monel 400 clad gauging

Can anyone describe the proper setup for ultrasonically gauging
monel 400 cladding with hy80,100 backing steel,using a .5" 5mhz
transducer with a panametrics epoch 3b? The bondline is defined
well enough to get a good response. My problem is a cannot obtain
readings without an error rate of approx .030 one would think one could get much more accurate than this.The min clad is .125
and the min backing is 1.187



    
 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
06:55 Oct-11-2002
Re: monel 400 clad gauging
: Can anyone describe the proper setup for ultrasonically gauging
: monel 400 cladding with hy80,100 backing steel,using a .5" 5mhz
: transducer with a panametrics epoch 3b? The bondline is defined
: well enough to get a good response. My problem is a cannot obtain
: readings without an error rate of approx .030 one would think one could get much more accurate than this.The min clad is .125
: and the min backing is 1.187

If the error is a constant 0.030" across the full thickness range, that sounds like a zero calibration problem. Are you calibrating velocity and zero on two reference standards of the same Monel material that you're measuring? If so, there is also the possibility that the monel/steel boundary is giving you an echo that is phase inverted with respect to an air-backed reference standard. (I don't have exact impedance numbers for those alloys available to check whether that's likely.) You can check this by going to the RF display mode and observing whether the echo inverts between the reference standard and the actual cladding. If so, you need to compensate. There are a couple ways of doing this depending on the exact echo shape, but what I'd probably do is either work in RF mode with gating on both positive and negative lobes of the echo, or alternate between half wave negative and half wave positive rectification as appropriate for each boundary condition. You will also want to look carefully at how the thickness measurement gate interacts with the echo peak to be sure that you're getting consistent detection.

--Tom Nelligan
GE Panametrics


    
 
 

Product Spotlight

Ultrasonic Testing Immersion Tanks with Unmatched Scanning Features

TecScan’s non-destructive testing Ultrasonic Immersion Tanks & scanners are designed for high perf
...
ormance and demanding NDT testing applications. Our Scan3D™ line of High Precision Immersion Tanks are specifically designed for automated ultrasonic testing of complex composites parts used in aerospace and industrial applications.
>

HD-CR 35 NDT Computed Radiography System

Portable high-resolution CR scanner for all radiography applications - weld testing, profile images
...
and aerospace. No matter what type of radiographic testing you are performing, the unique TreFoc Technology of the HD-CR 35 NDT imaging plate scanner always guarantees the highest image quality.
>

Research and Applications Development For NDT

The Research and Applications Development (RAD) group is a newly formed team within Acuren dedicat
...
ed to tackling challenging inspection problems. Our focus is the development of novel, field deployable, advanced inspection techniques for use in cases where standard NDT methods are ineffective. We don't wait for new innovations, we engineer them. From concept to commissioning.
>

Exertus Dual 120

The Exertus Dual 120 Projector has the ability to accept Iridium 192 sources or Selenium 75 source
...
s. This projector incorporates design and safety features that make it flexible, compact and lightweight. The Projector is lighter than most of its competitors. It incorporates an improved source channel, based on a new helicoidal design, which makes maintenance easier. The helicoidal design also allows smoother movement of the source assembly inside the device, making it easier for the operator and improving safety. The Projector also has a unique safety feature not found in competitive products. The source assembly locking mechanism is triggered by the source holder capsule at the front of the source assembly, thereby always assuring the operator that the source has returned to the safe position. The Exertus Dual 120 is ISO3999:2004 compliant.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window
s