monel 400 clad gauging

Can anyone describe the proper setup for ultrasonically gauging
monel 400 cladding with hy80,100 backing steel,using a .5" 5mhz
transducer with a panametrics epoch 3b? The bondline is defined
well enough to get a good response. My problem is a cannot obtain
readings without an error rate of approx .030 one would think one could get much more accurate than this.The min clad is .125
and the min backing is 1.187



    
 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
06:55 Oct-11-2002
Re: monel 400 clad gauging
: Can anyone describe the proper setup for ultrasonically gauging
: monel 400 cladding with hy80,100 backing steel,using a .5" 5mhz
: transducer with a panametrics epoch 3b? The bondline is defined
: well enough to get a good response. My problem is a cannot obtain
: readings without an error rate of approx .030 one would think one could get much more accurate than this.The min clad is .125
: and the min backing is 1.187

If the error is a constant 0.030" across the full thickness range, that sounds like a zero calibration problem. Are you calibrating velocity and zero on two reference standards of the same Monel material that you're measuring? If so, there is also the possibility that the monel/steel boundary is giving you an echo that is phase inverted with respect to an air-backed reference standard. (I don't have exact impedance numbers for those alloys available to check whether that's likely.) You can check this by going to the RF display mode and observing whether the echo inverts between the reference standard and the actual cladding. If so, you need to compensate. There are a couple ways of doing this depending on the exact echo shape, but what I'd probably do is either work in RF mode with gating on both positive and negative lobes of the echo, or alternate between half wave negative and half wave positive rectification as appropriate for each boundary condition. You will also want to look carefully at how the thickness measurement gate interacts with the echo peak to be sure that you're getting consistent detection.

--Tom Nelligan
GE Panametrics


    
 
 

Product Spotlight

TVC awarded UKAS accreditation

TVC are delighted to finally announce we have been awarded UKAS accreditation for our calibration
...
laboratory. Laboratory accreditation to ISO/IEC 17025:2005 enables us to conduct the Electrical Verification of Ultrasonic Flaw Detection Equipment to BS EN 12668-1:2010. It has taken many months of hard work and we want to thank our staff for all their efforts during this massive undertaking.
>

Varex Imaging Large Field of View (FOV) Digital Detector Arrays (DDAs)

A larger FOV DDA can reduce the space and volume of the X-ray inspection system on the factory floor
...
, enable faster scanning times, better throughput and better resolution images at a lower dose. Customers can also save time and money. With these benefits in mind, Varex Imaging has designed a family of large FOV detectors (4343HE, XRD 1611, 4343DX-I, 4343CT) for our industrial imaging customers.
>

FlexoFORM™

The FlexoFORM™ solution integrates flexible phased array probe technology in a scanner to solv
...
e the challenges that come with inspecting pipe elbows. When combined with a water column, this solution can be used to collect easy-to-interpret data on elbows with diameters ranging from 4.5 in. OD up to flat and offers many benefits.
>

GEKKO - Portable Phased Array Testing with TFM in Real-Time

The portable phased array testing system GEKKO provides 64 parallel test channels. On creating testi
...
ng parameters the operator is assisted by the CIVA software. Due to its modular set-up the GEKKO instrument is suitable for operators of all skill levels.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window