Flaw Detection based on 2:1 Noise to Flaw Ratio

Does anyone have any information on this? We are using a 2:1 base metal-to-defect ratio for detection in Friction Stir Weldments. Example: lack-of-penetration is defined when the indication is 6dB higher (2:1) than the noise floor in base metal.

Any information on techniques that use this would be helpful.

Thanks -

Jeff


    
 
 
Godfrey Hands
Engineering,
PRI Nadcap, United Kingdom, Joined Nov 1998, 287

Godfrey Hands

Engineering,
PRI Nadcap,
United Kingdom,
Joined Nov 1998
287
09:39 Jun-15-2002
Re: Flaw Detection based on 2:1 Noise to Flaw Ratio
: Does anyone have any information on this? We are using a 2:1 base metal-to-defect ratio for detection in Friction Stir Weldments. Example: lack-of-penetration is defined when the indication is 6dB higher (2:1) than the noise floor in base metal.
.
: Any information on techniques that use this would be helpful.
.
: Thanks -

Dear Jeff,
The problem with a 2:1 Flaw/Noise ratio is that the noise can add to the flaw amplitude or subtract from the flaw amplitude. This means that in the worst case scenario where the flaw amplitude (independent of noise) is twice the noise amplitude, and this is then reduced in amplitude by the noise, you have no resultant signal left above the noise.
Conversely, when the amplitude is just below the "threshold" without noise (i.e. acceptable imperfection), and is added to by the noise, this then exceeds the threshold and becomes an apparently rejectable condition.
A 3:1 ratio is more often used, as this gives closer calls to the real condition.
Is thereany way that you can improve the signal to noise ratio and work with 3:1 ? e.g. Higher frequency probes etc ?

Good Luck,

Godfrey Hands
.
: Jeff
.



    
 
 
David Ponder
David Ponder
07:04 Jun-15-2002
Re: Flaw Detection based on 2:1 Noise to Flaw Ratio
: : Does anyone have any information on this? We are using a 2:1 base metal-to-defect ratio for detection in Friction Stir Weldments. Example: lack-of-penetration is defined when the indication is 6dB higher (2:1) than the noise floor in base metal.
: .
: : Any information on techniques that use this would be helpful.
: .
: : Thanks -
.
: Dear Jeff,
: The problem with a 2:1 Flaw/Noise ratio is that the noise can add to the flaw amplitude or subtract from the flaw amplitude. This means that in the worst case scenario where the flaw amplitude (independent of noise) is twice the noise amplitude, and this is then reduced in amplitude by the noise, you have no resultant signal left above the noise.
: Conversely, when the amplitude is just below the "threshold" without noise (i.e. acceptable imperfection), and is added to by the noise, this then exceeds the threshold and becomes an apparently rejectable condition.
: A 3:1 ratio is more often used, as this gives closer calls to the real condition.
: Is there any way that you can improve the signal to noise ratio and work with 3:1 ? e.g. Higher frequency probes etc ?
.
: Good Luck,
.
: Godfrey Hands
: .
: : Jeff
: .
.Dear Fellow Technicians

I think that increasing the frequency of the probes will only increase the metal noise. The heat affected zone will give you all kinds of false indications with higher frequency. You will not be able to tell the difference between an actual defect and grain boundaries in the material. If you even have the option go with lower frequency. If the data you are using is anything like the AWS specs all they are saying is, any indications that are 6dB higher than the metal noise shall be considered lack of penetration. AWS uses alot of the same type stuff, check AWS D1.5



    
 
 
Terry Oldberg
Engineering, Mechanical Electrical Nuclear Software
Consultant, USA, Joined Oct 1999, 42

Terry Oldberg

Engineering, Mechanical Electrical Nuclear Software
Consultant,
USA,
Joined Oct 1999
42
09:14 Jun-29-2002
Re: Flaw Detection based on 2:1 Noise to Flaw Ratio
The subject of this posting contains an internal inconsistency that makes the subject itself nonsensical. In particular, "noise" and the related concept "signal" are defined statistically but flaw detection tests do not obey statistics ( see http://www.ndt.net/article/v04n05/oldberg/oldberg.htm ).




    
 
 

Product Spotlight

Research Center IDEKO develops an ultrasonic train wheel inspection system for CAF

The Basque Research centre IK4-IDEKO has developed a state-of-the-art ultrasonic inspection system f
...
or the train wheel. This system secures a sound condition of train wheels and is thus a significant contribution to rail transport safety. The device was delivered to CAF recently and its use allows the manufacturer to become an approved supplier of rolling stock in Italy, as it meets the demanding homologation standards of this country. \\\\r\\\\n
>

Trade-In Offer

We are offering our existing clients trade-in against Wavemaker® G3 systems, Standard (Blue) inflat
...
able rings and modules for excellent deals on the latest Wavemaker® G4mini system.
>

EPOCH 6LT

The EPOCH 6LT flaw detector is optimized for one-handed operation in rope access and high-portab
...
ility applications. Lightweight with a grip-oriented weight distribution, the instrument fits securely in a user’s hand or can be strapped to their leg using the rope access accessory kit. The clear, bright screen makes A-scans readable in any light while the intuitive software enables fast navigation.
>

GEKKO, Standard and Advanced phased-array for Easier inspection

With a 64-channel parallel architecture, GEKKO is a flaw detector offering at the same time conven
...
tional UT, standard PAUT, TOFD and real-time Total Focusing Method (TFM). GEKKO improves the detection, with better defects characterization and sizing as well as for misoriented defects. Some procedures require standard UT, others TOFD or PA. GEKKO includes all these techniques to offer a versatile and field-ready equipment. - 64:64 parallel channels - + 4 additional TOFD - Conventional UT channels - International code compliance: ASME, AWS, API, ASTM, ISO-EN Watch the video: https://youtu.be/kuEY_RwWS3Q
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window
s