Technical Discussions
yasseen
yasseen
01:59 Feb-05-2001
Re: Sound velocity measurements
: information on sound measuring equipement
: : Accuracy of sound velocity measurements( <=0.1% ?)

: : During the revision of the ISO 2400 Standard for the IIW block Nr. 1 we had to discuss
: : the problem of the accuracy at sound velocity measurements achievable with standard laboratory equipments.
: : This debate showed us that the difficulties with this problem tend to be underestimated. An absolute accuracy
: : of 0.1 % can be realized only with a certain amount care during the measurements.
: : As a reference I would like to indicate a contribution presented at the ASTM New Methods Seminar
: : E0-7.98 from January 1998 from Chris Fortunko:
: : C.M. Fortunko, T.P. Lerch and M. C. Renken
: : Velocity Measurements in IIW-Type Blocks
: : Materials Science and Engineering Laboratory
: : National Institute of Standards and Technology
: : 325 Broadway, Boulder, CO 80303
: : IIW Doc Nr. Vc-927/98/OE

: I have been working with Resonant Methods of NDT for the past few years.
: As you know, resonances depend upon the ultrasound velocity in materials.
: We are able to measure resonant frequencies with a repeatability in the order of 0.05%, and frequently much better (up to 0.005% measured some times).
: This will measure the transverse as well as the longitudinal velocities of the material.

: Contact me if you need more details.




    
 
 
Hermann Wuestenberg
R & D
BAM Berlin, Germany, Joined Nov 1998, 26

Hermann Wuestenberg

R & D
BAM Berlin,
Germany,
Joined Nov 1998
26
07:44 Nov-09-1998
Sound velocity measurements

Accuracy of sound velocity measurements( <=0.1% ?)

During the revision of the ISO 2400 Standard for the IIW block Nr. 1 we had to discuss
the problem of the accuracy at sound velocity measurements achievable with standard laboratory equipments.
This debate showed us that the difficulties with this problem tend to be underestimated. An absolute accuracy
of 0.1 % can be realized only with a certain amount care during the measurements.
As a reference I would like to indicate a contribution presented at the ASTM New Methods Seminar
E0-7.98 from January 1998 from Chris Fortunko:
C.M. Fortunko, T.P. Lerch and M. C. Renken
Velocity Measurements in IIW-Type Blocks
Materials Science and Engineering Laboratory
National Institute of Standards and Technology
325 Broadway, Boulder, CO 80303
IIW Doc Nr. Vc-927/98/OE



    
 
 
hendiy
hendiy
06:46 Nov-28-1998
Re: Sound velocity measurements
: Accuracy of sound velocity measurements( <=0.1% ?)

: During the revision of the ISO 2400 Standard for the IIW block Nr. 1 we had to discuss
: the problem of the accuracy at sound velocity measurements achievable with standard laboratory equipments.
: This debate showed us that the difficulties with this problem tend to be underestimated. An absolute accuracy
: of 0.1 % can be realized only with a certain amount care during the measurements.
: As a reference I would like to indicate a contribution presented at the ASTM New Methods Seminar
: E0-7.98 from January 1998 from Chris Fortunko:
: C.M. Fortunko, T.P. Lerch and M. C. Renken
: Velocity Measurements in IIW-Type Blocks
: Materials Science and Engineering Laboratory
: National Institute of Standards and Technology
: 325 Broadway, Boulder, CO 80303
: IIW Doc Nr. Vc-927/98/OE




    
 
 
Godfrey Hands
Engineering,
PRI Nadcap, United Kingdom, Joined Nov 1998, 291

Godfrey Hands

Engineering,
PRI Nadcap,
United Kingdom,
Joined Nov 1998
291
09:35 Dec-04-1998
Re: Sound velocity measurements

: Accuracy of sound velocity measurements( <=0.1% ?)

: During the revision of the ISO 2400 Standard for the IIW block Nr. 1 we had to discuss
: the problem of the accuracy at sound velocity measurements achievable with standard laboratory equipments.
: This debate showed us that the difficulties with this problem tend to be underestimated. An absolute accuracy
: of 0.1 % can be realized only with a certain amount care during the measurements.
: As a reference I would like to indicate a contribution presented at the ASTM New Methods Seminar
: E0-7.98 from January 1998 from Chris Fortunko:
: C.M. Fortunko, T.P. Lerch and M. C. Renken
: Velocity Measurements in IIW-Type Blocks
: Materials Science and Engineering Laboratory
: National Institute of Standards and Technology
: 325 Broadway, Boulder, CO 80303
: IIW Doc Nr. Vc-927/98/OE

I have been working with Resonant Methods of NDT for the past few years.
As you know, resonances depend upon the ultrasound velocity in materials.
We are able to measure resonant frequencies with a repeatability in the order of 0.05%, and frequently much better (up to 0.005% measured some times).
This will measure the transverse as well as the longitudinal velocities of the material.

Contact me if you need more details.


    
 
 

Product Spotlight

NDTkit RT

NDTkit RT, TESTIA's Digital Radiography software The NDTkit product line software for X-ray analysi
...
s. NDTkit RT is a software benefiting from the Ultis kernel which is dedicated to radiographic image analysis. It offers a set of tools and filtering processes to assist RT operators in finding relevant flaws.
>

FAAST-PA! OEM Patented phased Array for high speed UT inspection

Multiangle, Multifocus, Multifrequency, Multibeam. Instead of stacking UT electronics and having m
...
any PA probes, FAAST-PA is able to transmit all delay laws within ONE single shot in Real time.
>

Exertus Dual 120

The Exertus Dual 120 Projector has the ability to accept Iridium 192 sources or Selenium 75 source
...
s. This projector incorporates design and safety features that make it flexible, compact and lightweight. The Projector is lighter than most of its competitors. It incorporates an improved source channel, based on a new helicoidal design, which makes maintenance easier. The helicoidal design also allows smoother movement of the source assembly inside the device, making it easier for the operator and improving safety. The Projector also has a unique safety feature not found in competitive products. The source assembly locking mechanism is triggered by the source holder capsule at the front of the source assembly, thereby always assuring the operator that the source has returned to the safe position. The Exertus Dual 120 is ISO3999:2004 compliant.
>

Teletest Focus+

Teletest Focus+ electronics have superior capabilities than rivals on the market. Beyond the usual
...
test features, Focus+ has 24 transmit channels and 24 receive channels, with an additional on-board switching capabilities. The instrument's frequency range is 10–300 kHz.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window