Intial pulse in UT instrument

Dear all,

i'm using two UT instruments. The older equipment is giving good intial pulse even at 10mm Range. But the new equipment which i got recently, that is not giving intial pulse at 10mm range. All the parameter settings are same with the comparision of older instrument. But the old instrument is giving good intial pulse. Where as the new instrument is giving multiple echoes instead of intial pulse.

Can anybody explain the reason behind it. Sorry for not revealing the Instrument's NAME.

Please clarify.

    
 
 
ezio
Other, Retired ex Laboratory Technical Manager
OMECO Research Centre, Italy, Joined Sep 2008, 273

ezio

Other, Retired ex Laboratory Technical Manager
OMECO Research Centre,
Italy,
Joined Sep 2008
273
21:05 Mar-02-2010
Re: Intial pulse in UT instrument
In Reply to Phanindra Kumar at 06:31 Mar-02-2010 (Opening).

if the probe is the same the second instrument is broken or wrong used (my opinion)

    
 
 
Michel Couture
NDT Inspector,
consultant, Canada, Joined Sep 2006, 834

Michel Couture

NDT Inspector,
consultant,
Canada,
Joined Sep 2006
834
04:08 Mar-03-2010
Re: Intial pulse in UT instrument
In Reply to ezio at 21:05 Mar-02-2010 .

Phanindra,

It is a little difficult to make a proper diagnostic regarding your new instruement since you do not divulge the make of this new UT set and the old one as well. But one thing that comes to mind with this new instrument is the possibility of the PRF (pulse repetition rate) being too high. If this is the situation, you will have multiple false echo on your screen resulting from the interference of the sound pulse going into the part with the previous one coming back to your transducer.

If yor PRF is adjustable, try adjusting it to about 400 or if it is a Low - Medium - High type selection, try lowering it to a medium setting. With a static inspection, the PRF doesn't need to be very high.

Other thing you are not mentioning either is the frequency and probe size. This may also have an effect on your IP. Change your probe and see if your results are different between the instruments. Like Ezio said, it maybe your probe. This would eliminate it.

    
 
 

Product Spotlight

GEKKO - Portable Phased Array Testing with TFM in Real-Time

The portable phased array testing system GEKKO provides 64 parallel test channels. On creating testi
...
ng parameters the operator is assisted by the CIVA software. Due to its modular set-up the GEKKO instrument is suitable for operators of all skill levels.
>

Compact NDT inspection-heads for measurements with active thermography

The compact inspection head is suitable for thermographic ndt tasks. The uncooled infrared camera
...
is specially developed for NDI-tasks and offers a thermal sensitivity until now known only from thermal imagers with cooled detector. All required components and functions are integrated into the inspection-head. You will only need an ethernet cable to connect the sensor with the evaluation system.
>

NovaScope 6000

The all-digital Novascope 6000 is a portable, ultra-high precision thickness gauge for high-speed
...
thickness measurement. Novascope 6000 has unmatched capabilities and unique features including: •Superior Resolution with high contrast, high-speed color RF display •High pulser voltage •Real-time video output •Increased internal/external data storage •Programmable SetUp features •Battery & AC Powered
>

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window