monel 400 clad gauging

Can anyone describe the proper setup for ultrasonically gauging
monel 400 cladding with hy80,100 backing steel,using a .5" 5mhz
transducer with a panametrics epoch 3b? The bondline is defined
well enough to get a good response. My problem is a cannot obtain
readings without an error rate of approx .030 one would think one could get much more accurate than this.The min clad is .125
and the min backing is 1.187



    
 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
06:55 Oct-11-2002
Re: monel 400 clad gauging
: Can anyone describe the proper setup for ultrasonically gauging
: monel 400 cladding with hy80,100 backing steel,using a .5" 5mhz
: transducer with a panametrics epoch 3b? The bondline is defined
: well enough to get a good response. My problem is a cannot obtain
: readings without an error rate of approx .030 one would think one could get much more accurate than this.The min clad is .125
: and the min backing is 1.187

If the error is a constant 0.030" across the full thickness range, that sounds like a zero calibration problem. Are you calibrating velocity and zero on two reference standards of the same Monel material that you're measuring? If so, there is also the possibility that the monel/steel boundary is giving you an echo that is phase inverted with respect to an air-backed reference standard. (I don't have exact impedance numbers for those alloys available to check whether that's likely.) You can check this by going to the RF display mode and observing whether the echo inverts between the reference standard and the actual cladding. If so, you need to compensate. There are a couple ways of doing this depending on the exact echo shape, but what I'd probably do is either work in RF mode with gating on both positive and negative lobes of the echo, or alternate between half wave negative and half wave positive rectification as appropriate for each boundary condition. You will also want to look carefully at how the thickness measurement gate interacts with the echo peak to be sure that you're getting consistent detection.

--Tom Nelligan
GE Panametrics


    
 
 

Product Spotlight

Robotic laser shearography enables 100% inspection of complex, flight-critical composite structures

An article in “Composites World Magazine” showcases Non Destructive Testing of aero-structures
...
with Laser Shearography. Over the years Dantec Dynamics has supplied many solutions for the aerospace industry. Referring to specific customer projects several of these cases are examined to outline the advantages of using Laser Shearography for automated defect detection.
>

NDT Master Lecturer

In the program both university professors and practitioners will give lectures, which guarantees the
...
oretical depth and practical inside. Academic Director: Prof. Dr. Christian Boller
The following lecturers are not complete: Prof. Tadeusz Stepinski, Prof. Wieslaw Staszewski, Prof. Frank Walther, Prof. Giovanni Bruno, Prof. Gerd Dobmann, Prof. Philippe Guy
>

PROlineTOP Plug & Play Ultrasonic inspection device

As Plug & Play solution it units all control and operation elements in a small housing and therefo
...
re replaces the typical control cabinet...
>

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window