Calibration for DAC (UT)

I would like to use the standardized DAC-diagrams for UT-inspection of welds.
Therefore, for the sensitivity calibration I need to establish the dB value of the back wall. For angle beam probes, it is recommended to use the radii of calibration blocks.
My intention is to perform the sensitivity calibration at the test piece (sheet metal with parallel surfaces)by performing a V-or W- transmission similar to the first part of the transfer correction procedure.
Here is my question:
I am not sure if I can use the obtained signal for a back wall signal of the corresponding sound path (i.e. I set the dB to 80% FSH and add the neccesdary dB-values defined in the standard DAC-diagram to get the neccesary search sensitivity).
If not, is there a possibilty to use correction factors similar to the correction factors for the radii on the calibration blocks. If so, what are the values of these factors.



    
 
 Reply 
 
Udo Schlengrmann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 177

Udo Schlengrmann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
177
04:37 Aug-21-2002
Re: Calibration for DAC (UT)
Reply by Udo Schlengermann:
You obviously mean DGS-method, resp. DGS-diagram and not DAC-curves.
If you talk about V- or W-path transmission I guess you mean through transmission using two separate angle-beam probes of the same type, one transmitting and one receiving.
This through transmission in most cases does not give a signal equal to a backwall echo, because reflection at the surfaces of the plate is not total (especially for beam angles away from 45°) and because a plate is not a solid semi-infinite body compared to the beam size. Additional reflections cause mode conversion, i.e. transfer of energy to other wave modes than to transverse waves used for evalution.

For weld testing with the DGS-method you have to use a single probe and and a reference echo from a known reflector. It has not necessarily to be a backwall, it can be a disk-shaped reflector of known size.
There are recommendations for the minimum sizes of reference blocks and reflector positions.
European Standard EN 583-2(UT - Sensitivity and range setting) gives guidance on this matter.
The V- and W-path calibration is not recommended in standards.

Kind regards
Udo Schlengermann
Agfa NDT GmbH
+++
: I would like to use the standardized DAC-diagrams for UT-inspection of welds.
: Therefore, for the sensitivity calibration I need to establish the dB value of the back wall. For angle beam probes, it is recommended to use the radii of calibration blocks.
: My intention is to perform the sensitivity calibration at the test piece (sheet metal with parallel surfaces)by performing a V-or W- transmission similar to the first part of the transfer correction procedure.
: Here is my question:
: I am not sure if I can use the obtained signal for a back wall signal of the corresponding sound path (i.e. I set the dB to 80% FSH and add the neccesdary dB-values defined in the standard DAC-diagram to get the neccesary search sensitivity).
: If not, is there a possibilty to use correction factors similar to the correction factors for the radii on the calibration blocks. If so, what are the values of these factors.
.



    
 
 Reply 
 
Harald Germ
, Joined ,

Harald Germ

,
Joined

05:35 Aug-21-2002
Re: Calibration for DAC (UT)
Reply by Harald Germ
Thank you for your reply.
Your are right, I meant the DGS-method. The reason why I want to use the method I discribed below for setting the sensitivity is that with the standard method with reference blocks, you have to make a transfer correction get reasonable results. Additionally, in my understanding, the calibration with a reference block such as K1 or K2 has also some range of error.
Therefore, I thought that if I use the arrangement for the transfer correction, I would have taken into account the couplinmg losses and will not run into problems with different accoustic properties between the reference block and the test piece. So, although the signal I use for calibration is not ideally from a backwall, I think that the error might be in the same range as for a calibration with a reference block.
Could you please give a short comment if I am wrong with this assumption.

Thank you in advance!

: Reply by Udo Schlengermann:
: You obviously mean DGS-method, resp. DGS-diagram and not DAC-curves.
: If you talk about V- or W-path transmission I guess you mean through transmission using two separate angle-beam probes of the same type, one transmitting and one receiving.
: This through transmission in most cases does not give a signal equal to a backwall echo, because reflection at the surfaces of the plate is not total (especially for beam angles away from 45°) and because a plate is not a solid semi-infinite body compared to the beam size. Additional reflections cause mode conversion, i.e. transfer of energy to other wave modes than to transverse waves used for evalution.
.
: For weld testing with the DGS-method you have to use a single probe and and a reference echo from a known reflector. It has not necessarily to be a backwall, it can be a disk-shaped reflector of known size.
: There are recommendations for the minimum sizes of reference blocks and reflector positions.
: European Standard EN 583-2 (UT - Sensitivity and range setting) gives guidance on this matter.
: The V- and W-path calibration is not recommended in standards.
.
: Kind regards
: Udo Schlengermann
: Agfa NDT GmbH
: +++
: : I would like to use the standardized DAC-diagrams for UT-inspection of welds.
: : Therefore, for the sensitivity calibration I need to establish the dB value of the back wall. For angle beam probes, it is recommended to use the radii of calibration blocks.
: : My intention is to perform the sensitivity calibration at the test piece (sheet metal with parallel surfaces)by performing a V-or W- transmission similar to the first part of the transfer correction procedure.
: : Here is my question:
: : I am not sure if I can use the obtained signal for a back wall signal of the corresponding sound path (i.e. I set the dB to 80% FSH and add the neccesdary dB-values defined in the standard DAC-diagram to get the neccesary search sensitivity).
: : If not, is there a possibilty to use correction factors similar to the correction factors for the radii on the calibration blocks. If so, what are the values of these factors.
: .
.



    
 
 Reply 
 

Product Spotlight

PROlineTOP Plug & Play Ultrasonic inspection device

As Plug & Play solution it units all control and operation elements in a small housing and therefo
...
re replaces the typical control cabinet...
>

Immersion systems

ScanMaster ultrasonic immersion systems are designed for high throughput, multi shift operation in a
...
n industrial or lab environment. These fully integrated systems provide various scanning configurations and incorporate conventional and phased arrays technologies to support diverse applications, such as inspection of disks, bars, shafts, billets and plates. All of ScanMaster immersion systems are built from high accuracy scanning frames allowing for scanning of complex parts and include a multi-channel ultrasonic instrument with exceptional performance. The systems are approved by all major manufacturers for C-scan inspection of jet engine forged discs. Together with a comprehensive set of software modules these flexible series of systems provide the customer with the best price performance solutions.
>

YXLON Cougar EVO

Scalable small footprint X-ray inspection systems for assembly and laboratory applications. The
...
YXLON Cougar EVO series was designed to provide the "best-in- class" inspection solutions for SMT, semiconductor, and laboratory assembly applications, while maintaining a small system footprint for maximum convenience. With optimized software and hardware, these systems produce higher quality and more consistent results than other electronics inspection systems currently on the market.
>

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window