09:01 Apr-17-2002
Urgent!!!

the following data was obtained
1- the peak number inreases with the increase of transducer (probe ) diameter when the other parameter are constant.
2- The band width increases with decrease of transducer (probe) diameter when the other parameters are constant.

please help getting an interpritation for these two results.


    
 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1261

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1261
00:01 Apr-17-2002
Re: Urgent!!!
Some questions must be answered before this urgent matter can be resolved.
What do you mean by Peak Number.
Is it the number of peaks observed in the near zone?
Is it the peak frequency?
Is it the peak voltage?
Or do you mean centre frequency?
Or something else?

You then state that the bandwidth increases with diameter and claim that all other parameters remain constant.
I SUSPECT the effects of mechanical damping are the main cause of bandwidth variation. But unless you know that the entire construction process of the probe is identical it will be difficult to assert that ALL other parameters are constant.
Perhaps there are parameters varying that you have not considered.
Are these probes contact or immersion evaluated?

Ed

: the following data was obtained
: 1- the peak number inreases with the increase of transducer (probe ) diameter when the other parameter are constant.
: 2- The band width increases with decrease of transducer (probe) diameter when the other parameters are constant.
.
: please help getting an interpritation for these two results.
.



    
 
 Reply 
 
ISRA
Consultant
isr, Brazil, Joined Apr 2002, 3

ISRA

Consultant
isr,
Brazil,
Joined Apr 2002
3
09:37 May-01-2002
Re: Urgent!!!
I sent this mail 10 days before but i don't know why you don't receive it.

Dear expert
Thanks for giving me of your time to reply my question

by peak number?
i mean the number of half peaks have an amplitude greater than 20% of the
hiegst one (observed by viewing the second backwall echo received from the
V1 block on the oscilloscope)

the kind of probes are broad band normal contact probe (supposed to be
typically constructed)

a third question is
is there any limit for the band width% to judge if it is narrow band or wide
band?





    
 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1261

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1261
04:10 May-01-2002
Re: Urgent!!!
ISRA:
The number of "peaks" you define is the number of cycles. This varifies my postulation of mechancial variation due to clamping effects on the element.
Smaller dimensioned elements tend to be easier to clamp and thereby limit ring-time. Increased ring-time results in a higher Q-factor [Q=(fc/(f2-f1)] where f2 and f1 are the upper and lower 6dB drop frequencies from the fc or centre frequency.

For a polycrystalline piezo material you will need more damping to reduce the number of cycles over the referenced amplitude for a larger area element as compared to a smaller area element lapped to the same nominal resonant frequency.

As to specific assignments of bandwidth categories; i.e. what percentage bandwidth constitutes broadband or narrowband I do not know what the convention is. I personnally prefer a broadband probe with 1-1.5 cycles and this is usually about a 75-90% bandwidth. But I do not know if a 2 cycled pulse or something less than 70% bandwidth constitutes a narrow band probe.

Band widths greater than 100% are of course possible. But damping rigid crystalline materials to greater and greater amounts eventually kills all useful output. The best option for a useful high bandwidth signal can be had from PVDF and PVTrF materials. These have a very good receiver efficiency (about 7 times more than PZT).

Another potential source of variation is the electrical matching required if there is different capacitance between the electrodes of the large and small probes.
If using a negative spike pulser you can alter the frequency content of the ring by adjusting the pulse duration.

Why are you using the 'second' backwall signal instead of the first? Since you are testing these probes in contact mode you are coupling energy back into the probe's matching plate. Also, because you are testing the frequency content which is, as just discussed above, dependent on the damping of the element you have added another variable in the pressure put on the probe. Now you must consider effects of front loading as well as back loading.

Pages 65-100 of the NDT Handbook Second edition vol 7 by ASNT may be helpful to your design considerations.
Ed

: I sent this mail 10 days before but i don't know why you don't receive it.
.
: Dear expert
: Thanks for giving me of your time to reply my question
.
: by peak number?
: i mean the number of half peaks have an amplitude greater than 20% of the
: hiegst one (observed by viewing the second backwall echo received from the
: V1 block on the oscilloscope)
.
: the kind of probes are broad band normal contact probe (supposed to be
: typically constructed)
.
: a third question is
: is there any limit for the band width% to judge if it is narrow band or wide
: band?
.
.



    
 
 Reply 
 

Product Spotlight

Varex Imaging Large Field of View (FOV) Digital Detector Arrays (DDAs)

A larger FOV DDA can reduce the space and volume of the X-ray inspection system on the factory floor
...
, enable faster scanning times, better throughput and better resolution images at a lower dose. Customers can also save time and money. With these benefits in mind, Varex Imaging has designed a family of large FOV detectors (4343HE, XRD 1611, 4343DX-I, 4343CT) for our industrial imaging customers.
>

NDT Master Lecturer

In the program both university professors and practitioners will give lectures, which guarantees the
...
oretical depth and practical inside. Academic Director: Prof. Dr. Christian Boller
The following lecturers are not complete: Prof. Tadeusz Stepinski, Prof. Wieslaw Staszewski, Prof. Frank Walther, Prof. Giovanni Bruno, Prof. Gerd Dobmann, Prof. Philippe Guy
>

FAAST-PA! OEM Patented phased Array for high speed UT inspection

Multiangle, Multifocus, Multifrequency, Multibeam. Instead of stacking UT electronics and having m
...
any PA probes, FAAST-PA is able to transmit all delay laws within ONE single shot in Real time.
>

FD800 Bench Top Flaw Detectors

The bench-top FD800 flaw detector range combines state-of-the-art flaw detection with advanced mater
...
ial thickness capabilities. Designed for use in the laboratory these gauges are the tool you need for all your flaw detecting needs.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window