Fangjun Hong
Fangjun Hong
00:26 Sep-09-2002
How to detect the solid/liquid metal interface? (re-post)

I want to determine the postion of interface of liquid/solid alloy-158 metal. the attched is my experimental setup. I use Panamatrics 5900 pulser/reciver. I found out I canot detect the echo from the interface of liquid/solid metal, but the signal from the top of solid metal. because the maxium gain of 5900 pulser/receiver is 54db, so I am wondering whether the gain is not high enough to detect the singal. if I add a singal amplifier, could I detect the singal?
is the frequency of the transducer I used too high?
Could anyone give me suggestion on my experimental setup, for example, the selection of the transducer, puler/receiver, and etc.





 
 Reply 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
03:24 Sep-09-2002
Re: How to detect the solid/liquid metal interface? (re-post)
The Panametrics Model 5900PR that you are using is not really designed for applications requiring very high sensitivity. Rather, that instrument is optimized for use with very high frequency transducers (50 MHz and higher)in applications where thin material resolution, rather than penetration, is most important.

Because it is optimized for very high frequency transducers, it is going to under-drive the 20 MHz transducer that you are using. You could probably pick up at least a factor of 3 in sensitivity by using a higher energy excitation pulse, which can be obtained from other Panametrics pulsers such as the Model 5800PR or Model 5072PR (as well as some instruments from other manufacturers).

It also occurs to me that since your sound path is fairly long, you could perhaps use a lower frequency transducer (10 MHz, even 5 MHz), whose signal would be less subject to attenuation in the liquid metal. You might also consider a larger diameter transducer if the interface between liquid and solid metal is rough or uneven; a larger diameter may be a little less sensitive to geometrical effects.

--Tom Nelligan
GE Panametrics


 
 Reply 
 
Fangjun Hong
Fangjun Hong
07:30 Sep-10-2002
Re: How to detect the solid/liquid metal interface? (re-post)
: The Panametrics Model 5900PR that you are using is not really designed for applications requiring very high sensitivity. Rather, that instrument is optimized for use with very high frequency transducers (50 MHz and higher)in applications where thin material resolution, rather than penetration, is most important.
.
: Because it is optimized for very high frequency transducers, it is going to under-drive the 20 MHz transducer that you are using. You could probably pick up at least a factor of 3 in sensitivity by using a higher energy excitation pulse, which can be obtained from other Panametrics pulsers such as the Model 5800PR or Model 5072PR (as well as some instruments from other manufacturers).
.
: It also occurs to me that since your sound path is fairly long, you could perhaps use a lower frequency transducer (10 MHz, even 5 MHz), whose signal would be less subject to attenuation in the liquid metal. You might also consider a larger diameter transducer if the interface between liquid and solid metal is rough or uneven; a larger diameter may be a little less sensitive to geometrical effects.
.
: --Tom Nelligan
: GE Panametrics
.
Dear Mr. Tom Nelligan,
Thanks for your suggestion.
I am wondering whether V112-RM 10MHz, 6mm Finger Contact Transducer and V110-RM 5MHz, 6mm is ok for my application.
I also want to use external pulser (Function & Arbitrary Waveform Generator) to excite the transducers with burst tone pulse and use 5900 Receiver to detect the signal. is that OK?
What should I pay attention to when I use external pulser, beside the power limit of the transducer?




 
 Reply 
 

Product Spotlight

Magnetic X-Ray Pipeline Crawler

Zhong Yi brand pipeline crawler is magnetic/micro-video controlled crawler with DC X ray unit inst
...
alled for checking welding of pipeline. Move steadily inside the pipeline 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometers and provide the efficient inspection of the pipeline.
>

Panther

M2M PANTHER is a phased-array equipment designed for both desktop and industrial applications, offer
...
ing unparalleled performance in a compact unit. It combines the speed required for industrial integrated Phased-Array Ultrasound (PAUT) solutions, with the most complete set of total focusing method (TFM) imaging techniques, making it the ultimate tool for R&D and procedure qualification.
>

Conformable wedge transducer

The conformability is obtained with a flexible membrane filled with water between the transducer and
...
the inspected component. The coupling between the membrane and the component requires a small quantity of water or couplant. The conformable wedge combines the acoustic performance of immersion technique with good coupling and low attenuation.
>

YXLON Cougar EVO

Scalable small footprint X-ray inspection systems for assembly and laboratory applications. The
...
YXLON Cougar EVO series was designed to provide the "best-in- class" inspection solutions for SMT, semiconductor, and laboratory assembly applications, while maintaining a small system footprint for maximum convenience. With optimized software and hardware, these systems produce higher quality and more consistent results than other electronics inspection systems currently on the market.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window