yasseen
yasseen
01:59 Feb-05-2001
Re: Sound velocity measurements
: information on sound measuring equipement
: : Accuracy of sound velocity measurements( <=0.1% ?)

: : During the revision of the ISO 2400 Standard for the IIW block Nr. 1 we had to discuss
: : the problem of the accuracy at sound velocity measurements achievable with standard laboratory equipments.
: : This debate showed us that the difficulties with this problem tend to be underestimated. An absolute accuracy
: : of 0.1 % can be realized only with a certain amount care during the measurements.
: : As a reference I would like to indicate a contribution presented at the ASTM New Methods Seminar
: : E0-7.98 from January 1998 from Chris Fortunko:
: : C.M. Fortunko, T.P. Lerch and M. C. Renken
: : Velocity Measurements in IIW-Type Blocks
: : Materials Science and Engineering Laboratory
: : National Institute of Standards and Technology
: : 325 Broadway, Boulder, CO 80303
: : IIW Doc Nr. Vc-927/98/OE

: I have been working with Resonant Methods of NDT for the past few years.
: As you know, resonances depend upon the ultrasound velocity in materials.
: We are able to measure resonant frequencies with a repeatability in the order of 0.05%, and frequently much better (up to 0.005% measured some times).
: This will measure the transverse as well as the longitudinal velocities of the material.

: Contact me if you need more details.




    
 
 Reply 
 
Hermann Wuestenberg
R & D
BAM Berlin, Germany, Joined Nov 1998, 26

Hermann Wuestenberg

R & D
BAM Berlin,
Germany,
Joined Nov 1998
26
07:44 Nov-09-1998
Sound velocity measurements

Accuracy of sound velocity measurements( <=0.1% ?)

During the revision of the ISO 2400 Standard for the IIW block Nr. 1 we had to discuss
the problem of the accuracy at sound velocity measurements achievable with standard laboratory equipments.
This debate showed us that the difficulties with this problem tend to be underestimated. An absolute accuracy
of 0.1 % can be realized only with a certain amount care during the measurements.
As a reference I would like to indicate a contribution presented at the ASTM New Methods Seminar
E0-7.98 from January 1998 from Chris Fortunko:
C.M. Fortunko, T.P. Lerch and M. C. Renken
Velocity Measurements in IIW-Type Blocks
Materials Science and Engineering Laboratory
National Institute of Standards and Technology
325 Broadway, Boulder, CO 80303
IIW Doc Nr. Vc-927/98/OE



    
 
 Reply 
 
hendiy
hendiy
06:46 Nov-28-1998
Re: Sound velocity measurements
: Accuracy of sound velocity measurements( <=0.1% ?)

: During the revision of the ISO 2400 Standard for the IIW block Nr. 1 we had to discuss
: the problem of the accuracy at sound velocity measurements achievable with standard laboratory equipments.
: This debate showed us that the difficulties with this problem tend to be underestimated. An absolute accuracy
: of 0.1 % can be realized only with a certain amount care during the measurements.
: As a reference I would like to indicate a contribution presented at the ASTM New Methods Seminar
: E0-7.98 from January 1998 from Chris Fortunko:
: C.M. Fortunko, T.P. Lerch and M. C. Renken
: Velocity Measurements in IIW-Type Blocks
: Materials Science and Engineering Laboratory
: National Institute of Standards and Technology
: 325 Broadway, Boulder, CO 80303
: IIW Doc Nr. Vc-927/98/OE




    
 
 Reply 
 
Godfrey Hands
Consultant,
PRI Nadcap, United Kingdom, Joined Nov 1998, 301

Godfrey Hands

Consultant,
PRI Nadcap,
United Kingdom,
Joined Nov 1998
301
09:35 Dec-04-1998
Re: Sound velocity measurements

: Accuracy of sound velocity measurements( <=0.1% ?)

: During the revision of the ISO 2400 Standard for the IIW block Nr. 1 we had to discuss
: the problem of the accuracy at sound velocity measurements achievable with standard laboratory equipments.
: This debate showed us that the difficulties with this problem tend to be underestimated. An absolute accuracy
: of 0.1 % can be realized only with a certain amount care during the measurements.
: As a reference I would like to indicate a contribution presented at the ASTM New Methods Seminar
: E0-7.98 from January 1998 from Chris Fortunko:
: C.M. Fortunko, T.P. Lerch and M. C. Renken
: Velocity Measurements in IIW-Type Blocks
: Materials Science and Engineering Laboratory
: National Institute of Standards and Technology
: 325 Broadway, Boulder, CO 80303
: IIW Doc Nr. Vc-927/98/OE

I have been working with Resonant Methods of NDT for the past few years.
As you know, resonances depend upon the ultrasound velocity in materials.
We are able to measure resonant frequencies with a repeatability in the order of 0.05%, and frequently much better (up to 0.005% measured some times).
This will measure the transverse as well as the longitudinal velocities of the material.

Contact me if you need more details.


    
 
 Reply 
 

Product Spotlight

Ultrasonic Squirter/Gantry System

TecScan’s NDT Ultrasonic Gantry Systems are industrial Squirter scanners designed for non-destru
...
ctive quality testing and raster scanning of large structures and parts. The Gantry Systems are usually composed of an Industrial Mechanical Scanner and a Control Room with System Workstation.
>

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

Compact NDT inspection-heads for measurements with active thermography

The compact inspection head is suitable for thermographic ndt tasks. The uncooled infrared camera
...
is specially developed for NDI-tasks and offers a thermal sensitivity until now known only from thermal imagers with cooled detector. All required components and functions are integrated into the inspection-head. You will only need an ethernet cable to connect the sensor with the evaluation system.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window