, India, Joined Jul 2008, 5 |
 Rupesh Nair NDT Inspector, Instructor Jubail Industrial College, India, Joined Jul 2008
5
| 13:23 Sep-30-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Slawomir Mackiewicz at 12:51 Sep-30-2009 .
Slawomir
Thank you for the detailed reply.Even with the presence of rayleigh wave a depth equivalent to approx. 1 Lamda may get misinterpreted,which even is missed with respect to the Lateral wave.The chances of detecting & evaluating the surface breaking defects may get in uncertainity.But does it effect the volume embedded indications.Please comment.
Regards
Rupesh |
|
|
|
| Nigel Armstrong Engineering, - Specialist services United Kingdom, Joined Oct 2000, 1094 |
| 15:05 Sep-30-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Rupesh Nair at 13:23 Sep-30-2009 .
Rupesh/Slawomir
A Google search on the phrase "TOFD using shear wave probes" brings up some interesting papers on the topic. G. Baskaran, Krishnan Balasubramaniam and C.V. Krishnamurthy report on success with this method for thin-walled materials (6 - 7 mm).
I would advise very careful validation of any such inspection especially due to the lower diffraction coefficient and beamspread of shear wave from those of longitudinal wave.
do you have a specific application in mind Rupesh? |
|
|
|
| Rolf Director, Editor NDT.net, Germany, Joined Nov 1998, 603  |
 Rolf Director, Editor NDT.net, Germany, Joined Nov 1998
603 
| 15:44 Sep-30-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Nigel Armstrong at 15:05 Sep-30-2009 .
Nigel, I belive you are using the same Google :-)
No results found for "TOFD using shear wave probes".
Results for TOFD using shear wave probes (without quotes): |
|
|
|
| Nigel Armstrong Engineering, - Specialist services United Kingdom, Joined Oct 2000, 1094 |
| 19:30 Sep-30-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Rolf at 15:44 Sep-30-2009 .
Thanks for the corrrection Rolf. No quotation marks Rupesh! |
|
|
|
| Rupesh Nair NDT Inspector, Instructor Jubail Industrial College, India, Joined Jul 2008, 5 |
 Rupesh Nair NDT Inspector, Instructor Jubail Industrial College, India, Joined Jul 2008
5
| 19:35 Sep-30-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Rolf at 15:44 Sep-30-2009 .
Nigel
I dont have any special application in mind,but the usage of L-wave literally generated a feeling of having some better advantages in usage as in comparison with S-wave .As I was unware of that I put it in discussion.
Anyway I found the aricle posted in the web by the name"TOFD using shear wave probes".I have put a request to the author (Prof.Krishnan Balasubramaniam )whom I have met before to get his precious comments if possible.Thank you for the guidance Nigel.
Regards |
|
|
|
| Nigel Armstrong Engineering, - Specialist services United Kingdom, Joined Oct 2000, 1094 |
| 02:35 Oct-01-2009 Re: Sensitivity comparison between Shear & Longitudinal wavee In Reply to Rupesh Nair at 19:35 Sep-30-2009 .
Thanks to you too, Rupesh for starting the discussion. From that paper it seeems that S-TOFD has applications. Many of us accept the givens (i.e. TOFD = L-wave) without questioning why its a given and does it always have to be so.
Cheers! |
|
|
|
 High Resolution Industrial CT System for Small/Medium Size Parts InspectionThe YXLON FF35 CT computed tomography system is designed to achieve extremely precise inspection res ... ults for a wide range of applications. Available in a single or dual tube configuration, it is perfect for very small to medium size parts inspection in the automotive, electronics, aviation, and material science industries. >  Trade-In OfferWe are offering our existing clients trade-in against Wavemaker® G3 systems, Standard (Blue) inflat ... able rings and modules for excellent deals on the latest Wavemaker® G4mini system. >  TVC awarded UKAS accreditationTVC are delighted to finally announce we have been awarded UKAS
accreditation for our calibration ... laboratory.
Laboratory accreditation to ISO/IEC 17025:2005 enables us to
conduct the Electrical Verification of Ultrasonic Flaw Detection
Equipment to BS EN 12668-1:2010. It has taken many months of hard
work and we want to thank our staff for all their efforts during
this massive undertaking. >  Navic - Steerable Modular Automated ScannerThe Navic is a modular, motorized, steerable scanner designed to
carry multiple attachments
used ... in various scanning and inspection applications. The Navic is
capable of weld scanning
(girth welds and long seam welds), automated corrosion mapping,
and tank scanning.
> |
|